Just a couple of comments from a RF guy and FMT-nut

10 MHz clock distribution for the lab

 

 In the small coax cables one might consider RG-223. This is a 1/4 inch more or 
less diameter cable that features two concentric braided shields. Each is a 
very dense weave and is silver plated. This a very high quality mil spec coax. 
the main drawback is that a normal BNC fitting will not pass the entire cable 
through the clamp nut or most of the shield through the hole in the internal 
ferrule. However, Pasternack, among others has the correct connectors for BNC, 
TNC, and type N. I have used this cable with good success to keep RF from 
radiating in test setup. It is way way better than even quality RG-58. I found 
part of a reel on eBay for next to nothing.

How many instruments? Some time ago I surveyed a number of instruments to 
determine how many had an actual 50 OHM input for the 10 MHz reference. Answer 
- some do, most don't, though many specify something like 1V rms from a 50 Ohm 
source.

I have been able to operate 1 spectrum analyzer, 2 Fluke synthesizers, A hp 
synthesizer, and a HP 5345 counter all at the same time from a LPRO RB source 
using Ts and only terminating at the end of the line. Actually the termination 
had no effect as I was not looking at phase. Most of these devices seem to 
present a relatively high Z load and as long as that is the case; the simple T 
scheme is OK as long as there are not too many and the distances between them 
are short in terms of wavelength. 

It is easy enough to crudely determine the input Z using a series 1 K resistor 
on the source side.and just measuring the voltage drop with a scope.

I did get a distribution amp as things got more complex around here, but I 
never had a problem with the simpler scheme either.

-john k6iql



 
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to