Sorry the screen shot's url is:

http://hkbbs.leowood.net:88/incomefiles/20139/19144671911.JPG




At 2013-09-01 09:05:56,"Hui Zhang" <[email protected]> wrote:
>Hello Bob:
>   That's good idea, I will try it at right time, in fact I collect about 300 
> data by measured a OCXO warm-up, and I made a excel chart to analize the 
> relative of output digit and frequency. but it's very crudely because I use 
> little time to do it. I don't have a high resolution DAC, I will find a way 
> to do this test late. Thank you.
>
>
>Here is a screen shot of my chart:
>
>
>http://hkbbs.leowood.net:88/incomefiles/20139/18581578616.JPG
>
>
>
>Hui Zhang
>
>At 2013-09-01 07:51:24,"Bob Camp" <[email protected]> wrote:
>>Hi
>>
>>One thing to look at -
>>
>>Take a large set of data and then sort it by reading. What you will find is 
>>that there are only certain numbers that come up. You can directly observe 
>>this with a good OCXO and a high resolution input to the EFC.
>>
>>Bob
>>
>>On Aug 31, 2013, at 11:06 AM, Hui Zhang <[email protected]> wrote:
>>
>>> Dear group:
>>> 
>>>  I found a strange thing about HP53132A’s output digit, I know the HP53132A 
>>> is a universal counter that have 12 digits per second performance by read 
>>> the manual. By use 10s gate time, I thought it will be 13 digits 
>>> resolution. But I found it can output more digits reads by use GPIB or 
>>> RS-232 way to collect datas. (This is simple way rather than use offset 
>>> method). I use RS-232 cable to connected to my counter to observe the 
>>> frequency of a OCXO, I got these output result ( a part of long sequence):
>>> 
>>> 9.999,967,286,098,0  MHz
>>> 9.999,981,814,726,9  MHz
>>> 9.999,991,573,702,0  MHz
>>> 9.999,997,482,253,0  MHz
>>> 9.999,999,945,685  MHz
>>> 10.000,000,472,030,4 MHz
>>> 10.000,000,460,991,3 MHz
>>> 10.000,000,352,570,9 MHz
>>> 10.000,000,254,108  MHz
>>> 10.000,000,183,759  MHz
>>> 10.000,000,136,842  MHz
>>> ..
>>> . (Omitted few lines)
>>> .
>>> 10.000,000,039,724  MHz
>>> 10.000,000,006,299  MHz
>>> 9.999,999,992,692  MHz
>>> 9.999,999,990,674  MHz
>>> 9.999,999,989,094  MHz
>>> 9.999,999,995,662  MHz
>>> 9.999,999,993,829  MHz
>>> 
>>> The question is in first four lines the reads is 14 digits, but in line 5, 
>>> the digits reduce to 13 digits, and then, it’s became 15 digits in line 
>>> 6(because over 10Mhz). After 3 lines, the reads reduce one digits again to 
>>> became 14 digits(also over 10Mhz),  again after that few lines the 
>>> frequency less that 10Mhz, it return to 13 digits like last few line. My 
>>> question is what determines the HP53132A’s resolution? Why the output is 
>>> variable digits even the had very close frequency? Any comment will be 
>>> appreciate.
>>> 
>>> 
>>> 
>>> Hui Zhang
>>> _______________________________________________
>>> time-nuts mailing list -- [email protected]
>>> To unsubscribe, go to 
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>
>>_______________________________________________
>>time-nuts mailing list -- [email protected]
>>To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>and follow the instructions there.
>_______________________________________________
>time-nuts mailing list -- [email protected]
>To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts