Thank you all for the answer to my query.

cheers, Graham ve3gtc

On 2014-11-21 21:17, Bob Camp wrote:

Hi

We’ve been around this one before. The KS boxes are powered by a switcher brick 
inside the box. There’s not a lot of reason go super crazy on the DC input. 
It’s fully isolated from the case ground. Even leakage / stray grounds should 
not be an issue in this situation. I would not run it off of raw un-filtered 
rectifier output. Anything that’s at least got a filter capacitor in it should 
do fine.

Bob


On Nov 21, 2014, at 3:19 PM, Neil Schroeder gign...@gmail.com wrote:

I drive my prs10 with a parallel pair of ldos balanced by an op amp. I'll
draw it out and post it if you like.

On Friday, November 21, 2014, Azelio Boriani azelio.bori...@gmail.com
wrote:


At most you need 2.4A (when the two OCXOs are heating up): a 2.5A
supply is needed.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-22 Thread Brian Lloyd
On Fri, Nov 21, 2014 at 8:17 PM, Bob Camp kb...@n1k.org wrote:

 Hi

 We’ve been around this one before. The KS boxes are powered by a switcher
 brick inside the box. There’s not a lot of reason go super crazy on the DC
 input. It’s fully isolated from the case ground. Even leakage / stray
 grounds should not be an issue in this situation. I would not run it off of
 raw un-filtered rectifier output. Anything that’s at least got a filter
 capacitor in it should do fine.


Or batteries. Two 12v batteries in series on float charge not only powers
it nicely, it keeps everything going should the mains fail. You won't have
to wait for it to reacquire and stabilize after a mains power failure.

-- 
Brian Lloyd
Lloyd Aviation
706 Flightline Drive
Spring Branch, TX 78070
br...@lloyd.aero
+1.210.802-8FLY (1.210.802-8359)
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread planophore
Good day all,

I have been following the discussions as best as I could. I recell seeing
the question asked or at least theorized as to what is the power
requirements of these units - either individually or together.

I have reviewed all messages or at least all that I can find in these
threads but I couldn't find a definitive answer unless of course the
answer is in one of the responses hiding where I haven't looked.

I purchased a set of these modules and they are currently in transit but
I don't have a 24V DC supply at the moment.

So, in order to try and get ahead before the arrival of these modules I
would like to at least plan out if not get a suitable power supply
ready. I don't want to over specify by too much and worse yet, under
specify.

Has anyone actually made any power used measurements of these modules? 2A
average, 4A peak? Somewhere inbetween or more?

cheers, Graham ve3gtc
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread Azelio Boriani
At most you need 2.4A (when the two OCXOs are heating up): a 2.5A
supply is needed.

On Fri, Nov 21, 2014 at 6:25 PM,  planoph...@aei.ca wrote:
 Good day all,

 I have been following the discussions as best as I could. I recell seeing
 the question asked or at least theorized as to what is the power
 requirements of these units - either individually or together.

 I have reviewed all messages or at least all that I can find in these
 threads but I couldn't find a definitive answer unless of course the
 answer is in one of the responses hiding where I haven't looked.

 I purchased a set of these modules and they are currently in transit but
 I don't have a 24V DC supply at the moment.

 So, in order to try and get ahead before the arrival of these modules I
 would like to at least plan out if not get a suitable power supply
 ready. I don't want to over specify by too much and worse yet, under
 specify.

 Has anyone actually made any power used measurements of these modules? 2A
 average, 4A peak? Somewhere inbetween or more?

 cheers, Graham ve3gtc
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread Anthony Roby
I have been using a Universal 24V 5A DC Switching Power Supply from eBay.  I 
have not seen my units draw more than 2A each - steady state is about 1.1A.

Anthony

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of 
planoph...@aei.ca
Sent: Friday, November 21, 2014 11:25 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Lucent KS-24361 24V DC power requirements

Good day all,

I have been following the discussions as best as I could. I recell seeing the 
question asked or at least theorized as to what is the power requirements of 
these units - either individually or together.

I have reviewed all messages or at least all that I can find in these threads 
but I couldn't find a definitive answer unless of course the answer is in one 
of the responses hiding where I haven't looked.

I purchased a set of these modules and they are currently in transit but I 
don't have a 24V DC supply at the moment.

So, in order to try and get ahead before the arrival of these modules I would 
like to at least plan out if not get a suitable power supply ready. I don't 
want to over specify by too much and worse yet, under specify.

