Mixers also have the curious phenomenon of non-reciprocity
On 10/17/2016 4:13 PM, Bob Camp wrote:
Hi
Simple answer is yes.
More complex answer gets into things like the noise of the mixer (not just it’s
floor),
the levels of the signals, noise being coherent rather than non-coherent, AM
<->
On 10/2/2016 12:56 AM, Adrian Rus wrote:
Hello list,
For those of you interested in phase noise measurement without using
fancy/dedicated gear,
There is no free lunch here. The crystal has its own
intrinsic flicker of frequency noise. You cannot
measure below this noise floor. The
On 9/25/2016 6:53 AM, Bob Camp wrote:
Hi
If you dropped in to a crystal manufacturing plant at any point over the next
30 years after that film
was made, the styles of dress had changed. The crystal holders had changed.
Most of the processes
were still the same and some of the gear was
On 9/25/2016 6:42 PM, Skip Withrow wrote:
1. What the heck is rubidium cell flooding, and how does the TEC in the
5065A fix this problem? None of the -many- rubidium oscillators that I
have been inside before has a TEC, and I have never seen the subject
addressed.
The manual suggests that it
As we all know, step #1 in making a clock is NOT
to build a thermometer :-)
I thought I would check the brain trust here to see
if anyone has seen a hobbyist grade temperature
testing chamber or kit or homebrew design. I
have some crystals, oscillators, and other
electronics I would like to
On 8/21/2016 3:59 PM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:
That said, I don't know why the author is using directional couplers. A
bridge is much wider bandwidth. It is more lossy though.
In general, a resistive bridge will always require a
transformer/180 degree
Another great posting, Attila.
When I was with Agilent, we looked at all kinds of
simplified network analyzer architectures, and I
would have to say the author is really well informed.
One issue he doesn't seem to be aware of is that the
ADL5801, when driven single ended, has some quirks
below
On 8/11/2016 3:47 PM, John Miles wrote:
Right, I'm speaking specifically of L(f). The device being driven by the
oscillator doesn't care about the NF of the driver stage, only what a PN
analyzer would measure at the output jack.
For any 50-ohm source, the practical L(f) floor is -177
Zo doesn't matter for these purposes.
dBm works just as well for 75 ohm systems.
On 8/11/2016 1:22 PM, John Miles wrote:
Remember that L(f) is expressed in dBc/Hz, not dBm/Hz. If it were dBm/Hz, then
kT would be the limit. But in dBc/Hz terms, the limit is 177 + the DUT's
output power in
On 8/11/2016 5:01 AM, Mike Feher wrote:
This may be a naive question, but, how can you achieve results that are so
close to, and sometimes at further out are below kT? Thanks & 73 - Mike
Mike B. Feher, EOZ Inc.
kT per see is not the relevant parameter. It is the
ratio between kT and
On 8/5/2016 12:47 PM, Tom Van Baak wrote:
Here's a new article, on IEEE's site (and this one's free):
"Oscillator Phase Noise: A 50-Year Review", by David B. Leeson
http://ieeexplore.ieee.org/stamp/stamp.jsp?arnumber=7464875
It has always irked me that no credit was given to Edson's
Around 35 years ago, I worked with the guys
who designed and manufactured the 10811. There
are a couple of things here that don't add up,
subject to remembering stuff from a LONG time
ago:
1. Back in those days at least, there were
vendors who supposedly specialized in providing
low noise
On 7/25/2016 10:42 PM, Scott Stobbe wrote:
dramatically different due to glitching on code transition. That being
said, they are kept separate not to confuse sources of error.
FWIW:
The 5071A has a "home brew" DDS that was designed by the late
(and great) Robin Giffard. He used what he
On 7/24/2016 6:44 AM, James Flynn wrote:
Hal Murray writes:
Has anybody put the DAC and all of the analog stuff inside the oven?
Seems
like an obvious idea so somebody has probably patented it.
I am using a 5 MHz custom built design which has everything inside the
outer oven
On 7/22/2016 10:15 AM, David wrote:
It is too bad voltage control of an oscillator cannot be made
ratiometric. Or can it? I have never heard of such a thing. That
would remove some of the demands on a low drift reference.
That's what we tried to do with the E1938A. A multiplying DAC
On 7/22/2016 2:22 AM, Attila Kinali wrote:
Hoi Rick,
On Thu, 21 Jul 2016 18:47:24 -0700
"Richard (Rick) Karlquist" <rich...@karlquist.com> wrote:
Also in 1996, phase microsteppers were already legacy
technology and didn't have a good reputation for spectral
purity. Ano
On 7/21/2016 4:56 PM, Nick Sayer via time-nuts wrote:
Oh my. That’s a bit more than I was originally considering… What I had in mind
was adding a DAC front end to an OCXO so that you could tune the EFC with
serial commands rather than analog and calling that a product.
20 years ago when
I vaguely remember seeing designs where a through hole
transistor like a 2N5179 had a ferrite bead slipped over
the base to keep it stable. Although this works, it
degrades the performance of the transistor. I prefer
to put a resistor in series with the collector instead
of the base. Since the
On 6/29/2016 10:20 AM, Brooke Clarke wrote:
PS Stanford Research is a company founded by physicists and makes some
really high quality stuff. In fact some of the products
HP/Agilent/Keysight sells are repackaged SR instruments.
http://www.prc68.com/I/TandFTE.shtml#SR620
In the old days,
On 6/29/2016 9:22 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:
There is also a Keysight 53230A with a slightly better (20 ps vs 25 ps)
single shot resolution. It looks more modern, but I don't know how well
they compare.
Dave.
The 53230A does not offer a 10811 option, but instead
More off topic:
I saw on the history channel a story about a
British radio installation in France (IIRC)
that was taken over by the Germans, who
proceeded to masquerade as British operators,
hoping to gather intelligence. The British
became suspicious and someone got the bright
to append "HH"
Probably not necessary on this reflector, but
often needed on other reflectors I read:
If you are making a legitimate post that is
not spam, don't use "teaser" subject lines,
etc. that mimic spam messages. Don't
post tinyURL links either, no telling
what those are. Shouldn't have to say
this,
On 6/11/2016 6:33 AM, Attila Kinali wrote:
Hoi Rick,
use a small uC board to interface with the PC. Saving the samples in
a wav file and using one of the many FFT tools shouldn't be a problem.
This gets even better. The free Pscope software comes with
FFT. I don't even need to fool with
On 6/11/2016 6:33 AM, Attila Kinali wrote:
Hoi Rick,
I don't know which phase noise measurement sturcture you are going
to use exactly or what your goals are, so I am guessing here a little bit...
I'm interested in at least 10-100 Hz offset, 1-1000 Hz offset
would be a bonus. Thus audio is
I remember cobbling together a home brew phase noise system 35
years ago using an HP3582 FFT analyzer. It had several available
windows: IIRC: flat top, rectangular, and hamming (hanning? I can
never get them straight). Basically, if you are looking for spurs, you
need to use the flat top.
I remember when we first got a prototype 10816
Mini-Rubidium standard working. We put in on
of those old paper strip chart recorders (this
was circa 1981). We were pretty cocky about
how it went straight down the page. You
couldn't do that with quartz. When we
came back the next day, you
On 6/8/2016 9:59 AM, Charles Steinmetz wrote:
You can achieve substantially lower jitter (phase noise) with a
regenerative divider, which also allows you to divide by 3/2 for a 10MHz
output. I've built several like that, and they work extremely well.
When I was at Agilent, they developed
On 5/31/2016 3:30 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:
Please excuse my ignorance, but how would one lock a 116 MHz 5th overtone
crystal oscillator to 10 MHz with no difficulty? Do you have a circuit you
share that would give low phase noise, and if so how low?
If you
I'm really surprised at all the complicated solutions
posted here. A 116 MHz crystal oscillator is not
rocket science. Croven Crystals will sell you a
custom resonator. The whole circuit is like a dozen
components.
