Dear Alexandra and Leonid,

I do routinely small angle measurements starting from 0.6 degrees 2theta 
(Cu radiation) with the 1/32 divergence and 1/16 anti-scatter slit 
without beam knife, but never with the standard Panalytical stainless 
steel sample holder. It appears that this one gives a high background at 
very low angles. So you better use some polymer or glass sample holder. 
This works fine.
In addition you should be sure that your sample stage is well aligned 
(height and theta/2theta coupling). I know of a similar Xpert machine, 
where low-angle measurements were up until recently not possible, just 
because of improper alignment.


best regards, Arie

Leonid Solovyov wrote the following on 01/12/2009 15:41:

Dear Alexandra,

The 1/4 divergence and 1/2 anti-scatter slits are narrow enough to give a 
reasonably low background after 5 degrees 2Theta. First I would check that the 
background is not related to the sample itself. It may be done by measuring an 
empty sample holder.
The background may also be due to scattering from sample holder edges if the 
holder diameter is small and the primary beam mask is wide or/and the Soller 
slit divergence is big.
The automated sample changer can not produce background, but I am not sure if 
the beam knife can be installed together with the changer. The beam knife cuts 
the air scattering which is essential in the low-angle region. The standard 
beam knife supplied by PANal is rather thick and if it is installed too close 
to the sample surface it may attenuate intensity at high angles. With fixed 1/4 
divergence slit the knife should be put about 3-4 mm under the sample surface 
which must be safe for measurements up to 100 degrees 2Theta. To check if the 
knife attenuates high angle reflections it is necessary to measure the same 
sample with and without knife and then compare the intensities.
On my X’Pert I simply replaced the standard knife by a thin (ca. 1 mm) 
stainless steel screen,  and installed it 3 mm above the sample surface which 
allows preserving intensity up to 150 degrees 2Theta even with the 1/2 
divergence primary slit and have a reasonable low-angle background. Of course, 
the knife should also be properly centered on the diffractometer axis and 
aligned vertically.

Best regards,
Leonid

***
Leonid A. Solovyov
Institute of Chemistry and Chemical Technology
660049, K. Marx 42, Krasnoyarsk, Russia
www.icct.ru/eng/content/persons/Sol_LA
sites.google.com/site/solovyovleonid
***

--- On Mon, 11/30/09, Alexandra Seclaman seclaman_alexan...@yahoo.com wrote:


From: Alexandra Seclaman seclaman_alexan...@yahoo.com
Subject: RE: Configuration of PANalytical X'pert Pro
To: rietveld_l@ill.fr
Date: Monday, November 30, 2009, 2:05 PM
Thank
you for all of your answers, it was more than I hoped for. 


I did a Rietveld refinement on stibnite (Sb2S3). And I
mesured it from 5 2theta to 90. But  from 5 to 20 2Θ i
had a serious background noise and general trend similar to
that of glass. This trend and the background noise are there
no matter what I measure. 
I can't get to the lab until Wednesday, but from what I

recall, from the X ray source to the X'celerator
detector: there are 1/4 +1/2 fixed divergence,the widthest
mask and the slit (i'm not shure if that is the proper
name) in front of the X'celerator is also the widthest.

The automated system of handling the sample is
installed, should i changed it? 
And I forgot to mention, the system is MPD. 

I can give you more details on Wednesday.  


I'm sorry for my not technical terms, but is very hard
for me
 to speak about diffraction in English. 



Alexandra Seclaman

seclaman_alexan...@yahoo.com

a.c.secla...@gmail.com

--- On Sun, 11/29/09, Tony Raftery
a.raft...@qut.edu.au wrote:

From: Tony Raftery a.raft...@qut.edu.au
Subject: RE: Configuration of PANalytical X'pert Pro
To: Alexandra Seclaman
seclaman_alexan...@yahoo.com
Date: Sunday, November 29, 2009, 11:57 PM

Alexandra,

while there is no such thing as a standard configuration,
choices can be made to help in the modelling. Another
corresponant (to the Rietveld list)  made some
suggestions that I only partly agree with. A lot depends on
you goals ans available instrumentation - for instance if
you have an X'cellerator detector and a monochromator.
If you are interested in the low angle region (as I think
you are) or not. If you were looking at some phases or
phases where the peaks began after
 25 deg 2theta, you could afford to go to larger divergence
than if the peaks started at 5 deg 2theta.

