List,

The other day I had my second HP 3586B power supply board smoked.  It burned 
both the PC board traces and the connector on the motherboard so badly it will 
have to be replaced.  Also an inch or so of one trace on the mother board but 
it looks like it will be repairable.  Fortunately I have a donor board for the 
connector but it will probably be an all-day project to do the repair properly.

An observation.  There AFAIK a fold-back current limiting circuit but it didn’t 
help and the correct value of primary fuse did not blow.  This made me an 
extremely unhappy camper.  I’m also going to try and figure out how to add 
additional fuses.  

So I went through the entire instrument and came up with a list of all the 
Sprague TVA electrolytics that I am going to replace as they are dated coded 
1983.

So two questions.  One, can I safely double the capacitance of the filter 
capacitors?  (I plan on using the 105C 10K hour high reliability Nichicon or 
Panasonic units.)  

Secondly, the tantalum filter caps seem OK but can they be replaced with the 
same high quality aluminum electrolytics perhaps of a higher value of 
capacitance?

Additionally, I plan on doing the same to all my vintage HP test equipment.  A 
smoked 5370B would really, really, really, ruin my week.

Opinions and solutions sought.

Regards,

Perrier 