Hi,

I agree with you. Some HP spectrum analyzer have  internal 50 and 100 Hz spurs 
down to -80dBc due the fan as described in the manual. It can be the spurs are 
from the test set and/or connection cables. If your oscillator is battery 
powered you can be sure the spurs are not from the source. Anyway I think the 
-130 dBc value is a very good response.

Luciano
www.timeok.it


From: "time-nuts" [email protected]
To: [email protected]
Cc: 
Date: Tue, 12 Jul 2016 11:44:31 +0100
Subject: [time-nuts] OXCO Spurious Output at Line Frequencies


Hello,
 
I have a customer who is measuring the phase noise of my 10 MHz ultra-low phase 
noise frequency standard.
 
He is seeing spurious signals at line frequencies (50 and 100 Hz as we are in 
Europe) at a level around -130 dBc.
 
My opinion is that it's impossible to get much better than that.  Even running 
on batteries make little difference, since the equipment is in a test rack with 
AC signals everywhere.
 
Even the £50k R&S test set he is using only quotes a spurious spec of -90 dBc.
 
What experience does anyone have here?
 
Best Regards
 
Martyn 
 
 