// How long should he run it continuously to recover? Before sending it for
calibration?

// How long should he run it continuously after changing the oven temperature?

David,

the real question is, if the LTZ1000A is affected at all by this timely 
hysteresis.
SN-18 gives no hint and no help, just lets you leap in the dark.

So I proposed to use a known good external reference (stable over time, near 
zero T.C.) for comparison. It's not requested by SN-18, that's correct, but if 
you have one, you can decide within 24h, simply by monitoring the 3458A over 
that time.
Prerequisite is a stable environment (+/-0.1°C over 24h), otherwise you'll only 
see the different T.C.s of the 3458A.

The drift of the U180 A/D hybrid should also be visible by such a measurement, 
but you can as well do that w/o an external reference.
But any temperature change will also influence CAL? 72 parameter, as this 
parameter is used for ACAL-compensation of any drifts of U180.

Maximum T.C. for U 180 is on the order of 0.35ppm/K, as given in the datasheet, 
i.e. 0.5ppm/K overall (w/o ACAL) minus 0.15ppm/K for the LTZ (with ACAL)

My U180 has about -0.24ppm/K, as in the picture which I'll append. I've 
measured this parameter at different temperatures, especially also at 25°C, 
making an ACAL DCV directly after power on.
You also see, that my instrument / U180 seems to be quite stable over time, 
during the season with constant temperature in the basement.

Sorry Illya, but your 0.01ppm/day criterion can not be determined practically, 
because it's hardly possible to extract such a low timely drift from thermal 
fluctuations .. this would require a stable room temperature on the order of < 
1/35°C over a whole week, or so.

A more practical approach is to observe CAL? 72 and TEMP?, after each ACAL DCV 
over one week, and then trying to separate the temperature effect from a 
possible gross timely effect.

Everything else, if it's an instrument which is OK only (like mine), or even a 
very good performer, has to be extracted from a longer observation time.

Frank