Rick N6RK
___
time-nuts mailing
On 5/30/2016 4:06 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:
I was thinking about designing a 2 m (144-146 MHz) ->HF (28-30 MHz)
transverter, using a 116 MHz local oscillator feeding a level 30 mixer.
116 + 28 = 144
116 + 30 = 146
I'm wondering what's the best way to generate 116 MHz
Does anyone have any experience with this system?
Rick Karlquist
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Len Cutler proudly displayed an HP106 in his
office. It was one of the examples of Len's
philosophy of making the best possible design
rather than a "good enough" design. I never
heard what happened to it when he passed
away.
Rick
___
time-nuts
I'll be there all day Wednesday. I'll try to
meet you guys at 5 PM.
Rick Karlquist N6RK
On 5/22/2016 9:06 AM, Bernd Neubig wrote:
Agreed!
See you on Wed at 5.00pm at Booth 714 (Dynamic Engineers)
-Ursprüngliche Nachricht-
Von: time-nuts [mailto:time-nuts-boun...@febo.com] Im Auftrag
On 5/21/2016 11:25 AM, Attila Kinali wrote:
I'd give a lot to read the design documents of the 5071.
There must be a lot of knowhow and techniques in them.
Read papers by me and my colleagues at the 1992 Frequency
Control Symposium. There is nothing else in the public
domain.
What
On 5/21/2016 8:37 AM, ws at Yahoo via time-nuts wrote:
How would one go about to prove beyond a doubt in that experiment that 18ns
of phase shift in 24 hrs was truly caused by freq offset due to Relativity
and not by some other combination of environment and handling issues that
effected phase
HP/Agilent/Keysight laser interferometers
measure at the kind of rates you are talking
about and (last time I heard) could divide
an interference fringe down to 1/512 of a
wavelength. As you say, they definitely use
an ASIC with a ring oscillator. Perhaps
there is some way you could repurpose
On 4/23/2016 6:27 PM, Bob Camp wrote:
Hi
The 5334 spanned the era when the 10811 was being introduced. I’ve never been
able to sort out exactly *what* was in each version as shipped and what got
swapped
around by various techs over the years. I always *thought* the 10811 was
standard for the
On 4/13/2016 12:28 PM, Pete Lancashire wrote:
that even with a $10K price they sold, either Bill or Dave were advised to
cancel the project due to its growing cost and complexity but
decided to keep it going in that HP would learn about designing with logic.
I have an internal HP logic design
This product came out in the early '70's when I was working
for Boeing. The company bought several and they were
very popular. This was an amazing advance for the time,
to be able to measure short term stability so easily.
This was before HPIB, so you couldn't easily connect a
counter to a
On 4/12/2016 3:16 PM, Michael Wouters wrote:
evaluations (for accuracy) to finally get its best estimate of what one
second is. Each participating lab is then told what the difference between
its clock and UTC is.
USNO participates in UTC and keeps within tens of ns of UTC.
It might also
On 3/30/2016 8:18 AM, Tim Shoppa wrote:
There's a real beauty to many of the NIST designs - using topology and
jellybean parts to achieve the performance, rather than selected devices.
Tim N3QE
If only this were true. Authors at NIST have consistently told me
that the conditions of
Nothing has come to my attention in the last
35 years that is superior for buffer amplifiers
to the simple cascade of grounded base transistors
as described by numerous NBS/NIST papers.
The chain usually starts with a common emitter
(with emitter degeneration resistor), which
is an even older NBS
exotic.
Bob
On Mar 24, 2016, at 12:25 AM, Richard (Rick) Karlquist <rich...@karlquist.com>
wrote:
On 3/23/2016 3:56 PM, Bob Camp wrote:
and standard covered. If you want to do spurs, spend $40 and build a phase
noise test set that will drive the sound card on your PC.
Lots of c
On 3/23/2016 3:56 PM, Bob Camp wrote:
and standard covered. If you want to do spurs, spend $40 and build a phase
noise test set that will drive the sound card on your PC.
