On 14/09/11 23:34, Paul A. Cianciolo wrote:
Murray,
Thank you.

When an OCXO is unpowered for some time, it undergoes a further ageing
process when it starts up next. This is specified in some data sheets, and
can typically require an hour of burn-in for each day left unpowered before
specifications are met.

This is one thing I was curious about.  I am asking these questions because
I am working with a Rakon/C-MaC  double oven OCXO.
As I make changes to the power supply, cabinet, EFC power supply,
surrounding this OCXO, I wondered  how long to let the unit run before I run
a standard test on it using TimeLab.
Maybe 2 days would be a good compromise?

I let my OCXOs at work simmer for a couple of days before I really trust the data... but using a direct ADEV is mearly an approximation as the upper end will be drift limited. I use HDEV to pre-view what a drift compensated curve would be, because Hadamard Deviation will remove first degree drift, but true OCXO drift isn't linear, so using a curve-matching gives better info.

It is actually recommended to run 1h test-runs back to back and see how the "high end" reveals itself more and more as the drift falls back while shorter taus remains fairly unchanged in the process.

TimeLab is good in that it will allow you to flip between ADEV and HDEV while running. Looking at the high-end curve flipping and flapping around while gently being pushed up by more samples is also educating.

Cheers,
Magnus