Has anyone actually made any power used measurements of these modules? 2A 
average, 4A peak? Somewhere inbetween or more?

cheers, Graham ve3gtc
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread Bert Kehren via time-nuts
I think Bob mentioned it before Laptop power supplies make good power  
sources. Older units like in my case an old Dell 20V 3.5 A is perfect. They  
still have plenty of input and output filtering and build for lomg term  
operation. Never throw away a laptop supply. Late models are lower power and  
less 
filtering. 
Bert Kehren
 
 
In a message dated 11/21/2014 4:42:29 P.M. Eastern Standard Time,  
ar...@antamy.com writes:

I have  been using a Universal 24V 5A DC Switching Power Supply from eBay.  
I  have not seen my units draw more than 2A each - steady state is about  
1.1A.

Anthony

-Original Message-
From: time-nuts  [mailto:time-nuts-boun...@febo.com] On Behalf Of 
planoph...@aei.ca
Sent:  Friday, November 21, 2014 11:25 AM
To: Discussion of precise time and  frequency measurement
Subject: [time-nuts] Lucent KS-24361 24V DC power  requirements

Good day all,

I have been following the discussions  as best as I could. I recell seeing 
the question asked or at least theorized  as to what is the power 
requirements of these units - either individually or  together.

I have reviewed all messages or at least all that I can find  in these 
threads but I couldn't find a definitive answer unless of course the  answer is 
in one of the responses hiding where I haven't looked.

I  purchased a set of these modules and they are currently in transit but I 
don't  have a 24V DC supply at the moment.

So, in order to try and get ahead  before the arrival of these modules I 
would like to at least plan out if not  get a suitable power supply ready. I 
don't want to over specify by too much  and worse yet, under specify.

Has anyone actually made any power used  measurements of these modules? 2A 
average, 4A peak? Somewhere inbetween or  more?

cheers, Graham  ve3gtc
___
time-nuts mailing  list -- time-nuts@febo.com To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions  there.
___
time-nuts mailing  list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread Bob Camp
Hi

I have run mine from anything from 22V to 30V and seen no problems. I had 
trouble on a pair with a 2A 24V supply. The pairs seem to settle down to about 
1A at 30V after warmup in my lab. 

The supplies are rated to 18V input. I would not trust them at that low a 
voltage. There are some coils running around in the box and voltages into the 
supply will be lower than at the input pins. 

Even if you buy the brand new supplies from TRC (or your local MeanWell 
distributor) a full up supply for a pair is  $30. The auction sites are indeed 
a bit cheaper on some auctions.

Bob


 On Nov 21, 2014, at 12:50 PM, Anthony Roby ar...@antamy.com wrote:
 
 I have been using a Universal 24V 5A DC Switching Power Supply from eBay.  I 
 have not seen my units draw more than 2A each - steady state is about 1.1A.
 
 Anthony
 
 -Original Message-
 From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of 
 planoph...@aei.ca
 Sent: Friday, November 21, 2014 11:25 AM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] Lucent KS-24361 24V DC power requirements
 
 Good day all,
 
 I have been following the discussions as best as I could. I recell seeing the 
 question asked or at least theorized as to what is the power requirements of 
 these units - either individually or together.
 
 I have reviewed all messages or at least all that I can find in these threads 
 but I couldn't find a definitive answer unless of course the answer is in one 
 of the responses hiding where I haven't looked.
 
 I purchased a set of these modules and they are currently in transit but I 
 don't have a 24V DC supply at the moment.
 
 So, in order to try and get ahead before the arrival of these modules I would 
 like to at least plan out if not get a suitable power supply ready. I don't 
 want to over specify by too much and worse yet, under specify.
 
 Has anyone actually made any power used measurements of these modules? 2A 
 average, 4A peak? Somewhere inbetween or more?
 
 cheers, Graham ve3gtc
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread Graham
Just to clarify, each unit (no more than 2 and steady about 1.1A) or 
both units together?


cheers, Graham ve3gtc

On 2014-11-21 12:50, Anthony Roby wrote:

I have been using a Universal 24V 5A DC Switching Power Supply from eBay.  I 
have not seen my units draw more than 2A each - steady state is about 1.1A.

Anthony

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of 
planoph...@aei.ca
Sent: Friday, November 21, 2014 11:25 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Lucent KS-24361 24V DC power requirements

Good day all,

I have been following the discussions as best as I could. I recell seeing the 
question asked or at least theorized as to what is the power requirements of 
these units - either individually or together.

I have reviewed all messages or at least all that I can find in these threads 
but I couldn't find a definitive answer unless of course the answer is in one 
of the responses hiding where I haven't looked.

I purchased a set of these modules and they are currently in transit but I 
don't have a 24V DC supply at the moment.

So, in order to try and get ahead before the arrival of these modules I would 
like to at least plan out if not get a suitable power supply ready. I don't 
want to over specify by too much and worse yet, under specify.

Has anyone actually made any power used measurements of these modules? 2A 
average, 4A peak? Somewhere inbetween or more?

cheers, Graham ve3gtc
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread Bob Camp
Hi

A bit over 2A for the pair at turn on. About 1A for the pair after warmup.