Lots of choices ….
Bob
Any documentation on this $40 phase noise test set?
Rick N6RK
On 3/13/2016 10:39 AM, Nick Sayer via time-nuts wrote:
Our WWVB clock switched way, way early last night - like 20:00 or so (PST).
but that clock hasn’t exhibited any whacky behavior as long as I can remember.
Same here with my clock.
Rick
___
Does anyone have a documented way to do this?
I asked the original designers of the 10811
(Burgoon and Wilson) about this a few years
ago and they couldn't remember what the circuit
mods were. I got hold of their lab notebooks
and it wasn't in there either. What I remember
them telling me 35
On 3/8/2016 10:11 AM, jimlux wrote:
Or is subject to export controls because of the specific application.
While you could probably "generalize" it to get out from under export
controls, that's a lot of work. Ulrich and Rick raise interesting points:
The people who really know about this
On 3/8/2016 3:18 AM, ka2...@aol.com wrote:
Good Morning,
technically you are correct, most buy what they find and live with a
compromise. But companies like mine, R, test equipment , need superior
performance and many parts which we need, we have made by foundries.
Numerically controlled
I know for me, I mainly use the "synthesizer on a chip" IC's
from Analog Devices/Hittite and National. Their data sheets
and ap notes serve as the "textbook". I'm not sure there
will be much call going forward for a book on fundamentals
that explains how to design synthesizers from first
On 3/1/2016 4:13 AM, Bob Camp wrote:
Is it worth getting it super close? Probably not without a temperature test
setup.
Bob
Right.
It is entirely possible that if you did a temperature test
in an environmental chamber, you would find that you
could get a better tempco by adjusting oven
On 2/28/2016 7:01 AM, Bob Camp wrote:
Hi
It’s not an electrical issue as much as a heat issue ….
Before you start, consider that you will be doing something like:
Move trimmer 1 turn CW
Wait 10 minutes
read frequency
Move trimmer 1 turn CW
wait / read
Move trimer 1/2 turn CCW
wait / read
If you replaced the thermistor with an exact replacement,
then you shouldn't need to change the pot. If you didn't
replace the thermistor with an exact replacement, then you
should carefully measure the resistance of the pot as you
found it. (I hope you did not fool with it already.) From
the
On 2/14/2016 11:20 AM, William H. Fite wrote:
They don't wonder; they know very well. But they're stuck. Consider
oscilloscopes. Why pay for a Keysight or Tectronix or LeCroy or, God
forbid, a Rohde & Schwarz when, for the vast majority of applications, a
Rigol will give you everything you
A few clarifications:
Before 1999, HP had a Medical Division that made
equipment you saw in hospitals and a Scientific
Instrument Division that made chemical analysis
equipment used in medical laboratories (and also
other laboratories). IIRC, both began as
acquisitions. The Agilent spin off in
On 2/12/2016 12:14 PM, Joseph Gray wrote:
I sent my HP 3457A in for cal. I should be getting it back next week.
I won't mention where I sent it, but it wasn't Keysight (I don't like
that name). I recently changed the SRAM battery and purposely did not
I left Agilent just before the split, but
Vintage HP equipment often had HP made power transformers,
which ran all the time whether the equipment was on or off.
The core loss while idling could be fairly high. The
core was usually just below saturation, so that if the
power switch was on 100V, it would really get hot.
There was some
On 1/22/2016 2:14 PM, Mathew Breton wrote:
I was gifted an HP 5370B with the usual problem: front-end problems, probably
due to overstress. It is currently up and running again with a set of 5345A
series A3/A4 boards as I wasn't able to get a cheap pair of 5088-706x hybrid
ICs.
