Hi Patrick:

There are a bunch of signal sources that could be used.

The rack sized HP signal generators that weigh 50+ pounds were designed to have very low phase noise and although now many years old are still very good for that.

The HP 3325() is the only signal generator that has adjustable amplitude that's calibrated to a small fraction of a dB. They still are the specified instrument for amplitude calibration of many other pieces of lab equipment and there is no newer instrument to replace them. They don't have good specs in the frequency domain.

The HP 8648() series are reasonably small and light synthesized signal generators. I got the 8648A Option 1EP which is specific for pager testing and includes many modulation enhancements. The prior versions of the 8648 did not have good enough specs to test pagers.
http://www.prc68.com/I/HP8648.shtml

A problem with injecting a signal into operating equipment is you may burn out the signal generator. For example just touching a DC point feeds a step change back into the sig gen. So it's good to have a series resistor and blocking cap to protect the sig gen.

You might be able to use a probe that generates something like narrow pulses at an audio rate as a universal signal source.

Have Fun,

Brooke Clarke
http://www.prc68.com

Patrick wrote:
Hey everyone

Sorry for the off topic post. I have received great advice in the past
with items for my little shop and I can't resist to ask again.

I am thinking about buying a signal generator. I suspect that I will
mostly use it to inject low uV/mV signals into the amplification stages
of the laboratory instruments I service.

Any feedback you have would be greatly appreciated-Patrick

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to