Re: Configuration of PANalytical X'pert Pro
Dear Alexandra and Leonid, I do routinely small angle measurements starting from 0.6 degrees 2theta (Cu radiation) with the 1/32 divergence and 1/16 anti-scatter slit without beam knife, but never with the standard Panalytical stainless steel sample holder. It appears that this one gives a high background at very low angles. So you better use some polymer or glass sample holder. This works fine. In addition you should be sure that your sample stage is well aligned (height and theta/2theta coupling). I know of a similar Xpert machine, where low-angle measurements were up until recently not possible, just because of improper alignment. best regards, Arie Leonid Solovyov wrote the following on 01/12/2009 15:41: Dear Alexandra, The 1/4 divergence and 1/2 anti-scatter slits are narrow enough to give a reasonably low background after 5 degrees 2Theta. First I would check that the background is not related to the sample itself. It may be done by measuring an empty sample holder. The background may also be due to scattering from sample holder edges if the holder diameter is small and the primary beam mask is wide or/and the Soller slit divergence is big. The automated sample changer can not produce background, but I am not sure if the beam knife can be installed together with the changer. The beam knife cuts the air scattering which is essential in the low-angle region. The standard beam knife supplied by PANal is rather thick and if it is installed too close to the sample surface it may attenuate intensity at high angles. With fixed 1/4 divergence slit the knife should be put about 3-4 mm under the sample surface which must be safe for measurements up to 100 degrees 2Theta. To check if the knife attenuates high angle reflections it is necessary to measure the same sample with and without knife and then compare the intensities. On my X’Pert I simply replaced the standard knife by a thin (ca. 1 mm) stainless steel screen, and installed it 3 mm above the sample surface which allows preserving intensity up to 150 degrees 2Theta even with the 1/2 divergence primary slit and have a reasonable low-angle background. Of course, the knife should also be properly centered on the diffractometer axis and aligned vertically. Best regards, Leonid *** Leonid A. Solovyov Institute of Chemistry and Chemical Technology 660049, K. Marx 42, Krasnoyarsk, Russia www.icct.ru/eng/content/persons/Sol_LA sites.google.com/site/solovyovleonid *** --- On Mon, 11/30/09, Alexandra Seclaman seclaman_alexan...@yahoo.com wrote: From: Alexandra Seclaman seclaman_alexan...@yahoo.com Subject: RE: Configuration of PANalytical X'pert Pro To: rietveld_l@ill.fr Date: Monday, November 30, 2009, 2:05 PM Thank you for all of your answers, it was more than I hoped for. I did a Rietveld refinement on stibnite (Sb2S3). And I mesured it from 5 2theta to 90. But from 5 to 20 2Θ i had a serious background noise and general trend similar to that of glass. This trend and the background noise are there no matter what I measure. I can't get to the lab until Wednesday, but from what I recall, from the X ray source to the X'celerator detector: there are 1/4 +1/2 fixed divergence,the widthest mask and the slit (i'm not shure if that is the proper name) in front of the X'celerator is also the widthest. The automated system of handling the sample is installed, should i changed it? And I forgot to mention, the system is MPD. I can give you more details on Wednesday. I'm sorry for my not technical terms, but is very hard for me to speak about diffraction in English. Alexandra Seclaman seclaman_alexan...@yahoo.com a.c.secla...@gmail.com --- On Sun, 11/29/09, Tony Raftery a.raft...@qut.edu.au wrote: From: Tony Raftery a.raft...@qut.edu.au Subject: RE: Configuration of PANalytical X'pert Pro To: Alexandra Seclaman seclaman_alexan...@yahoo.com Date: Sunday, November 29, 2009, 11:57 PM Alexandra, while there is no such thing as a standard configuration, choices can be made to help in the modelling. Another corresponant (to the Rietveld list) made some suggestions that I only partly agree with. A lot depends on you goals ans available instrumentation - for instance if you have an X'cellerator detector and a monochromator. If you are interested in the low angle region (as I think you are) or not. If you were looking at some phases or phases where the peaks began after 25 deg 2theta, you could afford to go to larger divergence than if the peaks started at 5 deg 2theta. I see little use for beam knifes except over restricted low angle ranges (they always interfere at high angle, where they interfere depends on how they are set). It see little use for varable divergence slits for all sorts of reasons (firstly as there is no rock solid conversion from automatic to fixed intensities, secondly the resolution changes
RE: Configuration of PANalytical X'pert Pro