This sounds
On 1/26/2016 11:52 AM, walter shawlee 2 wrote:
I have been working on a compact portable 10Mhz bench standard
using both an FE FE5680A Rb oscillator and an Oscilloquartz ovenized
It is important for a 10 MHz source to launch a pure sine wave
and also to have an accurate 50 ohm impedance at
It is interesting that the HP8662A multiplies 10 MHz to 640 MHz,
in steps of 2X. But there is a crystal filter at 80 MHz to
clean up the wideband noise of the 10811. In the 11729, they
filter the 640 MHz from the 8662 with a SAW filter, again to
eliminate multiplied up wideband noise. It's
On 1/14/2016 12:35 PM, Nathan Johnson wrote:
What does the group think of the HP 8660? Just scored a broken one too cheap to
pass up. I know it's not gonna be the last signal generator I buy, but for under
$100 shipped it should be an interesting project.
Nathan KK4REY
When I worked at HP,
for the multistage
zero crossing detectors discussed on this forum many
times.
Rick
On 1/10/2016 2:24 AM, Attila Kinali wrote:
Hi Rick,
On Sat, 9 Jan 2016 14:45:43 -0800
"Richard (Rick) Karlquist" <rich...@karlquist.com> wrote:
This circuit is very similar to one that was champione
Phase frequency detectors (starting with the legendary MC4044)
being made out of flip flops, had metastability and/or race
conditions. Motorola showed a block diagram made of gates,
as if it were combinatorial logic, but because of the feedback,
it is actually a state machine, as described in
On 1/9/2016 12:44 PM, Attila Kinali wrote:
The purpose of the input circuit is to convert the RF input signal
into a low-jitter square wave that can drive the PIC clock input.
The circuit is closely based on the one published by Wenzel at
http://www.wenzel.com/documents/waveform.html, with
On 1/7/2016 3:11 PM, Bob Camp wrote:
Hi
If your intention is to run a mixer with saturated inputs …. just run
an X-OR gate. It will handle the high level signals much better than
an over-driven analog part.
Bob
If you look at the schematic of an XOR gate IC and compare it
to the
On 1/7/2016 4:35 AM, Attila Kinali wrote:
How about using the Gilbert Cell as "digital" mixer,
ie driving the currents hard from one branch to the other
and replacing the current sources by resistors?
How much would that improve the noise? Would it be still much
worse than the diode mixer?
On 1/5/2016 12:07 PM, Bruce Griffiths wrote:
The noise of such Gilbert cell based analog multipliers far exceeds that of the
traditional mixer.
Bruce
Read Gilbert's paper or Gray and Meyers analog IC textbook and
you will see that the whole theory of operation of these
depends on keeping
On 12/13/2015 7:15 PM, Tom McDermott wrote:
It brings up a question: Is it possible to estimate the phase noise of that
internal crystal from the ADEV measurements? There are a bunch of
papers that go the other way: from Phase Noise to Adev. Searching
brings up only one paper that goes
It's been 20 years since I presented that paper in San Francisco
at FCS and I had just about forgotten about it. It is
flattering to realize that people are still reading it now.
It might be useful for the discussion here if I explained why I
wrote the paper. We had recently completed the
On 12/9/2015 2:26 AM, Attila Kinali wrote:
Moin,
I just tried to figure out how phase microsteppers are usually build,
but, beside the time-nuts discussion from 10 years ago and US patents
US4358741 and US4417352 my search turned out empty. I am pretty sure
that I used the wrong search terms
On 11/12/2015 1:01 PM, William Schrempp wrote:
has failed. I hear old machinists complaining about new machinists who can't
drill a hole if the drill-press isn't computer-controlled. And in my work,
nurse education, I see students who can't be bothered to learn how to take a
manual
On 11/12/2015 12:04 AM, Hal Murray wrote:
I think it was HP that measured the signal in the Silicon Valley area. NBS
published and distributed the offset.
Does anybody remember that booklet? Did I get the story reasonably accurate?
When I was hired by HP in 1979, my new boss (who was
Driscoll wrote a lot about oscillators over the years.
I couldn't find anything specific to discontinuous operation.
Do you have a titel of a paper related to this?
What Driscoll was talking about was self limiting in a
transistor. That is discontinuous operation, although
Driscoll doesn't
On 10/27/2015 10:11 PM, Bruce Griffiths wrote:
The answer to this conundrum is surely that the equation for PN doesn't apply
directly in this case
for offset frequencies outside the crystal bandwidth.
