> It seems there is no way to estimate frequency from looking at the phase

> data then?

 

Not to my knowledge.  The frequency count chart on the instrument itself is 
always correct, since it can take the internal baseband offset out of the 
incoming data.  Because TimeLab doesn't have access to that offset, its 
frequency count chart won't be accurate when acquiring data from the 51xx.

 

> Does TimeLab have automated collection for the frequency-counter data on

> the control port?

 

TimeLab doesn't, but I have another collection of quick-and-dirty utilities 
that includes a way to do it (http://www.ke5fx.com/gpib/readme.htm).   The 
tsccmd.bat file can be run without any arguments to get command help; this 
option in particular will pull the frequency count entries across as ASCII text:

 

------------

   C:\Program Files (x86)\KE5FX\GPIB>tsccmd 192.168.1.225 "show fcounter"

   Using Windows Sockets V2.2 (WinSock 2.0)

   Initializing host JMCORE64, address 192.168.1.200

   Attempting to connect to server 192.168.1.225:1299 . . .

   Connected to server 192.168.1.225:1299

   Transmitting "show fcounter" ...

   show fcounter

   Welcome to the Symmetricom 5125A

 

   =192.168.1.225 > Reference Frequency: 100.0 MHz (auto)

 

   Avg Time (s)    Frequency (MHz)

   1               9.9999999627589

   10              9.99999996274766

   100             9.999999962746890

 

   Done, 8 line(s) received

   Average rate = 11.0/sec

------------

 

 

There's probably a way (wget?) to do something similar in most other OSes -- 
you basically need to send "show fcounter" to port 1299 with an appropriate 
timeout interval.  Then you can pipe the response to something that parses the 
text that comes back, if desired.

 

-- john, KE5FX

Miles Design LLC