Hi

>From what I’ve seen, many 620’s that have been sitting for a few years will 
>not operate until the calibration procedure is run. They drift far enough out 
>that they pitch errors trying to do simple things. Let the beast warm up for 
>an hour and run the cal as soon as you get a used one. 

Bob

> On Nov 21, 2014, at 9:34 PM, Hui Zhang <[email protected]> wrote:
> 
> Hi Bob:
>    I agree with you. But the chapter is a easy way to check your new counter 
> when it arrived, and it's not only a calibration of just self test, it's 
> testing counter's performance on some level.
> 
> 
> 
> 
> Regards.
> 
> 
> 
> 
> Hui Zhang 
> 
> 
> 
> At 2014-11-22 10:22:53, "Bob Camp" <[email protected]> wrote:
>> Hi
>> 
>> Probably the best information out there on setting up a SR620 is buried in 
>> the archives of this list back a few years. The back and forth required on 
>> some of the adjustments is far from obvious. You do need to start from the 
>> manual and get it as far as the manual will allow. You then go into the 
>> “Time Nuts advanced” procedures. 
>> 
>> Bob
>> 
>>> On Nov 21, 2014, at 9:19 PM, Hui Zhang <[email protected]> wrote:
>>> 
>>> The SR620's manual is perfect, there is a whole chapter has detail 
>>> information of performance test. You can use that way for check the 
>>> counter's condition.
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> Hui Zhang
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> 
>>> At 2014-11-22 07:51:41, "Dr. David Kirkby (Kirkby Microwave Ltd)" 
>>> <[email protected]> wrote:
>>>> After regrettably selling my 5370B a few years ago, I decided to look for a
>>>> used SR620.
>>>> 
>>>> Are there any particular issues with these that could be tested by the
>>>> seller prior to shipping it? I have not downloaded the manual yet, but I
>>>> assume there's a basic test in there - probably similar to that in the
>>>> 5370B manual.
>>>> 
>>>> Were these available with a Rb oscillator?  I see a seller on eBay that he
>>>> has some Rb ones for $1500 extra, but I don't see it as an option on the
>>>> SRS web site.
>>>> 
>>>> Dr David Kirkby
>>>> Managing Director
>>>> Kirkby Microwave Ltd
>>>> Registered office: Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3
>>>> 6DT, United Kingdom
>>>> Registered in England and Wales as company number 08914892
>>>> http://www.kirkbymicrowave.co.uk/
>>>> Tel 07910 441670 / +44 7910 441670 (0900-2100 GMT)
>>>> _______________________________________________
>>>> time-nuts mailing list -- [email protected]
>>>> To unsubscribe, go to 
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>> _______________________________________________
>>> time-nuts mailing list -- [email protected]
>>> To unsubscribe, go to 
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>> 
>> _______________________________________________
>> time-nuts mailing list -- [email protected]
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to