Hello,
I’m using the NavSyncview software with the FTS125 GPS module.
Is it possible to set the time readout on the NavSync software to a local time?
I actually need UTC plus 8 hours.
It’s at UTC and I cannot see any way of changing it.
Best Regards
Garry
Hello,
Since the questions has been asked I will reply also apologies for the “advert”
We make ultra low noise distribution amplifiers.
1 Hz noise is –135 dBc and we can achieve –172 dBc noise floor if we
carefully set all the gains.
www.ptsyst.com/DA1-100-10-B.pdf
But these are expensive
Hi,
Apologies if this has been asked 1000 times before.
Is there any good software for reading the NMEA or Motorola binary code out of
a M12M.
I have winoncore, but wondered if there was anything a bit more modern.
Regards
Martyn
___
time-nuts
Hello,
Has anyone tried the NV08C-MCM GPS receiver chip?
I’m interested in a chip that will work with all the different satellite
systems.
Regards
Martyn
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
Hello,
Here's a timepod plot of my own amplifier that my company sells
www.ptsyst.com/phasenoise.pdf
Note this is a distribution amplifier with about 7 dB of gain.
We can get the floor to about -174 dBc if we reduce the gain to unity.
Best Regards
Martyn
Hi All,
I read Garry's email about using the 74AC04 as a distribution amplifier.
Did I read it right, does Gary say you will get -178 dBc/Hz at 10 kHz
offset???
Being one of only 3 or 4 companies in the world that manufactures
distribution amplifiers that break the -170 dBc/Hz region at 10
Garry,
Are you also saying your 100 MHz OXCO is making -178 dBc/Hz?
Unless your 100 MHz OXCO is some special $10k type, there's no way your OXCO
has a -178 dBc/Hz floor noise?
Regards
Martyn
___
time-nuts mailing list -- time-nuts@febo.com
To
Hi,
I stand corrected. Indeed Garry does make ULN 100 MHz OXCO's.
Garry I've already sent you an email directly to see if you can help me with
another part I need.
Best Regards
Martyn
This Email is from:
Martyn Smith
Precision Test Systems LTD
Gerard,
I just read your post about your DIY counter and asking for help.
I'm not sure when you actually wrote it.
I'm based in Chelmsford Essex and can give you some help.
I have multiple frequency standards as well as 4 x SR620 timer counters.
Let me know if I can be of help.
Regards
Hello,
Can anyone recommend software to calulate the Allan varaince from a set of
data?
I have taken many measurements with the SR620 using it's 500 second gate and
downloaded the results into Excel.
I can use the SR620 itself to calulate the Allan Variance. However, I want
to know the
I read with interest about the new 10 MHz distribution amplifier.
But I agree with everyone, it's the specs that are missing that are
important.
But I wanted to take one comment up about AGC.
My company makes a nice distribution amplifier and we do have AGC. The nice
thing here is that a +7
$1 to CE mark
a product. And if we add one small option, it must be CE marked again. To
the small manufacturer, this kills us and doesn't allow us to offer low
prices.
Steve
Best Regards
Martyn
This Email is from:
Martyn Smith
Precision Test Systems LTD
Tel: +44 (0) 1245 329608
Email
Sorry you are wrong.
There is no exemtion for CE marking. Nothing is excempt. All test
equipment, all electronic equipment, even toothpaste must be CE marked.
There is no exemtion for personal use.
You maybe are getting confused with RoHS where test equipment is exempt.
Martyn
Hi All,
Yes I can't spell. I've just realised my spell checker isn't working.
Not only can't I spell, I can't type!!!
I did mean exemption if my previous email.
Best Regards
Martyn
Best Regards
___
time-nuts mailing list --
Hello All,
I have a couple of HP3048A with option 301 phase noise test sets.
I run the systems on HP Brio computers with a GPIB interface, running
Windows 95.
I bought the units as working systems and have no knowledge on the 3048A
software and how it works.
I want to upgrade the system
Hello,
I'm trying to get a 82335 HPIB card to run on my computer.
The card works in an old computer running windows 95, so I know the card is
ok.
I have a Pentium 3 running windows XP home edition. This computer has an
ISA connector, so the HPIB fits ok.
