Bob,

Sorry, I'm not sure, if I've understood the issue - what exactly did you wrong?

Thank you

Volker


Am 08.07.2013 14:48, schrieb Robert Darby:
John,

After a night's sleep and a rereading of your post I finally realized what I was doing wrong. I did not understand the the role of the sampling interval setting and the display rate setting on the 5370. When I follow the process below the results are totally consistent.

Sorry to taken your time and thanks all for your help. Now to find the original issue.....

Bob Darby

On 7/7/2013 8:57 PM, John Miles wrote:
So the only difference between the test setups is the setting of the Display Rate control on the 5370, correct? You're allowing TimeLab to estimate the sample rate automatically, and giving it enough time to converge on a stable reading before hitting 'Start Measurement'? You're correct in that changing the real-world sample rate should yield results that are identical (or at least very similar) to resampling the phase data after the fact. In frequency mode, dead time between readings would make that an iffy proposition, but for data taken in TI mode the outcomes should be close. -- john, KE5FX Miles Design LLC