RE: IEC 60601-2-24 / Magnetic Field

2002-05-03 Thread Ned Devine

Hi,

Yes.  Triple checked.

Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail

Entela, Inc. A Certified Woman Owned Business
www.entela.com 





-Original Message-
From: plaw...@west.net [mailto:plaw...@west.net]
Sent: Thursday, May 02, 2002 11:24 AM
To: EMC-PSTC
Subject: Re: IEC 60601-2-24 / Magnetic Field



Are you sure about the IEC 60521 reference?
  IEC 60521 (1988-03)
  Class 0.5, 1 and 2 alternating-current watthour meters

  Applies only to newly manufactured induction type watt-hour meters of
  accuracy classes 0.5, 1 and 2, for the measurement of alternating current
  electrical active energy of a frequency in the range 45 Hz to 65 Hz and to
  their type tests only. This publication supersedes IEC 60043 (1960), 60170
  (1964) and 60280 (1968). 


Patrick Lawler
plaw...@west.net

On Mon, 29 Apr 2002 10:55:57 -0400, Ned Devine ndev...@entela.com wrote:
I can help much, but, for this paragraph, the rationale at the back of the
standard states that Annex AAA of the Collateral Standard IEC 60601-1-2
states that the limits and methodology are under consideration by technical
committee 77.  This Particular Standard, however, refers for the time being
to IEC 60521 in which 400 A/m is required.

For the Second Edition of IEC 60601-1-2, Clause 36.202.8.1, the limit is 3
A/m.

Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail

Entela, Inc. A Certified Woman Owned Business
www.entela.com 

-Original Message-
From: am...@westin-emission.no [mailto:am...@westin-emission.no]
Sent: Sunday, April 28, 2002 3:40 PM
To: emc-p...@majordomo.ieee.org
Subject: IEC 60601-2-24 / Magnetic Field

Hi all,
A colleague was checking the IEC 60601-2-24:98 and found the Magnetic Field
Req't (paragraph 36.202.6) far too high: 400 Ampere-per-meter! I'd like to
check this requirement for its correctness or history with backgrounds for
such a demand. I know that 3A/m, 10A/m and 30A/m are often used, but 400A/m
.

Can anybody help?

Best regards
Amund Westin, Oslo/Norway

PS: IEC 60601-2-24:98 - Medical electrical equipment - Part 2-24:
Particular requirements for the safety of infusion pumps and controllers


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: IEC 60601-2-24 / Magnetic Field

2002-05-02 Thread Crabb, John

I can imagine that it is a perfectly reasonable requirement
that electricity meters be required to operate correctly
when subjected to high magnetic fields, to overcome any
likely fraud attempts by enterprising consumers with 
permanent magnets, trying to reduce their electricity bills.
I suspect that is the reason for the requirement in 
IEC 60521.

Regards,
John Crabb, Development Excellence (Product Safety) , 
NCR  Financial Solutions Group Ltd.,  Discovery Centre, 
3 Fulton Road, Dundee, Scotland, DD2 4SW
E-Mail :john.cr...@scotland.ncr.com
Tel: +44 (0)1382-592289  (direct ). Fax +44 (0)1382-622243.


-Original Message-
From: plaw...@west.net [mailto:plaw...@west.net]
Sent: 02 May 2002 16:24
To: EMC-PSTC
Subject: Re: IEC 60601-2-24 / Magnetic Field



Are you sure about the IEC 60521 reference?
  IEC 60521 (1988-03)
  Class 0.5, 1 and 2 alternating-current watthour meters

  Applies only to newly manufactured induction type watt-hour meters of
  accuracy classes 0.5, 1 and 2, for the measurement of alternating current
  electrical active energy of a frequency in the range 45 Hz to 65 Hz and to
  their type tests only. This publication supersedes IEC 60043 (1960), 60170
  (1964) and 60280 (1968). 


Patrick Lawler
plaw...@west.net

On Mon, 29 Apr 2002 10:55:57 -0400, Ned Devine ndev...@entela.com wrote:
I can help much, but, for this paragraph, the rationale at the back of the
standard states that Annex AAA of the Collateral Standard IEC 60601-1-2
states that the limits and methodology are under consideration by technical
committee 77.  This Particular Standard, however, refers for the time being
to IEC 60521 in which 400 A/m is required.

For the Second Edition of IEC 60601-1-2, Clause 36.202.8.1, the limit is 3
A/m.

Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail

Entela, Inc. A Certified Woman Owned Business
www.entela.com 

-Original Message-
From: am...@westin-emission.no [mailto:am...@westin-emission.no]
Sent: Sunday, April 28, 2002 3:40 PM
To: emc-p...@majordomo.ieee.org
Subject: IEC 60601-2-24 / Magnetic Field

Hi all,
A colleague was checking the IEC 60601-2-24:98 and found the Magnetic Field
Req't (paragraph 36.202.6) far too high: 400 Ampere-per-meter! I'd like to
check this requirement for its correctness or history with backgrounds for
such a demand. I know that 3A/m, 10A/m and 30A/m are often used, but 400A/m
.

Can anybody help?

Best regards
Amund Westin, Oslo/Norway

PS: IEC 60601-2-24:98 - Medical electrical equipment - Part 2-24:
Particular requirements for the safety of infusion pumps and controllers

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


Re: IEC 60601-2-24 / Magnetic Field

2002-05-02 Thread Patrick Lawler

Are you sure about the IEC 60521 reference?
  IEC 60521 (1988-03)
  Class 0.5, 1 and 2 alternating-current watthour meters

  Applies only to newly manufactured induction type watt-hour meters of
  accuracy classes 0.5, 1 and 2, for the measurement of alternating current
  electrical active energy of a frequency in the range 45 Hz to 65 Hz and to
  their type tests only. This publication supersedes IEC 60043 (1960), 60170
  (1964) and 60280 (1968). 


Patrick Lawler
plaw...@west.net

On Mon, 29 Apr 2002 10:55:57 -0400, Ned Devine ndev...@entela.com wrote:
I can help much, but, for this paragraph, the rationale at the back of the
standard states that Annex AAA of the Collateral Standard IEC 60601-1-2
states that the limits and methodology are under consideration by technical
committee 77.  This Particular Standard, however, refers for the time being
to IEC 60521 in which 400 A/m is required.

For the Second Edition of IEC 60601-1-2, Clause 36.202.8.1, the limit is 3
A/m.

Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail

Entela, Inc. A Certified Woman Owned Business
www.entela.com 

-Original Message-
From: am...@westin-emission.no [mailto:am...@westin-emission.no]
Sent: Sunday, April 28, 2002 3:40 PM
To: emc-p...@majordomo.ieee.org
Subject: IEC 60601-2-24 / Magnetic Field

Hi all,
A colleague was checking the IEC 60601-2-24:98 and found the Magnetic Field
Req't (paragraph 36.202.6) far too high: 400 Ampere-per-meter! I'd like to
check this requirement for its correctness or history with backgrounds for
such a demand. I know that 3A/m, 10A/m and 30A/m are often used, but 400A/m
.

Can anybody help?

Best regards
Amund Westin, Oslo/Norway

PS: IEC 60601-2-24:98 - Medical electrical equipment - Part 2-24:
Particular requirements for the safety of infusion pumps and controllers


---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


Re: IEC 60601-2-24 / Magnetic Field

2002-04-30 Thread Nick Rouse

I'm no medical Guru either but I know the sort of field levels involved.
They range from about .5 to 3 teslas. This is not much higher 
than the fields in the gaps of electrical motors and generators
(up to 1.8T in normal silicon iron and  up to 2.1 T in cobalt 
iron used in aerospace machines) The difference is the length of
the effective magnetic dipole. The field in an MRI machine is almost
constant across the whole bore of the machine and for whole body 
machines this gives a dipole length of a metre or more.  You are
still in the near field up to a couple of dipole lengths before the
field strength starts drooping as the inverse cube of the distance.
Thus it is quite possible to have fields up 1T a metre or more 
from the machine this corresponds to an H field of 790 kA/m
and this is sufficient to lift large ferromagnetic objects. The dipole
length in a normal electrical machine is about the size of the air 
gap, about 1mm or less. Thus there are very strong magnetic
fields only out to a few centimetres. I don't know if this is the 
reasoning behind the field values in IEC 60601-2-24  but
if it is the levels do not seem unreasonable.