But the computer just doesn't
Hello,
I have a working 3048A phase noise test set. I'm trying to get a new
computer to work with the system.
I have finally got the computer to talk to the HPIB 82335A HPIB card.
I am using windows 98.
When I run the PN.exe program, it immediately says that it can't find the
HPIB card.
Hello,
Re my email on the 3048A software.
I got the unit to work. I have to start windows 98 in dos mode.
Don't know why but I'm happy with that.
It would be nice to have windows running properly, so if anyone does know
the answer, that would be great.
Regards
Martyn
Hello,
Today many manufacturers are producing 10 MHz oscillators that offer many
advantages over BVA.
Phase noise of -116 dBc @ 1 Hz with a -175 dBc noise floor, ageing of 5 x
10E-9 per month and a G sensitivity of 5 x 10E-11 /g in all axis is
available on a new product.
Also the short term
Hello,
I mentioned about there being other low Allan Dev oscillators available on
the market.
Said asked me to supply some results (And Yes Said, I meant 1 x 10E-13)
I'm not sure if attachments work on timeouts, so I have put a couple of
graphs on my website.
These graphs show typical results
Hello,
Re my last post about low noise 10 Mhz oscillators.
I found a better PN plot of our 10 MHz oscillators.
www.ptsyst.com/10MHzPN-2.pdf
Unfortunately don't have the Allan Dev for this one.
Best Regards
Steve
___
time-nuts mailing list --
Hello,
My colleague Steve posted an Allan Dev plot of one of our 10 MHz
oscillators.
It showed a funny hump at the beginning of the measurement.
The hump is just because the units were still warming up. They hadn’t been
on long before I started the measurement.
The problem is that we
Hello,
I’m a bit new to GPS receivers. I’m reading all about sawtooth correction of
the 1 pps output signal, on the ilotus M12M or M12+.
I understand that correction can either be hardware controlled or software
controlled.
Does anyone make a low cost stand alone board to carry out this
Hello,
I have been asked a question from a friend and would like to hear what the
time-nuts have to say about the M12+ and M12M GPS receivers.
His questions were:
1) It would be great if you could outline the broad algorithm and bits of
information that go into keeping the 1 pps aligned to
Hello,
I may have asked this before, but still haven't got an answer.
The Motorola M12+ or M12M quote 20 ns accuracy to UTC.
I have two receivers running off the same aerial through a GPS splitter.
If I measure the 1 pps signals on my SR620 counter (single shot), they are
up to 100 ns out.
Hello,
I just bought a HP8561B on ebay but it has a few faults.
1. The unit powers up and basically operates ok. But after a while, it
turns off, turns back on, goes through the start sequence, and then stays on
for a few minutes and turns off again. It cycles through this.
2. Sometimes
Hello,
I am trying to achieve accuracy with two frequency standards disciplined to
GPS.
I am trying to get their 1 pps outputs close to each other as possible.
The idea being you put each unit in a different part of the world and the 1
pps outputs are sycronised.
So far I have two units
Hello,
Can anyone advise how I can measure the frequency accuracy of my 10 MHz
standard at gate times of 1 ms and 100 microseconds.
I have the Wenzle program that calulates Allan Variance from phase noise,
but don't know how reliable that is.
The 1 second Allan Variance is about 1.2 x
Hello,
I am using my GPS disciplined rubidium frequency standard with a fibre optic
transmitter and receiver. The Tx and Rx is connected between the antenna
and the frequency standard with a meter of fibre optic cable.
The Tx and Rx devices are especially designed for the GPS frequency and
Hello,
I asked a question about fibre optic, but the problem has been solved.
It had nothing to do with the fibre optic unit.
But if anyone does have any thoughts on using a fibre optic link, and any
pitfalls, I'd be intertesed.
Cheers
Martyn
Hello,
Will exisitng GPS receivers (Motorola M12M for example) be able to work with
the new Galileo system when it comes on line.
I know the systems are compatible, but not sure about older GPS receivers.
Any help appreciated
Martyn
___
Hello,
Thanks Bj?rn, Magnus and Said for the input.
That's another upgrade I'll (not) look forward to.
