Hello Dave,

At first, this error 110 is probably caused by a hardware error. The full error message, if you scroll right might read: 110 "Calibration Required -- ACAL". On eevblog, we had a 3458A with exactly this message, which was caused by a faulty OpAmp in the input MUX, which was successfully repaired, with remote failure tracing.

So I doubt that a CAL is required.

Memory errors concerning the three nvSRAMs usually create different messages, I think it's # 212 or alike.

Therefore, please make a complete self test with the front panel TEST function, and report back, which additional errors show up.

Your serial number relates to a basic instrument from around 1995, which may have been refurbished, at least the front panel in the last 5 years.

The high calibration count denotes that it still has the old processor board with 6 EPROMs.
Probably it has old nvSRAM, with nearly exhausted batteries.
Both you can easily see if you open the upper case lid and read the date codes of the 3 DALLAS RAMs. Please report back.

I can't recommend at all to 'simply' let KS calibrate your instrument, for several reasons.

If the DALLAS RAMs are affected, they will replace the A5 µP board for 1-2k$. You could instead save the CAL RAM and exchange the RAMs on your own for much less money, even if you'd have to buy an appropriate programmer.

If there's a HW error on the analogue board, KS will as well exchange the whole board for 2-3k$, which you might repair yourself instead.

Now, I just want to wish you a successful knee surgery, and then discussing your case soon.

Frank



_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com