Arthur Dent wrote:
The +12VDC supply (internal to the sealed oscillator) supplies both the oven 
and the oscillator circuits. I think you'll find that internally this +12VDC 
goes to the heater circuit as well as through a regulator to the oscillator 
which is running on something like +7VDC. This is speculation at this point but 
a number of similar type oscillators I've checked are set up this way and even 
have a temperature compensated regulated reference voltage out of about +7VDC 
to feed a pot or EFC circuit. For instance the HP 10811 has an internal +5.7VDC 
regulated supply for the oscillator circuit

One crude way to check is to take a removed 10Mhz oscillator Trimble oscillator 
and power it from a bench supply at +12 and let it stabilize. Then reduce the 
supply voltage slowly and see where the output level and/or frequency start to 
change. If you see no change until you've reduced the voltage a few volts then 
the oscillator has an
 internal regulator for the oscillator circuit.
Every oscillator has a spec for frequency shift due to power supply variation, but maybe you'd see a big change when you lost regulation. Have to think about that.

Ed


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to