I see little use for beam knifes except over restricted low
angle ranges (they always interfere at high angle, where
they interfere depends on how they are set). It see little
use for varable divergence slits for all sorts of reasons
(firstly as there is no rock solid conversion from automatic
to fixed intensities, secondly the resolution changes

RE: Configuration of PANalytical X'pert Pro

2009-12-01 Thread Leonid Solovyov
Dear Alexandra,

The 1/4 divergence and 1/2 anti-scatter slits are narrow enough to give a 
reasonably low background after 5 degrees 2Theta. First I would check that the 
background is not related to the sample itself. It may be done by measuring an 
empty sample holder.
The background may also be due to scattering from sample holder edges if the 
holder diameter is small and the primary beam mask is wide or/and the Soller 
slit divergence is big.
The automated sample changer can not produce background, but I am not sure if 
the beam knife can be installed together with the changer. The beam knife cuts 
the air scattering which is essential in the low-angle region. The standard 
beam knife supplied by PANal is rather thick and if it is installed too close 
to the sample surface it may attenuate intensity at high angles. With fixed 1/4 
divergence slit the knife should be put about 3-4 mm under the sample surface 
which must be safe for measurements up to 100 degrees 2Theta. To check if the 
knife attenuates high angle reflections it is necessary to measure the same 
sample with and without knife and then compare the intensities.
On my X’Pert I simply replaced the standard knife by a thin (ca. 1 mm) 
stainless steel screen,  and installed it 3 mm above the sample surface which 
allows preserving intensity up to 150 degrees 2Theta even with the 1/2 
divergence primary slit and have a reasonable low-angle background. Of course, 
the knife should also be properly centered on the diffractometer axis and 
aligned vertically.

Best regards,
Leonid

***
Leonid A. Solovyov
Institute of Chemistry and Chemical Technology
660049, K. Marx 42, Krasnoyarsk, Russia
www.icct.ru/eng/content/persons/Sol_LA
sites.google.com/site/solovyovleonid
***

--- On Mon, 11/30/09, Alexandra Seclaman seclaman_alexan...@yahoo.com wrote:

 From: Alexandra Seclaman seclaman_alexan...@yahoo.com
 Subject: RE: Configuration of PANalytical X'pert Pro
 To: rietveld_l@ill.fr
 Date: Monday, November 30, 2009, 2:05 PM
 Thank
 you for all of your answers, it was more than I hoped for. 
 
 I did a Rietveld refinement on stibnite (Sb2S3). And I
 mesured it from 5 2theta to 90. But  from 5 to 20 2Θ i
 had a serious background noise and general trend similar to
 that of glass. This trend and the background noise are there
 no matter what I measure. 
 I can't get to the lab until Wednesday, but from what I
 recall, from the X ray source to the X'celerator
 detector: there are 1/4 +1/2 fixed divergence,the widthest
 mask and the slit (i'm not shure if that is the proper
 name) in front of the X'celerator is also the widthest.
 
 The automated system of handling the sample is
 installed, should i changed it? 
 And I forgot to mention, the system is MPD. 
 
 I can give you more details on Wednesday.  
 
 I'm sorry for my not technical terms, but is very hard
 for me
  to speak about diffraction in English. 
 
 
 Alexandra Seclaman
 
 seclaman_alexan...@yahoo.com
 
 a.c.secla...@gmail.com
 
 --- On Sun, 11/29/09, Tony Raftery
 a.raft...@qut.edu.au wrote:
 
 From: Tony Raftery a.raft...@qut.edu.au
 Subject: RE: Configuration of PANalytical X'pert Pro
 To: Alexandra Seclaman
 seclaman_alexan...@yahoo.com
 Date: Sunday, November 29, 2009, 11:57 PM
 
 Alexandra,
 
 while there is no such thing as a standard configuration,
 choices can be made to help in the modelling. Another
 corresponant (to the Rietveld list)  made some
 suggestions that I only partly agree with. A lot depends on
 you goals ans available instrumentation - for instance if
 you have an X'cellerator detector and a monochromator.
 If you are interested in the low angle region (as I think
 you are) or not. If you were looking at some phases or
 phases where the peaks began after
  25 deg 2theta, you could afford to go to larger divergence
 than if the peaks started at 5 deg 2theta.
 
