Well, in that case you do measure a form of noise-floor rather than measuring the OCXO performance, as OCXO performance is being corrected out in the PPS position (and sawtooth-correction).

If you want to measure that OCXO performance, you need to do that against an independent PPS.

Cheers,
Magnus

On 09/02/2014 06:37 PM, Bob Stewart wrote:
Hi Adrian,

Actually no.  This ADEV is a plot of the sawtooth-corrected 1PPS signal from 
the GPS receiver (LEA-6T) measured by the TIC in my GPSDO against the
output of its OCXO.

Bob


________________________________
  From: Adrian <[email protected]>
To: Bob Stewart <[email protected]>; Discussion of precise time and frequency 
measurement <[email protected]>
Sent: Tuesday, September 2, 2014 11:24 AM
Subject: Re: [time-nuts] OCXO Phase Noise Measurement in Primitive Conditions


Bob,

to me your ADEV plot looks pretty exactly like a measurement system
noise floor plot with its typical 10/decade shape.
I would assume that you connected the same signal to both counter
inputs, or, both sources have been phase locked.

Adrian

Bob Stewart schrieb:
Hi Bob,

I take it that the 3 segment ADEV pretty much defines a GPSDO.  So, is my plot 
an indication that I'm riding the OCXO too hard to get the results I think I 
should be getting?  What I think I should be getting is a flat phase line.  But 
is a GPS receiver, even after applying quantization error corrections, accurate 
enough to give that?  I see three things at play here: hardware, software, and 
my expectations.  I'm trying to figure out which one (or several) of these 
needs to be addressed.  I suppose I could hook the Rb back up and see the 
effect on the ADEV of the Rb vs OCXO as I loosen the gain values.


Bob



________________________________
   From: Bob Camp <[email protected]>
To: Bob Stewart <[email protected]>; Discussion of precise time and frequency 
measurement <[email protected]>
Sent: Tuesday, September 2, 2014 6:38 AM
Subject: Re: [time-nuts] OCXO Phase Noise Measurement in Primitive Conditions


Hi

Your ADEV is (mostly) a plot of 1-2 ns resolution out over your sample range. 
It’s not a “typical” three segment ADEV. It’s more likely that any departure 
from a straight line is random rather than systematic. Yes I’ve spent a lot of 
time looking at random plots that I was *sure* had significant information in 
them…..

Bob


On Sep 1, 2014, at 9:31 PM, Bob Stewart <[email protected]> wrote:

Hi Poul,

I've made a few hardware changes and some software changes, and I was wondering 
if you (and/or others) would take a look at this plot.  The green is phase over 
70,000+ seconds with ~100ps resolution, while the blue is the ADEV.  (The sharp 
departures in phase are from large temperature changes and not quite perfect 
temperature compensation software.)  I don't see your horizontal line down at 
the bottom, but there are some worrying horizontal lines between about 3tau and 
150tau.  Should I take that to mean that I have some noise/oscillations from 
about 1/3Hz to 1/150Hz?  Those up at about 150tau I think I can see by eye on 
another plot of different data.  I've got a lot of data collected from this run 
(still running) including pTerm, iTerm, dTerm, and tTerm (temperature) updates 
to the DAC.  I hope to try to tease something out of that later, though the 
pTerm and iTerm are mostly +1,0,-1 stuff.  FWIW, in real terms, the dGain is at 
about 4.2.  The phase
starts to get ugly if I bring it much below that.  iGain is .01 and pGain is 
.05.  There are misc damping and limiting factors applied, as well.


Just about ready to do a second board with updates from what I've learned, so 
any help is appreciated.  The plan is to release the source code at some (not 
too) future date and make boards available, if there's interest.  But I've 
still got a lot to do before that.


http://evoria.net/AE6RV/TIC/ADEV3.png


Bob




________________________________
From: Poul-Henning Kamp <[email protected]>
To: Bob Stewart <[email protected]>; Discussion of precise time and frequency 
measurement <[email protected]>
Sent: Thursday, August 28, 2014 1:04 AM
Subject: Re: [time-nuts] OCXO Phase Noise Measurement in Primitive Conditions


--------
In message <[email protected]>, Bob Ste
wart writes:

This looks like a step in the right direction.

The "correct" allan plot will have a clearly visible horizontal
segment somewhere in the 100-10000 second range (depending on OCXO quality).

This is where the GPS long-term stability "takes over" from the OCXO's
better short-term stability.
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to