Hi

If you collect data over the entire second and average that down for a single 
point, then no, your ADEV will not be correct. 
There are a number of papers on this. What ADEV wants to see is a single phase 
“sample” at one second spacing. This is
also at the root of how you get 10 second ADEV. You don’t average the ten 1 
second data points. You throw nine data points
away and use one of them ( = you decimate the data ). 

What happens if you ignore this? Your curve looks “to good”. The resultant 
curve is *below* the real curve when plotted. 

A quick way to demonstrate this is to do ADEV with averaged vs decimated data ….

Bob

> On May 10, 2018, at 4:46 AM, Oleg Skydan <[email protected]> wrote:
> 
> Hi
> 
> I have got a pair of not so bad OCXOs (Morion GK85). I did some measurements, 
> the results may be interested to others (sorry if not), so I decided to post 
> them.
> 
> I ran a set of 5minutes long counter runs (two OCXOs were measured against 
> each other), each point is 1sec gate frequency measurement with different 
> number of timestamps used in LR calculation (from 10 till 5e6). The counter 
> provides continuous counting. As you can see I reach the HW limitations at 
> 5..6e-12 ADEV (1s tau) with only 1e5 timestamps. The results looks 
> reasonable, the theory predicts 27ps equivalent resolution with 1e5 
> timestamps, also the sqrt(N) law is clearly seen on the plots. I do not know 
> what is the limiting factor, if it is OCXOs or some counter HW.
> 
> I know there are HW problems, some of them were identified during this 
> experiment. They were expectable, cause HW is still just an ugly construction 
> made from the boards left in the "radio junk box" from the other 
> projects/experiments. I am going to move to the well designed PCB with some 
> improvements in HW (and more or less "normal" analog frontend with good 
> comparator, ADCMP604 or something similar, for the "low frequency" input). 
> But I want to finish my initial tests, it should help with the HW design.
> 
> Now I have some questions. As you know I am experimenting with the counter 
> that uses LR calculations to improve its resolution. The LR data for each 
> measurement is collected during the gate time only, also measurements are 
> continuous. Will the ADEV be calculated correctly from such measurements? I 
> understand that any averaging for the time window larger then single 
> measurement time will spoil the ADEV plot. Also I understand that using LR 
> can result in incorrect frequency estimate for the signal with large drift 
> (should not be a problem for the discussed measurements, at least for the 
> numbers we are talking about).
> 
> Does the ADEV plots I got looks reasonable for the used "mid range" OCXOs 
> (see the second plot for the long run test)?
> 
> BTW, I see I can interface GPS module to my counter without additional HW 
> (except the module itself, do not worry it will not be another DIY GPSDO, 
> probably :-) ). I will try to do it. The initial idea is not try to lock the 
> reference OCXO to GPS, instead I will just measure GPS against REF and will 
> make corrections using pure math in SW. I see some advantages with such 
> design - no hi resolution DAC, reference for DAC, no loop, no additional 
> hardware at all - only the GPS module and software :) (it is in the spirit of 
> this project)... Of cause I will not have reference signal that can be used 
> outside the counter, I think I can live with it. It worth to do some 
> experiments.
> 
> Best!
> Oleg UR3IQO 
> <Снимок экрана (1148).png><Снимок экрана (1150).png><Снимок экрана 
> (1149).png>_______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.