I don't have experience on this DMM but usually the Ohms sense terminals and the correspondent current source ones are internally connected by resistors that become shorted at the Kelvin connected ends when used in 4 wire connection. If any of this resistors became open you meter will do weird things when in 2 wire Ohms measurements.

Regards,
Ignacio EB4APL


El 18/05/2013 3:23, Volker Esper wrote

Hi,

I bought a 34401A lately. It's all fine, except it's got a fault when measuring low ohm with 2-wires: the value runs (monotonic) from 0 to 30 oder 40 Ohms at a speed of say 1 ohm per second and after a while it decides to run back... When using 4-wires-measurement everything's alright.

Did anyone have that problem before?

Thank you

Volker