); SAEximRunCond expanded to false Errors-To: [EMAIL PROTECTED] Tom Van Baak wrote: > Richard, > > A 1.6 us window mean you have almost no issues with > the accuracy or stability of your 100 MHz sample clock. > 10 ns out of 1.6 us is 1/2 percent; clock counts won't > exceed 160; a quartz timebase is overkill. > > Do I understand correctly: you make each raw 1pps > time interval measurement down to 10 ns resolution, > then (in software, I presume, one second later) apply > a negative sawtooth correction with 1 ns resolution, then > average 120 of those sums, and then expect a 84 ps > resolution result? Something doesn't sound quite right. > > /tvb > Tom
If the quantisation error of the sawtooth correction is random then averaging will reduce the noise in the average sawtooth correction to somewhat below 1ns. If the 100MHz oscillator phase drifts randomly with respect to the phase of the OCXO being disciplined then averaging will indeed reduce the noise of the phase measurement to below 10ns, however although the calculated resolution is 83ps the noise in the result will be somewhat higher than this. Indeed a quartz timebase is perhaps too stable for effective averaging unless the PPS signal has around 10ns or so rms jitter on its leading edge. However if a less stable timebase is used to ensure it has sufficient short term instability to ensure accurate averaging it will be necessary to measure/calibrate its frequency perhaps every second to correct for drift in the timebase oscillator. Relying on the 100MHz crystal oscillator phase drifting around in a random fashion so that the phase error averages are unbiased estimates of the true phase error is perhaps expecting too much unless heroic measures are taken to ensure that the 100MHz oscillator doesnt phase lock to the OCXO output via injection locking. The phase error averaging will be improved by adding sufficient jitter to the PPS signal to increase its random timing jitter to around 10ns rms or so. If this is done the 100MHz clock can be derived from the oscillator being disciplined and the averaged phase will still have a very small bias. Bruce _______________________________________________ time-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.