On 18/08/11 05:10, Paul Cianciolo wrote:
Hello Folks,
I am looking for an instrument that is better than Fluke 103a comparator.
The purpose of this piece of equipment is help me learn more about oscillators 
and characterizing them

The HP5370B is the TIC I keep hearing about but I am not glued to that make or 
model.
I wou;d like to hear suggestion from the group.

Could get you started. For good quality oscillators the resolution will still be a limiting factor for short taus. For longer taus it should give you good bang for the buck. GPIB interface should be something you also invest in.

Another way to go might be to build one of those units where there is a common 
oscillator is split and feeds the LO of 2 mixers.
The RF side comes from the DUT and the REF
Sorry I cannot remember the acronym... MDMM????

Dual Mixer Time Difference (DMTD)

You may have to look for a suitable offset oscillator, since you do want a low noise one. Also consider that you need to feed it two signals, one reference and one DUT. You will see the sum of both signals.

With a third mixer you can do three-cornered hat things to resolve the phase noise of the individual oscillators. Naturally there is some problems with all that, and isolation of the sources becomes important. Ah well.

Cheers,
Magnus

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to