The Crystal actually bandpass filters the signal and PN noise generated by
oscillator.
For
Do you have a specific URL for "hacking oscillators"? I can't
find it on Rubiola's web site.
Rick
On 10/28/2015 1:32 PM, Gerhard Hoffmann wrote:
Am 28.10.2015 um 19:22 schrieb KA2WEU--- via time-nuts:
This oscillator seems to have been more a frequency standard then a noise
standard. Today's
On 10/28/2015 10:38 AM, Hal Murray wrote:
rich...@karlquist.com said:
The 2N5179 in the 10811 is selected for minimum beta and Ft at 20 mA, which
is the start up condition due to the ALC being at full gain. It has a
special HP part number, so you wouldn't know this just looking at the parts
On 10/26/2015 9:15 PM, Bruce Griffiths wrote:
The 10811A ocxo uses an oscillator of this type albeit with a lower crystal
current, an overtone crystal. However the output stages spoil the PN
floor..Cascaded transformer coupled CB stages are somewhat quieter.
Bruce
That's right. Burgoon
The oscillator transistor and buffer amplifier are basically
the same as the HP 10811, except for the absence of mode
suppressors. The difference here is that the oscillator
self limits in the oscillator transistor, whereas the 10811
has ALC. The discontinuous operation of the transistor,
as
The thermal fuse plugs into pin sockets. It cannot be
soldered for the obvious reason that the solder
would melt it...at 109 degrees as it is marked.
My suggestion would be to jumper it out of the circuit.
Rick Karlquist N6RK
On 10/19/2015 12:31 PM, Dimitri.p wrote:
How common is it to find
There used to be a color at HP called mint gray FWIW.
Rick
On 8/13/2015 1:58 AM, tim...@timeok.it wrote:
I am interested in knowing the exact name or number of the colors used by HP
for the HP5065A front panels.
___
time-nuts mailing list --
OTOH, the other cure for high base spreading resistance is
to simply parallel multiple devices. This avoids the bad
side effects you mention. The other key noise parameter
in a BJT is RF current gain, and this cannot be cured
by any circuit design tricks.
Rick Karlquist N6RK
On 7/23/2015 8:29
On 7/24/2015 11:58 AM, Brooke Clarke wrote:
Hi Charles:
Does hFE (DC) have much relevance to this? Would hfe (AC) be the
important one?
Only insofar as DC current gain is an upper bound on AC current gain.
If your operating frequency is less than f-sub-t divided by beta,
then DC current
On 7/20/2015 8:12 AM, Charles Steinmetz wrote:
Rick wrote:
Base spreading resistance can be overcome
by using a sufficiently high source impedance
This sounds like the all-too-common noise figure fallacy (increasing
input impedance to get a lower NF). All this does is raise the source
On 7/18/2015 2:16 AM, Attila Kinali wrote:
I always wonder how you figure out whether a transistor is low noise
or not. What part of the datasheet hints at which transistors have low
noise and which have not? Even if it's just try and measure, how
do you find good candidates to measure?
On 6/20/2015 6:25 AM, Jim Lux wrote:
This makes a good case for the 30dB/decade very close in
Somewhat had asked about how close in the 30 dB/decade is
good for. There is a reference about this issue. The
book Edson: Vacuum Tube Oscillators has what I believe is
the first published
On 6/20/2015 1:16 PM, Alex Pummer wrote:
Actually a YIG, even standalone has very good phase noise performance,
as long as the tuning current is quite, once upon the time HP made some
cheaper version of the 856x-es spectrum analyzers [ perhaps that was the
95xx ] they had the first LO just
On 6/17/2015 11:36 PM, Attila Kinali wrote:
Do you have any recomendation, where an ordinary engineer could
read up on this topic?
Attila Kinali
There is always Floyd Gardners, Phase Lock Techniques.