Bob

 On Nov 21, 2014, at 6:23 PM, Graham planoph...@aei.ca wrote:
 
 Just to clarify, each unit (no more than 2 and steady about 1.1A) or both 
 units together?
 
 cheers, Graham ve3gtc
 
 On 2014-11-21 12:50, Anthony Roby wrote:
 I have been using a Universal 24V 5A DC Switching Power Supply from eBay.  I 
 have not seen my units draw more than 2A each - steady state is about 1.1A.
 
 Anthony
 
 -Original Message-
 From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of 
 planoph...@aei.ca
 Sent: Friday, November 21, 2014 11:25 AM
 To: Discussion of precise time and frequency measurement
 Subject: [time-nuts] Lucent KS-24361 24V DC power requirements
 
 Good day all,
 
 I have been following the discussions as best as I could. I recell seeing 
 the question asked or at least theorized as to what is the power 
 requirements of these units - either individually or together.
 
 I have reviewed all messages or at least all that I can find in these 
 threads but I couldn't find a definitive answer unless of course the answer 
 is in one of the responses hiding where I haven't looked.
 
 I purchased a set of these modules and they are currently in transit but I 
 don't have a 24V DC supply at the moment.
 
 So, in order to try and get ahead before the arrival of these modules I 
 would like to at least plan out if not get a suitable power supply ready. I 
 don't want to over specify by too much and worse yet, under specify.
 
 Has anyone actually made any power used measurements of these modules? 2A 
 average, 4A peak? Somewhere inbetween or more?
 
 cheers, Graham ve3gtc
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread Neil Schroeder
I drive my prs10 with a parallel pair of ldos balanced by an op amp. I'll
draw it out and post it if you like.

On Friday, November 21, 2014, Azelio Boriani azelio.bori...@gmail.com
wrote:

 At most you need 2.4A (when the two OCXOs are heating up): a 2.5A
 supply is needed.

 On Fri, Nov 21, 2014 at 6:25 PM,  planoph...@aei.ca javascript:;
 wrote:
  Good day all,
 
  I have been following the discussions as best as I could. I recell seeing
  the question asked or at least theorized as to what is the power
  requirements of these units - either individually or together.
 
  I have reviewed all messages or at least all that I can find in these
  threads but I couldn't find a definitive answer unless of course the
  answer is in one of the responses hiding where I haven't looked.
 
  I purchased a set of these modules and they are currently in transit but
  I don't have a 24V DC supply at the moment.
 
  So, in order to try and get ahead before the arrival of these modules I
  would like to at least plan out if not get a suitable power supply
  ready. I don't want to over specify by too much and worse yet, under
  specify.
 
  Has anyone actually made any power used measurements of these modules? 2A
  average, 4A peak? Somewhere inbetween or more?
 
  cheers, Graham ve3gtc
  ___
  time-nuts mailing list -- time-nuts@febo.com javascript:;
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com javascript:;
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Lucent KS-24361 24V DC power requirements

2014-11-21 Thread Bob Camp
Hi

We’ve been around this one before. The KS boxes are powered by a switcher brick 
inside the box. There’s not a lot of reason go super crazy on the DC input. 
It’s fully isolated from the case ground. Even leakage / stray grounds should 
not be an issue in this situation. I would not run it off of raw un-filtered 
rectifier output. Anything that’s at least got a filter capacitor in it should 
do fine.

Bob

 On Nov 21, 2014, at 3:19 PM, Neil Schroeder gign...@gmail.com wrote:
 
 I drive my prs10 with a parallel pair of ldos balanced by an op amp. I'll
 draw it out and post it if you like.
 
 On Friday, November 21, 2014, Azelio Boriani azelio.bori...@gmail.com
 wrote:
 
 At most you need 2.4A (when the two OCXOs are heating up): a 2.5A
 supply is needed.
 
 On Fri, Nov 21, 2014 at 6:25 PM,  planoph...@aei.ca javascript:;
 wrote:
 Good day all,
 
 I have been following the discussions as best as I could. I recell seeing
 the question asked or at least theorized as to what is the power
 requirements of these units - either individually or together.
 
 I have reviewed all messages or at least all that I can find in these
 threads but I couldn't find a definitive answer unless of course the
 answer is in one of the responses hiding where I haven't looked.
 
 I purchased a set of these modules and they are currently in transit but
 I don't have a 24V DC supply at the moment.
 
 So, in order to try and get ahead before the arrival of these modules I
 would like to at least plan out if not get a suitable power supply
 ready. I don't want to over specify by too much and worse yet, under
 specify.
 
 Has anyone actually made any power used measurements of these modules? 2A
 average, 4A peak? Somewhere inbetween or more?
 
 cheers, Graham ve3gtc
 ___
 time-nuts mailing list -- time-nuts@febo.com javascript:;
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com javascript:;
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.