Martyn
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow
Hello,
I knew it would happen. A customer has just asked for a GPS system based on
Galileo.
What's the staus of Galileo? When is it expected to be running?
Does anyone know of Galileo GPS receivers or dual receivers?
Thanks
Martyn
___
Hello,
I have a question about TAI (Temps Atomique International) time.
As I understand it TAI is UTC time without the leap seconds.
Is that correct?
So is that the same as GPS time?
I need to be able to generate accurate TAI time as opposed to UTC time.
I'm a bit confused as to what the
Helo,
Re my last question.
I've read that GPS time is ahead of UTC time by 15 seconds and TAI time is
ahead of UTC by 34 seconds.
I need to generate time to a display based on TAI time.
Any ideas how I can do that with a normal M12M GPS receiver.
Martyn
Hello,
I'm trying to measure the Allan Deviation of an amplifier and need help with
the maths.
My measurement process uses the SR620 in time interval mode and I make one
measurement per second for about a day.
I then use Ulrichs excellent plotter to calculate the Allan Deviation.
The Allan
Hello,
My measurement technique is from Stanford's own application note.
The two 10 MHz are fed to Ch A and B. The measurement is a time interval
gated at 1 kHz from the SR620's ref out.
So we make 1000 measurements per second.
So the one shot 25 ps SR620 spec is reduced by square root of
Hi,
I'm looking for the sensitivity specification of the Synergy M12M GPS
receiver.
Has anyone data on this. I can't find it in any brochure or manual.
Regards
Steve
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
Hello,
Has anyone out there circuit diagrams for a HP8713C network analyzer?
The generate port is fine, but the transmission or RF IN port is faulty
below about 1.5 GHz. It's response is all over the place. Above 1.5 GHz it
seems to be fine.
I suspect the A5 receiver especially the RF input
Hello,
I need to produce a pulse per minute, derived from 10 MHz or 1 pps.
So a divide by 60 if 1 pps is used.
Does anyone make such a divider? That would save me from designing my own.
Regards
Steve
___
time-nuts mailing list --
Hello,
I'm trying to compare the stability specifications of two oscillators.
The first spec is 2 x 10E-11 Allan Variance.
The second spec is 3 x 10E-11 Allan Deviation.
As Allan deviation (ADEV), is the square root of Allan variance, I thought
the first spec is the better of the two, but a
Hi,
Thanks for the answers but I'm still confused.
The Stanford PRS10 has an Allan Variance of 2E-11 but a Symmetricomm unit
has a Allan Deviation of 3E-11.
So according to the answers you gave, the Symmetricomm unit has a stability
thousands of times better than the PRS10.
This, dispite
Hello,
The answer to my original question is that Allan Variance is not to be used
but rather Allan deviation.
However, David Allan does not mention Allan Deviation on his website, only
Allan Variance.
Also I spoke to the designer of the PRS10 and he definitely said his specs
are
Allan
Hello,
Sorry that I'm harping on about this subject but I want to make sure my
brochures are correct.
I use the SR620 to measure Allan Variance. When I look at the SR620 manual
the formulae they give for Allan Variance is actually root Allan Variance.
So root Allan Variance is the same as
Robert thanks for telling me my typo. Should be www.ptsyst.com.
I'm glad you agree. I'd like to see what everyone else says then I can tell
the customer he's wrong, with confidence!!!
Best Regards
Martyn
This Email is from:
Martyn Smith
Precision Test Systems LTD
Tel: +44 (0) 1245 329608
10 MHz oscillator.
It makes -121 dBc/Hz @ 1 Hz. Noise floor is only -162 dBc/Hz at the moment,
but we are working on reducing this a further 5 to 10 dB.
Enter your order number here..
Best Regards
Martyn
This Email is from:
Martyn Smith
Precision Test Systems LTD
Tel: +44 (0) 1245
Rick,
Your just the person I need
Any chance of a cesium before Christmas?
Seriously, I do know how complicated they are.
Although we supply hydrogen masers, everyone prefers cesium's. Also I hear
there's a large inquiry out there for tens of units. Maybe 100's.
The goods news on our
oscillators for under $1000, the
cost is not as much as what would be needed to keep an OXCO.