However, a better tutorial would be the one written by
HP's Dieter
On 6/18/2015 1:20 PM, Magnus Danielson wrote:
The trick is to convert the 2nd degree loop to a 3rd degree loop, which
then allows for a 12 dB/oct slope, to counteract the 9 dB/oct slope.
No this is not correct. A very conventional Type 2 loop, where the
loop filter consists of an integrator
On 6/17/2015 8:22 AM, Jim Lux wrote:
I'm looking for some representative data for inexpensive microwave VCOs
(in the 2.5-6 GHz range, in general). Not in a locked loop situation,
If the phase noise data you have goes to a low enough frequency to
get below the 1/f corner (which is the case
On 6/17/2015 8:22 AM, Jim Lux wrote:
I'm looking for some representative data for inexpensive microwave VCOs
(in the 2.5-6 GHz range, in general). Not in a locked loop situation,
If you are working up to 2.5 GHz, you can get a low power
chip for $2 from Analog Devices that has a VCO and
That's interesting. I worked for the HP Santa Clara Division
from 1979 until just before it was closed in 1998. I
forget who invented MDA at SCD, but it was hyped like
it was some new concept and I never heard anything about
the HP9540.
Many times someone would come to me and ask me about
some
The counter only had to run at ~50 MHz, on account of our
mode locked laser ran at that frequency. I don't remember
what the CPLD was rated at.
Rick
On 6/5/2015 8:19 PM, Hal Murray wrote:
rich...@karlquist.com said:
I used a CPLD in a 900 GHz (that's right 900 GHz) optical sampling scope
I used a CPLD in a 900 GHz (that's right 900 GHz) optical
sampling scope timebase. It was great because you just
write a 17 bit counter in VHDL and there it is. You
don't have to know anything about building digital
hardware any more (40 years of experience wasted).
Nobody cares about look
Can someone explain to me how this is going to work in
light of the fact that each clock is in a different
gravitational field? Or is accuracy not the measurement,
but rather stability? No, that can't be because any
lab that wants to measure stability merely needs to build
two or three copies
The only gates that seem to do very well are high speed (as in 74AC or faster)
silicon CMOS. You need to run them with a fairly clean supply and feed them
with a p-p input that matches the supply voltage. Other than that, not a lot
of magic. Are they ideal - surely not. Will they hit 2x10^-13
On 5/20/2015 11:22 AM, Magnus Danielson wrote:
The older HP counter manuals explained it very nicely too, as they
illustrated the slew-rate amplitude noise to time-noise conversion.
What do amazes me is the fact that I've yet to see a counter input
channel which takes care to square up the
On 5/8/2015 2:19 PM, Bob Camp wrote:
On May 7, 2015, at 11:09 AM, Attila Kinali att...@kinali.ch wrote:
On Wed, 06 May 2015 18:09:03 -0700
Richard (Rick) Karlquist rich...@karlquist.com wrote:
A standard input on a frequency counter is not a very demanding thing in the
hierarchy
On 5/6/2015 3:24 PM, Bob Camp wrote:
Hi
A standard input on a frequency counter is not a very demanding thing in the
hierarchy of
TimeNut signals. You can drive any of them with some pretty simple logic gate
based
circuits. No need to spend a lot of money.
Bob
Logic gate, yes.
On 4/26/2015 3:51 AM, Bryan _ wrote:
All:
P
I was looking at the project from David partridges web site
http://www.perdrix.co.uk/FrequencyDivider/index.html
-=Bryan=-
___
This is a comparator based
On 5/4/2015 5:25 PM, J. L. Trantham wrote:
Charles,
Thanks for the help. I need to learn how to add text to .PDF documents.
Go to:
www.tracker-software.com
and download (for free):
PDF-Xchange viewer.
Rick Karlquist N6RK
___
time-nuts
On 4/23/2015 12:20 AM, VK2DAP wrote:
Dear time-nuts,
I have a question about the HP5328A and HP5328B universal counters.
1) Generally speaking, would it be correct to say that when a product model
number changes from A to B, that represents an improvement or major update to a
product?
2) I
101 - 200 of 409 matches