Hello:
  Thanks TVB and Magnus's comment, I think I should ignore that last digit 
value, it's useless in result. Yes the oscillator in my test is very poor, 
beacuse it's a OCXO warm-up procedure, the counter's output digit was not 
regular, I don't know when it add or reduce the digit, but I found in a 
stablility frequency very close 10Mhz, it not happened. So I powered on a OCXO 
and observe the output digit in a large range of frequency change. 
 


Hui Zhang








At 2013-09-01 01:49:49,"Tom Van Baak" <[email protected]> wrote:
>Hi Hui Zhang,
>
>Yes, I agree it looks strange, but this is normal and correct for the 53132; 
>or modern frequency counters like it. When it drops a decimal place on the 
>right it is doing you a favor because that last digit is known to be noise 
>rather than a real digit of measurement. HP/Agilent has always been concerned 
>with accuracy and truth.
>
>In frequency mode, these counters use oversampling algorithms to report the 
>*best fit* frequency. It works well until the frequency gets very, very close 
>an exact fractional/multiple of 10 MHz. Then the algorithms no longer improve 
>phase resolution and the instrument becomes more like a conventional 10 or 11 
>digit/second frequency counter. The output is still accurate, but the 
>precision is reduced accordingly. Again, this is normal. If you've ever 
>designed a high-resolution frequency counter you understand why.
>
>Amazingly (and to their credit), the 53132 counter firmware detects this 
>sub-standard condition and removes digits from the output. It does this 
>because those digits are recognized to be meaningless. You will also notice in 
>statistics mode it replaces low order digit(s) with a '*' character. Some 
>engineer at HP did a lot of work to make sure the instrument did not report 
>more resolution than was valid.
>
>There is a footnote in the 53131/53132 manual that explains this:
>http://leapsecond.com/pages/53132/53132-reduced-resolution.gif
>
>In your case, your oscillator is quite poor. Your oscillator varies by 
>hundreds of microHertz each measurement. The ADEV is about 1e-10. In this 
>case, it is not a counter problem; it is instability in your oscillator (or 
>reference). The good news is that for this type of oscillator, your 53132 
>counter performs well and you do not need a more expensive time/frequency 
>comparison instrument.
>
>If you need more information, let me know.
>
>/tvb
>
>----- Original Message ----- 
>From: "Hui Zhang" <[email protected]>
>To: "Discussion of precise time and frequency measurement" <[email protected]>
>Sent: Saturday, August 31, 2013 8:06 AM
>Subject: [time-nuts] HP53132A's LSD question
>
>
>> Dear group:
>> 
>>  I found a strange thing about HP53132A’s output digit, I know the HP53132A 
>> is a universal counter that have 12 digits per second performance by read 
>> the manual. By use 10s gate time, I thought it will be 13 digits resolution. 
>> But I found it can output more digits reads by use GPIB or RS-232 way to 
>> collect datas. (This is simple way rather than use offset method). I use 
>> RS-232 cable to connected to my counter to observe the frequency of a OCXO, 
>> I got these output result ( a part of long sequence):
>> 
>> 9.999,967,286,098,0  MHz
>> 9.999,981,814,726,9  MHz
>> 9.999,991,573,702,0  MHz
>> 9.999,997,482,253,0  MHz
>> 9.999,999,945,685  MHz
>> 10.000,000,472,030,4 MHz
>> 10.000,000,460,991,3 MHz
>> 10.000,000,352,570,9 MHz
>> 10.000,000,254,108  MHz
>> 10.000,000,183,759  MHz
>> 10.000,000,136,842  MHz
>> ..
>> . (Omitted few lines)
>> .
>> 10.000,000,039,724  MHz
>> 10.000,000,006,299  MHz
>> 9.999,999,992,692  MHz
>> 9.999,999,990,674  MHz
>> 9.999,999,989,094  MHz
>> 9.999,999,995,662  MHz
>> 9.999,999,993,829  MHz
>> 
>> The question is in first four lines the reads is 14 digits, but in line 5, 
>> the digits reduce to 13 digits, and then, it’s became 15 digits in line 
>> 6(because over 10Mhz). After 3 lines, the reads reduce one digits again to 
>> became 14 digits(also over 10Mhz),  again after that few lines the frequency 
>> less that 10Mhz, it return to 13 digits like last few line. My question is 
>> what determines the HP53132A’s resolution? Why the output is variable digits 
>> even the had very close frequency? Any comment will be appreciate.
>> 
>> 
>> 
>> Hui Zhang
>
>
>_______________________________________________
>time-nuts mailing list -- [email protected]
>To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts