Thanks for these useful comments!

It seems there is no way to estimate frequency from looking at the phase
data then?
Does TimeLab have automated collection for the frequency-counter data on
the control port?

Anders


On Fri, Dec 11, 2015 at 2:51 AM, John Miles <[email protected]> wrote:

> > Hi all,
> >
> > A 5115A phase noise test set landed in our lab and I am wondering about
> the
> > data collection.
> > It has two telnet ports one for commands and one for phase data.
> > When issuing "start" on the command-port it starts spitting out
> > phase-difference values on the data port.
> >
> > However it seems to me that the phase data alone (REF-DUT phase, in units
> > of the REF period) is almost useless without knowing the internal
> workings
> > of the device.
> > The phase values are clearly not the true phase difference, which would
> > quickly accumulate to a huge number with e.g. REF=10MHz and DUT=11MHz.
> > My
> > understanding is that digital downconversion (DDC) is performed on both
> REF
> > and DUT signals, and without knowing the LO frequencies for the DDC the
> raw
> > phase data alone is almost useless.
> >
> > The command prompt does have commands for displaying the internally
> > calculated ADEV, L(f), frequency-counter readings, etc. and these seem to
> > show correct and OK values, but there seems to be no way to reproduce
> e.g.
> > the ADEV or frequency-counter values/statistics from the raw phase values
> > on the data port??
> > Is this correct? Any other experiences with the 5115A or higher end
> 5120A?
>
> The shorter-term ADEV values on the 5120A/5125A test sets are
> mathematically backed out of the phase noise data.  They will never
> perfectly match the ADEV values you get from plotting the phase data stream
> in my experience.
>
> I don't actually know what they do on the 5115A, though -- since it
> doesn't do cross-correlated PN, there's presumably no reason to derive the
> short-term ADEV plot from the PN data pipeline.  I'd expect the plots to
> match in that case, apart from any errors due to different measurement
> bandwidths and ADEV bin distributions.
>
> It's also true that there will always be an arbitrary phase slope error
> due to the mismatch between the frequency estimate used to tune the
> internal DDCs and the "real" frequency of the incoming data.  Knowledge of
> the actual DDC core frequencies would not be enough to correct for this
> behavior, because you would also need to know how far off they are.  The
> true frequency offset can't be measured ahead of time with perfect
> certainty, so it has to be estimated, and the resulting error will often
> dominate the slope of the phase-difference graph.
>
> The TimePod and 3120A test sets allow you to specify the input and
> reference frequencies explicitly to obtain a valid phase slope in residual
> measurements and other cases where the exact frequencies are known by the
> user.  But this is something you have to remember to do prior to the
> measurement, and you still don't get a calibrated absolute offset.
>
> (Also, note that TimeLab can read both phase and PN data from 51xx test
> sets without the need to use a Telnet client.  Not that it helps with this
> particular issue, of course.)
>
> -- john, KE5FX
> Miles Design LLC
>
>
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to