Thank you, Bob, yes, that was my first guess, too. I opened it and
shortened the relay contacts with an extremly low ohms short and the
display changed significantly when coming to K101. I have to get one.
Besides, after fumbling and knocking and all that handling the
instrument is now quite stable - that lasts till the next hyper
important unattended long time measurement...
Thanks a lot
Volker
Am 18.05.2013 03:46, schrieb Bob Camp:
Hi
Never seen that problem, but it sounds very much like a relay / mux problem in
the ohms section.
Bob
On May 17, 2013, at 9:23 PM, Volker Esper<[email protected]> wrote:
Hi,
I bought a 34401A lately. It's all fine, except it's got a fault when measuring
low ohm with 2-wires: the value runs (monotonic) from 0 to 30 oder 40 Ohms at a
speed of say 1 ohm per second and after a while it decides to run back... When
using 4-wires-measurement everything's alright.
Did anyone have that problem before?
Thank you
Volker