Nick Rouse
.  
- Original Message - 
From: Chris Maxwell chris.maxw...@nettest.com
To: am...@westin-emission.no; emc-p...@majordomo.ieee.org
Sent: Tuesday, April 30, 2002 4:22 PM
Subject: RE: IEC 60601-2-24 / Magnetic Field


 
 Hi Amund,
 
 I read your email (below).  I'm no medical device expert; but I'm
 wondering if the magnetic strength limit was set so high due to MRI
 (Magnetic Resnonance Imaging) devices.  My understanding is that MRI
 devices produce HUGE magnetic field levels.  According to a recent
 newspaper story that I read, these fields are strong enough to draw a
 metal oxygen tank across the room to the MRI machine.
 
 Is there any chance that the device in question could be used around an
 MRI?   It would be a real bummer if someone's infusion pump suddenly
 changed its run rate due to a nearby magnetic field.
 
 Could that be what the standard is safeguarding against?  Any medical
 device gurus care to comment?
 
 
 
 Chris Maxwell | Design Engineer - Optical Division
 email chris.maxw...@nettest.com | dir +1 315 266 5128 | fax +1 315 797



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: IEC 60601-2-24 / Magnetic Field

2002-04-30 Thread Chris Maxwell

Hi Amund,

I read your email (below).  I'm no medical device expert; but I'm
wondering if the magnetic strength limit was set so high due to MRI
(Magnetic Resnonance Imaging) devices.  My understanding is that MRI
devices produce HUGE magnetic field levels.  According to a recent
newspaper story that I read, these fields are strong enough to draw a
metal oxygen tank across the room to the MRI machine.

Is there any chance that the device in question could be used around an
MRI?   It would be a real bummer if someone's infusion pump suddenly
changed its run rate due to a nearby magnetic field.

Could that be what the standard is safeguarding against?  Any medical
device gurus care to comment?



Chris Maxwell | Design Engineer - Optical Division
email chris.maxw...@nettest.com | dir +1 315 266 5128 | fax +1 315 797
8024

NetTest | 6 Rhoads Drive, Utica, NY 13502 | USA
web www.nettest.com | tel +1 315 797 4449 | 

 -Original Message-
 From: am...@westin-emission.no [mailto:am...@westin-emission.no]
 Sent: Sunday, April 28, 2002 3:40 PM
 To: emc-p...@majordomo.ieee.org
 Subject: IEC 60601-2-24 / Magnetic Field
 
 
 
 Hi all,
 
 A colleague was checking the IEC 60601-2-24:98 and found the Magnetic
 Field
 Req't (paragraph 36.202.6) far too high: 400 Ampere-per-meter! I'd
 like to
 check this requirement for its correctness or history with backgrounds
 for
 such a demand. I know that 3A/m, 10A/m and 30A/m are often used, but
 400A/m
 .
 
 Can anybody help?
 
 Best regards
 Amund Westin, Oslo/Norway
 
 PS: IEC 60601-2-24:98 - Medical electrical equipment - Part 2-24:
 Particular requirements for the safety of infusion pumps and
 controllers
 
 
 

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list


RE: IEC 60601-2-24 / Magnetic Field

2002-04-29 Thread Ned Devine

Hi,

I can help much, but, for this paragraph, the rationale at the back of the
standard states that Annex AAA of the Collateral Standard IEC 60601-1-2
states that the limits and methodology are under consideration by technical
committee 77.  This Particular Standard, however, refers for the time being
to IEC 60521 in which 400 A/m is required.

For the Second Edition of IEC 60601-1-2, Clause 36.202.8.1, the limit is 3
A/m.

Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI  49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail

Entela, Inc. A Certified Woman Owned Business
www.entela.com 





-Original Message-
From: am...@westin-emission.no [mailto:am...@westin-emission.no]
Sent: Sunday, April 28, 2002 3:40 PM
To: emc-p...@majordomo.ieee.org
Subject: IEC 60601-2-24 / Magnetic Field



Hi all,

A colleague was checking the IEC 60601-2-24:98 and found the Magnetic Field
Req't (paragraph 36.202.6) far too high: 400 Ampere-per-meter! I'd like to
check this requirement for its correctness or history with backgrounds for
such a demand. I know that 3A/m, 10A/m and 30A/m are often used, but 400A/m
.

Can anybody help?

Best regards
Amund Westin, Oslo/Norway

PS: IEC 60601-2-24:98 - Medical electrical equipment - Part 2-24:
Particular requirements for the safety of infusion pumps and controllers



---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list

---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://ieeepstc.mindcruiser.com/
Click on browse and then emc-pstc mailing list