 I see little use for beam knifes except over restricted low
 angle ranges (they always interfere at high angle, where
 they interfere depends on how they are set). It see little
 use for varable divergence slits for all sorts of reasons
 (firstly as there is no rock solid conversion from automatic
 to fixed intensities, secondly the resolution changes with
 angle with variable slits that can't be modelled -
 unless a more sofisticated model than that of Highcore Plus
 is used).
 
 What I agree with is that you need to use and properly
 adjust the anti-scatter slits.
 
 Fro memory (I am away from the lab for a few weeks) my
 conditions are for scans 3.5 deg to 140 deg 2theata - a
 general scan, well crystalled phases
 fixed divergence 0.5 deg, a-scatter (incident) 1 deg,
 sollers (incident) 0.04 rad
 3.4mm (diffracted) a-scatter, mono fot
  X'cellerator, X'cellerator set full )about 2 deg)
 step 0.02 deg (or what X'cellerator allows, in my case
 0.0167 deg), step-time to give about 10,000 counts full
 scale

Re: Configuration of PANalytical X'pert Pro

2009-12-01 Thread Payzant, E. Andrew
Alexandra,

For your PANalytical MPD, at 5 degrees 2theta, a 1/2 deg divergence slit gives 
you a beam length (on the sample) of about 48mm. Probably this is much larger 
than the actual size of your sample!

At 20 degrees 2theta, the length is down to about 12mm for the same divergence 
(1/2 deg).

So if you see that the background is fine above 20 deg, but not below that, 
your problem may be due to unwanted background scatter from the sample holder. 
Also consider that for a 12mm long sample, above 20 degrees 2theta, all the 
beam is illuminating the sample, whereas below this angle, some of the beam 
misses the sample, and by 5 degrees, only a quarter of the beam hits the 
sample. So your peak intensities are all compromised at low angles.

One solution is to reduce the divergence so that the beam does not overflow the 
sample. A variable slit does this for you, but (1) it changes the resolution as 
a function of angle as compared with fixed slits, (2) you need to be careful 
that at high angles you don’t have too large a divergence (you don’t want the 
divergence to become greater than half the anti-scatter slit, otherwise the 
anti-scatter slit will cut off the beam).

A smaller fixed slit will also do the trick, but will of course compromise your 
intensities (and potentially the counting statistics, unless you use a sample 
spinner) at higher angles. If your sample is 12mm in length, then to avoid beam 
overflow at 5 degrees 2theta a divergence slit of 1/8 degree must be used.

I hope this is helpful,

Andrew
--
E. Andrew Payzant
Senior RD Staff Member
High Temperature Materials Laboratory

Oak Ridge National Laboratory
1 Bethel Valley Road
PO Box 2008, MS 6064
Building 4515, Room 113
Oak Ridge, TN, 37831-6064

ph: (865) 574-6538   FAX: (865) 574-3940
web: http://www.ms.ornl.gov/DTP/payzant.shtml




On 12/1/09 9:41 AM, Leonid Solovyov l_solov...@yahoo.com wrote:

Dear Alexandra,

The 1/4 divergence and 1/2 anti-scatter slits are narrow enough to give a 
reasonably low background after 5 degrees 2Theta. First I would check that the 
background is not related to the sample itself. It may be done by measuring an 
empty sample holder.
The background may also be due to scattering from sample holder edges if the 
holder diameter is small and the primary beam mask is wide or/and the Soller 
slit divergence is big.
The automated sample changer can not produce background, but I am not sure if 
the beam knife can be installed together with the changer. The beam knife cuts 
the air scattering which is essential in the low-angle region. The standard 
beam knife supplied by PANal is rather thick and if it is installed too close 
to the sample surface it may attenuate intensity at high angles. With fixed 1/4 
divergence slit the knife should be put about 3-4 mm under the sample surface 
which must be safe for measurements up to 100 degrees 2Theta. To check if the 
knife attenuates high angle reflections it is necessary to measure the same 
sample with and without knife and then compare the intensities.
On my X’Pert I simply replaced the standard knife by a thin (ca. 1 mm) 
stainless steel screen,  and installed it 3 mm above the sample surface which 
allows preserving intensity up to 150 degrees 2Theta even with the 1/2 
divergence primary slit and have a reasonable low-angle background. Of course, 
the knife should also be properly centered on the diffractometer axis and 
aligned vertically.

