On 05/09/2013 01:52 AM, Bob Camp wrote:
Hi

Clip a DVM on the power leads of the OCXO and see if the supply swings at the same 
rate as the frequency. It doesn't have to be anything fancy. If the problem is the 
supply, the swing will be obvious with>= 3 1/2 digits of resolution.

To be *really* sure, put a DVM in series with the oven supply lead and watch 
the current. That will also tell you if it's cycling due to an internal problem.

I've seen ovens that had this issue, and it out of luck that I saw the current fluctuate on power-up (as the HP supply had both voltage and current reading) and then I hooked up a DMM in series and could see the same fluctuation in the more fine-grained structure too. Turned out to be a badly designed oven, or rather, the original design was good enough, but then transferred to another product with new conditions, it was no longer stable.

It can be good to understand what that "bump" comes from, it can be a severe systematic flaw, but ADEV doesn't help you to understand what it is.

Cheers,
Magnus
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to