Best Regards
Martyn
This Email is from:
Martyn Smith
Precision Test Systems LTD
Tel: +44 (0) 1245 329608
Email: [EMAIL PROTECTED]
Web: www.ptsyst.com
NOTICE - This message contains legally privileged
at 1 Hz with a -183 dBc/Hz floor noise. Now
that's what I call low noise.
Anyone know any more about this? I'm trying to make contact with the guy
but no success yet.
Best Regards
Martyn
This Email is from:
Martyn Smith
Precision Test Systems LTD
Tel: +44 (0) 1245 329608
Email: [EMAIL PROTECTED
. This takes
some 40-50 pages of my forthcoming book on
phase noise and frequency stability in oscillators.
You can't enter an order number, you must wait for
Cambridge University Press, which is a non-profit
company.
Cheers,
Enrico
On 6 Dec 2007, at 15:12 , Martyn Smith wrote:
Hi
John,
I've tried different level spurs from -40 to -120 dBc. From -40 dBc to -100
dBc the spur, as shown on the HP3048A, is always 6 dB lower than it really
is. So if I put a -40 dBc spur in, the HP3048A says its -46 dBc.
However, when I try from -100 to -120 dBc, the spur is a bit random and
Hi All,
This comment is bound to get you all going.
Maybe I'm being stupid, but why does everyone use Allan Variance and not
plan old accuracy?
I am very familiar with David Allan's full article on Allan Variance.
However Allan Variance isn't the same as accuracy.
Accuracy is what is important
Pete,
That's my point.
Tom has plotted Allan Variance. I have plotted frequency accuracy.
I get similar Allan Variance results to Tom. But if I just measure the
accuracy, e.g peak to peak frequency variations, rubidium's out- perform
OXCO's even when both are disciplined to GPS.
I hope I'm
In replying to Didier's question.
The SR620's reference is my in-house standard. In fact a Russian high
performance rubidium that has been drifting 5E-13/month for the past five
years. It's now ten years old and I will very much miss it when it finally
dies as it is very stable.
The
David,
I'm sure I can find you an oscillator.
Give me a call on 01245329608 to discuss your requirements and budget.
Best Regards
Martyn
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
David,
Got it. OXCO will leave tomorrow via special post.
Full specs and pin-out are included.
Best Regards
Martyn
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
Hi,
Two questions for all of you today.
1st Question
I can measure 10 MHz frequency to an accuracy of about 2 parts in 10E-12 in
a one second gate time.
I use a SR620 time interval counter and make a timeAB measurement.
I make two measurements of phase (between my UUT and my reference) and
Hi All,
I have an ongoing problem and realized why don't I ask the time nuts!!
My frequency and temperature data is usually plotted in a comma separator
value (csv) file.
I use excel to import and plot the results.
Excel only lets me plot 32000 points at a time. I often have over a million
Hello,
Thanks to all for your replies and suggesting especially Brooke, Bruce and
Pete.
I am already using the Stanford technique with the 1 kHz output. That's how
I get 2 x 10E-12/sec accuracy.
Using this technique, the accuracy should be 7.9 uHz at 10 MHz (7.9E-13).
But I don't reliably
Hello,
Ulrich was asking about 10 MHz oscillators.
We make a 10 MHz OXCO that has –118 dBc @ 1 Hz with a –175 dBc noise floor.
Not cheap though.
Also Low G sensitive oscillators. 5 x 10E-11/g.
Contact me off list as I’m trying not to advertise on this forum.
Best Regards
Martyn
Hello,
I’m responding to John Miles comment about 100 MHz oscillators.
We’ve managed to design a 100 MHz OXCO that makes –140 dBc @ 100 Hz with –182
dBc @ 1 MHz.
We also have a design on the table that should improve the 100 Hz to –145 dBc,
but haven’t had the time or money to actually build
Hello,
I am using a LEA-6T GPS chip with the DS1123LE-50 delay line to correct the
jitter of the 1 pps output from the GPS chip.
The correction works very well, dropping the peak to peak jitter of the 1pps
output from about 20 ns to 3 ns.