Best regards,
Leonid

***
Leonid A. Solovyov
Institute of Chemistry and Chemical Technology
660049, K. Marx 42, Krasnoyarsk, Russia
www.icct.ru/eng/content/persons/Sol_LA
sites.google.com/site/solovyovleonid
***

--- On Mon, 11/30/09, Alexandra Seclaman seclaman_alexan...@yahoo.com wrote:

 From: Alexandra Seclaman seclaman_alexan...@yahoo.com
 Subject: RE: Configuration of PANalytical X'pert Pro
 To: rietveld_l@ill.fr
 Date: Monday, November 30, 2009, 2:05 PM
 Thank
 you for all of your answers, it was more than I hoped for.

 I did a Rietveld refinement on stibnite (Sb2S3). And I
 mesured it from 5 2theta to 90. But? from 5 to 20 2Θ i
 had a serious background noise and general trend similar to
 that of glass. This trend and the background noise are there
 no matter what I measure.
 I can't get to the lab until Wednesday, but from what I
 recall, from the X ray source to the X'celerator
 detector: there are 1/4 +1/2 fixed divergence,the widthest
 mask and the slit (i'm not shure if that is the proper
 name) in front of the X'celerator is also the widthest.

 The automated system of handling the sample is
 installed, should i changed it?
 And I forgot to mention, the system is MPD.

 I can give you more details on Wednesday.?

 I'm sorry for my not technical terms, but is very hard
 for me
  to speak about diffraction in English.


 Alexandra Seclaman

 seclaman_alexan...@yahoo.com

 a.c.secla...@gmail.com

RE: Configuration of PANalytical X'pert Pro

2009-11-30 Thread Alexandra Seclaman
Thank you for all of your answers, it was more than I hoped for. 

I did a Rietveld refinement on stibnite (Sb2S3). And I mesured it from 5 2theta 
to 90. But  from 5 to 20 2Θ i had a serious background noise and general trend 
similar to that of glass. This trend and the background noise are there no 
matter what I measure. 
I can't get to the lab until Wednesday, but from what I recall, from the X ray 
source to the X'celerator detector: there are 1/4 +1/2 fixed divergence,the 
widthest mask and the slit (i'm not shure if that is the proper name) in front 
of the X'celerator is also the widthest. 
The automated system of handling the sample is installed, should i changed 
it? 
And I forgot to mention, the system is MPD. 

I can give you more details on Wednesday.  

I'm sorry for my not technical terms, but is very hard for me to speak about 
diffraction in English. 


Alexandra Seclaman

seclaman_alexan...@yahoo.com

a.c.secla...@gmail.com

--- On Sun, 11/29/09, Tony Raftery a.raft...@qut.edu.au wrote:

From: Tony Raftery a.raft...@qut.edu.au
Subject: RE: Configuration of PANalytical X'pert Pro
To: Alexandra Seclaman seclaman_alexan...@yahoo.com
Date: Sunday, November 29, 2009, 11:57 PM

Alexandra,

while there is no such thing as a standard configuration, choices can be made 
to help in the modelling. Another corresponant (to the Rietveld list)  made 
some suggestions that I only partly agree with. A lot depends on you goals ans 
available instrumentation - for instance if you have an X'cellerator detector 
and a monochromator. If you are interested in the low angle region (as I think 
you are) or not. If you were looking at some phases or phases where the peaks 
began after 25 deg 2theta, you could afford to go to larger divergence than if 
the peaks started at 5 deg 2theta.

I see little use for beam knifes except over restricted low angle ranges (they 
always interfere at high angle, where they interfere depends on how they are 
set). It see little use for varable divergence slits for all sorts of reasons 
(firstly as there is no rock solid conversion from automatic to fixed 
intensities, secondly the resolution changes with angle with variable slits 
that can't be modelled - unless a more sofisticated model than that of Highcore 
Plus is used).

What I agree with is that you need to use and properly adjust the anti-scatter 
slits.

