Holzworth's HX2410 <http://www.holzworth.com/Spec_sheets/HX2410_Web_Datasheet.pdf> claims femtosecond jitter albeit with CMOS outputs. This is difficult to achieve (especially at the lower end of its input frequency range) with a comparator with 1ps of internal jitter.

A simple differential pair with the appropriate gain and bandwidth will have somewhat lower jitter than a comparator, however as long as the jitter is lower than that of the following logic achieving the lowest possible output jitter from the sine to LVDS shaper may be somewhat pointless.

Since the input signal slew rate is so high a transformer with a centre tap biased halfway between the LVDS input levels with a pair of inverse parallel clamp diodes across the secondary plus a couple of low value series resistors between the secondary and the clamp diodes should suffice. Similar schemes are used by most of the high resolution pipeline ADCs with sample clock frequencies of 50MHz or more.
These ADCs typically have sub picosecond sampling jitter.
However some LVDS receivers have built in attenuators to protect the input so adding the diodes may not be required.


Bruce

Bruce

Luis Cupido wrote:
I had similar need some time ago and
I found that a differential pair with two (pnp) BFR93
worked much better than any comparator(three or four tested, but not the adcmp604).
(was a pll reference and I judged the impact of such
observing the phase noise at microwaves).

lc.
ct1dmk.



On 9/9/2011 12:29 PM, Javier Herrero wrote:
Thanks for the reminder :)

Te available 180MHz signal has that level, and it is quite heavily
bandpass filtered, since it comes from a multiplier chain from an 45MHz
OCXO, so the ADCMP604 will fit nicely

Best regards,

Javier

El 09/09/2011 13:07, Bruce Griffiths escribió:
Javier Herrero wrote:
Hello all,

I think that the same question that has been discuted here a zillion
times but usually around 10MHz... anyway, what would be the best way
to convert a sine wave to a LVDS clock (preferably duty cycle 50%) at
180MHz?

Texas Instruments suggest a LVDS receiver as a comparator
http://www.ti.com/lit/an/slyt180/slyt180.pdf but time ago this was
discussed here, and not very favoured due to the high hysteresis of
the LVDS receivers.

Regards,

Javier

Use a true LVDS comparator (e.g. ADCMP604).
With 1:1 transformer coupled input and a pair of inverse parallel
schottky diodes across the inputs together with series resistors
between the transformer secondary to the comparator inputs should
suffice for inputs of +16dBm or more.
With sufficient input from a low noise source a cycle to cycle jitter
of 1-2ps should be feasible.

Sub picosecond jitter is feasible if one cascades a series of low pass
filtered limiter stages.

Bruce

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to