However, I randomly get a big jump of the 1 pps
Hello,
I have both the HP3048A and Timepod Phase Noise test Sets.
I have two low noise 100 MHz PLL's. The 100 MHz output is locked to a 10
MHz reference input.
I want to measure their phase noise. Their spec is -100 @ 10 Hz with a -168
noise floor.
I connect both 100 MHz outputs to the
Hello,
I have some questions on GPS and GNSS.
Do all the civilian GPS receivers only operate on the L1 frequency?
Are there any GPS frequency standards out there that use L1 and L2 and that can
be purchased by non-military customers?
I am playing with the new Lea-M8T receiver.
How do I know
you need a medium power transistor
amplifier or opamp.
Bob
On Wednesday, March 4, 2015 6:09 AM, Martyn Smith mar...@ptsyst.com
wrote:
Hello,
A quick question.
My output driver for a simple amplifier.
I use three gates (in parallel with resistors) from a 74AC14 to give me about
0
Hello,
A quick question.
My output driver for a simple amplifier.
I use three gates (in parallel with resistors) from a 74AC14 to give me about
0-3.2V into 50 ohm.
I want to have a driver that gives me a full 0-5V (at least 0-4.5V) swing into
50 ohms.
Can anyone recommend an IC that can
Hello,
I apologize if this question has been asked before.
I have a frequency standard that uses the M12+ to discipline a rubidium
oscillator.
I set the M12+ to do a site survey and then set it to position hold mode for
best accuracy.
My question is:
If I then move the unit, for example 30
Hello,
Does anyone have an EXCEL spreadsheet that calculates the individual phase
noise of 3 oscillators when they are compared against each other, e.g A vs
B, A vs C, B vs C.
I.e the 3 corner hat technique.
I do have a Timepod and I thought Timelab could do that, have haven't found
how
frequency standards as I will always use the 3 source method of measurement
on the timepod.
John have saved me about 10 hours a week of measurements!!!
Regards
Martyn Smith
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https
. The Timepod is the worlds best bit of electronics
ever, and if John Miles was English, I would recommend him for a knighthood.
John, hope you're blushing now.
Regards
Martyn
Today's Topics:
1. measuring os latency for pps (folkert)
2. 3 corner hat for phase noise (Martyn Smith)
3. Re: SRS PRS10
Hello,
I asked for a 3 corner hat measurement for phase noise.
Managed to work it out myself.
In case anyone else wants to know:
First of all Timepod does do it but only for Allan Deviation.
But I wanted it for phase noise.
So if you make three phase noise measurements A vs B (call result
Hello,
Re my measurements with the timepod.
John Miles has now saved me a ton of work by telling me how to using three
sources in the timepod, to automatically get the absolute phase noise of a
device (without having to result to 3 corner formulae).
John, I have one more question. What about
Hello,
I use the Connor Winfield FTS125 Disciplined GPS module in a system we make.
I have three customers all complaining about the same thing.
The NMEA output suddenly stops for no reason. It needs a power cycle to
restore.
But it only occurs every few months and at completely random
Hello,
I'm just about to order a timepod at $9400 from Micro Semi.
Does any one have a used one for sale?
Best Regards
Martyn
Precision Test Systems UK & USA Contact Details:
This email is from Martyn Smith
United Kingdom
Precision Test Systems Ltd
The Studio, Whitehouse Farm,
New
Hello,
First of all quick apologies for sending an email last week without deleting
all the old information from previous postings.
I have been playing around with producing 32.768 kHz from 10 MHz using a PIC
chip.
I have a real time clock calendar chip that requires a 32.768 kHz
Hello,
I have a strange fault on a PRS10. Wondered if anyone had seen it.
I am running a PRS10 and at the same time monitoring its FC (the frequency
control value) and Status byte data values.
So normally, the Status shows are all zeros, except 2nd from last is a 1,
showing the unit is not
Hello,
I have a customer who is measuring the phase noise of my 10 MHz ultra-low phase
noise frequency standard.
He is seeing spurious signals at line frequencies (50 and 100 Hz as we are in
Europe) at a level around -130 dBc.
My opinion is that it's impossible to get much better than that.
77 matches
Mail list logo