Fro memory (I am away from the lab for a few weeks) my conditions are for scans 
3.5 deg to 140 deg 2theata - a general scan, well crystalled phases
fixed divergence 0.5 deg, a-scatter (incident) 1 deg, sollers (incident) 0.04 
rad
3.4mm (diffracted) a-scatter, mono fot X'cellerator, X'cellerator set full 
)about 2 deg)
step 0.02 deg (or what X'cellerator allows, in my case 0.0167 deg), step-time 
to give about 10,000 counts full scale

From: Alexandra Seclaman [seclaman_alexan...@yahoo.com]
Sent: Saturday, 28 November 2009 3:09 AM
To: rietveld_l@ill.fr
Subject: Configuration of PANalytical X'pert Pro

Hello,

I have done a Rietveld refinement using a PANalytical X'pert Pro but the 
standard configuration is giving me a lot of troubles. This configuration is 
somewhat blind to the small 2 theta angles and it introduces (because of the 
poorly focused X ray beam) a heavy background. I've manage to remove that 
background by refinement but I'm sure that the quality of the acquisition can 
be improved.
I am no engineer, so the inner workings of that machine are somewhat a mystery 
to me. I am asking if someone else has used a PANalytical for Rietveld and if 
you can give me a better configuration of slots, masks etc. I have tried 
different configurations with no success.

Please take into consideration, while writing your reply, that I'm still a 
student and I have yet a lot to learn.

Thank you!

Alexandra Seclaman
seclaman_alexan...@yahoo.com
a.c.secla...@gmail.com




  

Configuration of PANalytical X'pert Pro

2009-11-27 Thread Alexandra Seclaman
Hello, 





I have done a Rietveld refinement using a PANalytical X'pert
Pro but the standard configuration is giving me a lot of troubles. This
configuration is somewhat blind to the small 2 theta angles and it
introduces (because of the poorly focused X ray beam) a heavy
background. I've manage to remove that background by refinement but I'm
sure that the quality of the acquisition can be improved.  

I am no engineer, so the inner workings of that machine are somewhat a
mystery to me. I am asking if someone else has used a PANalytical for
Rietveld and if you can give me a better configuration of slots, masks
etc. I have tried different configurations with no success.




Please take into consideration, while writing your reply, that I'm still a 
student and I have yet a lot to learn. 





Thank you!


Alexandra Seclaman

seclaman_alexan...@yahoo.com

a.c.secla...@gmail.com


  

Re: Configuration of PANalytical X'pert Pro

2009-11-27 Thread Leonid Solovyov
Dear Alexandra,

The term “standard configuration” is rather indefinite since PANal has a lot of 
various configurations and accessories. The most straightforward option for 
reducing the low-angle background is the programmable divergence and 
anti-scatter slits. For a fixed slit system the low-angle background may be 
reduced by primary slits and mask of lower divergence and a proper adjustment 
of the beam knife (above the sample) that cuts the air scattering. The 
alignment of the sample holder in the primary beam is also important for 
low-angle measurements.

Best regards,
Leonid

***
Leonid A. Solovyov
Institute of Chemistry and Chemical Technology
660049, K. Marx 42, Krasnoyarsk, Russia
www.icct.ru/eng/content/persons/Sol_LA
sites.google.com/site/solovyovleonid
***

--- On Fri, 11/27/09, Alexandra Seclaman seclaman_alexan...@yahoo.com wrote:

 From: Alexandra Seclaman seclaman_alexan...@yahoo.com
 Subject: Configuration of PANalytical X'pert Pro
 To: rietveld_l@ill.fr rietveld_l@ill.fr
 Date: Friday, November 27, 2009, 5:09 PM
 Hello,
 
 
 
 
 
 
 I have done a Rietveld refinement using a PANalytical
 X'pert
 Pro but the standard configuration is giving me a lot of
 troubles. This
 configuration is somewhat blind to the small 2
 theta angles and it
 introduces (because of the poorly focused X ray beam) a
 heavy
 background. I've manage to remove that background by
 refinement but I'm
 sure that the quality of the acquisition can be
 improved.  
 
 I am no engineer, so the inner workings of that machine are
 somewhat a
 mystery to me. I am asking if someone else has used a
 PANalytical for
 Rietveld and if you can give me a better configuration of
 slots, masks
 etc. I have tried different configurations with no
 success.
 
 
 
 
 Please take into consideration, while writing your reply,
 that I'm still a student and I have yet a lot to learn. 
 
 
 
 
 
 Thank you!

 
 
 Alexandra Seclaman
 
 seclaman_alexan...@yahoo.com
 