_Correction_

Oops, Its the CNT91 not the CNT81 that actually does the regression fit.

How did you achieve MDEV ~1E-13 @10s with a counter rated at 50ps single shot resolution?


Bruce

Bruce Griffiths wrote:
Read the data sheet and the various application notes/white papers on the Pendulum site. The intrinsic resolution of the Pendulum counter (50ps) is slightly inferior to that of the SR620 and HP5370A/B. What they do is statistically process the results of a series of measurements of the input phase taken at short intervals. They actually fit a regression line to the resultant series of phase measurements.
This process inherently filters out some of the noise of the counter.
If it were possible to do the same thing using an SR620 or HP5370 the noise in the output resolution would be even lower.

If one is building a conventional DMTD one doesn't actually need resolution for the timestamping device/counter much better than 10ns or so to achieve a resolution of around 1E-13/Tau with say a 100Hz beat frequency and 10MHz inputs to the mixer/phase detector.

Bruce

Bob Camp wrote:
Hi

At least from what I've seen, the Pendulum's seem to work a bit better than the other counters you mention. That may simply be a function of their being designed much more recently. It could also be the issue of comparing beat up stuff on the bench to brand new stuff on the bench. The CNT-81 is rated to have a much better single shot time resolution than the others.

Yes I realize that in no way addresses the question you asked.

MDEV and ADEV measure slightly different things. Depending on what you are looking for MDEV may give you better insight.

Bob


On Feb 5, 2010, at 10:59 PM, Pete Rawson wrote:

Efforts are underway to develop a low cost DMTD apparatus with
demonstrated stability measurements of 1E-13 in 1s. It seems that
existing TI counters can reach this goal in 10s. (using MDEV estimate
or 100+s. using ADEV estimate). The question is; does the MDEV tool
provide an appropriate measure of stability in this time range, or is
the ADEV estimate a more correct answer?

The TI performance I'm referring to is the 20-25 ps, single shot TI,
typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
from my CNT81showing MDEV<  1E-13 in 10s. and I believe the
other counters behave similarly.

I would appreciate any comments or observations on this topic.
My motivation is to discover the simplest scheme for making
stability measurements at this performance level; this is NOT
even close to the state-of-the-art, but can still be useful.

Pete Rawson

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to