Dear Alexandra, The 1/4 divergence and 1/2 anti-scatter slits are narrow enough to give a reasonably low background after 5 degrees 2Theta. First I would check that the background is not related to the sample itself. It may be done by measuring an empty sample holder. The background may also be due to scattering from sample holder edges if the holder diameter is small and the primary beam mask is wide or/and the Soller slit divergence is big. The automated sample changer can not produce background, but I am not sure if the beam knife can be installed together with the changer. The beam knife cuts the air scattering which is essential in the low-angle region. The standard beam knife supplied by PANal is rather thick and if it is installed too close to the sample surface it may attenuate intensity at high angles. With fixed 1/4 divergence slit the knife should be put about 3-4 mm under the sample surface which must be safe for measurements up to 100 degrees 2Theta. To check if the knife attenuates high angle reflections it is necessary to measure the same sample with and without knife and then compare the intensities. On my X’Pert I simply replaced the standard knife by a thin (ca. 1 mm) stainless steel screen, and installed it 3 mm above the sample surface which allows preserving intensity up to 150 degrees 2Theta even with the 1/2 divergence primary slit and have a reasonable low-angle background. Of course, the knife should also be properly centered on the diffractometer axis and aligned vertically. Best regards, Leonid *** Leonid A. Solovyov Institute of Chemistry and Chemical Technology 660049, K. Marx 42, Krasnoyarsk, Russia www.icct.ru/eng/content/persons/Sol_LA sites.google.com/site/solovyovleonid *** --- On Mon, 11/30/09, Alexandra Seclaman seclaman_alexan...@yahoo.com wrote: From: Alexandra Seclaman seclaman_alexan...@yahoo.com Subject: RE: Configuration of PANalytical X'pert Pro To: rietveld_l@ill.fr Date: Monday, November 30, 2009, 2:05 PM Thank you for all of your answers, it was more than I hoped for. I did a Rietveld refinement on stibnite (Sb2S3). And I mesured it from 5 2theta to 90. But from 5 to 20 2Θ i had a serious background noise and general trend similar to that of glass. This trend and the background noise are there no matter what I measure. I can't get to the lab until Wednesday, but from what I recall, from the X ray source to the X'celerator detector: there are 1/4 +1/2 fixed divergence,the widthest mask and the slit (i'm not shure if that is the proper name) in front of the X'celerator is also the widthest. The automated system of handling the sample is installed, should i changed it? And I forgot to mention, the system is MPD. I can give you more details on Wednesday. I'm sorry for my not technical terms, but is very hard for me to speak about diffraction in English. Alexandra Seclaman seclaman_alexan...@yahoo.com a.c.secla...@gmail.com --- On Sun, 11/29/09, Tony Raftery a.raft...@qut.edu.au wrote: From: Tony Raftery a.raft...@qut.edu.au Subject: RE: Configuration of PANalytical X'pert Pro To: Alexandra Seclaman seclaman_alexan...@yahoo.com Date: Sunday, November 29, 2009, 11:57 PM Alexandra, while there is no such thing as a standard configuration, choices can be made to help in the modelling. Another corresponant (to the Rietveld list) made some suggestions that I only partly agree with. A lot depends on you goals ans available instrumentation - for instance if you have an X'cellerator detector and a monochromator. If you are interested in the low angle region (as I think you are) or not. If you were looking at some phases or phases where the peaks began after 25 deg 2theta, you could afford to go to larger divergence than if the peaks started at 5 deg 2theta. I see little use for beam knifes except over restricted low angle ranges (they always interfere at high angle, where they interfere depends on how they are set). It see little use for varable divergence slits for all sorts of reasons (firstly as there is no rock solid conversion from automatic to fixed intensities, secondly the resolution changes with angle with variable slits that can't be modelled - unless a more sofisticated model than that of Highcore Plus is used). What I agree with is that you need to use and properly adjust the anti-scatter slits. Fro memory (I am away from the lab for a few weeks) my conditions are for scans 3.5 deg to 140 deg 2theata - a general scan, well crystalled phases fixed divergence 0.5 deg, a-scatter (incident) 1 deg, sollers (incident) 0.04 rad 3.4mm (diffracted) a-scatter, mono fot X'cellerator, X'cellerator set full )about 2 deg) step 0.02 deg (or what X'cellerator allows, in my case 0.0167 deg), step-time to give about 10,000 counts full scale
Re: Configuration of PANalytical X'pert Pro
Alexandra, For your PANalytical MPD, at 5 degrees 2theta, a 1/2 deg divergence slit gives you a beam length (on the sample) of about 48mm. Probably this is much larger than the actual size of your sample! At 20 degrees 2theta, the length is down to about 12mm for the same divergence (1/2 deg). So if you see that the background is fine above 20 deg, but not below that, your problem may be due to unwanted background scatter from the sample holder. Also consider that for a 12mm long sample, above 20 degrees 2theta, all the beam is illuminating the sample, whereas below this angle, some of the beam misses the sample, and by 5 degrees, only a quarter of the beam hits the sample. So your peak intensities are all compromised at low angles. One solution is to reduce the divergence so that the beam does not overflow the sample. A variable slit does this for you, but (1) it changes the resolution as a function of angle as compared with fixed slits, (2) you need to be careful that at high angles you don’t have too large a divergence (you don’t want the divergence to become greater than half the anti-scatter slit, otherwise the anti-scatter slit will cut off the beam). A smaller fixed slit will also do the trick, but will of course compromise your intensities (and potentially the counting statistics, unless you use a sample spinner) at higher angles. If your sample is 12mm in length, then to avoid beam overflow at 5 degrees 2theta a divergence slit of 1/8 degree must be used. I hope this is helpful, Andrew -- E. Andrew Payzant Senior RD Staff Member High Temperature Materials Laboratory Oak Ridge National Laboratory 1 Bethel Valley Road PO Box 2008, MS 6064 Building 4515, Room 113 Oak Ridge, TN, 37831-6064 ph: (865) 574-6538 FAX: (865) 574-3940 web: http://www.ms.ornl.gov/DTP/payzant.shtml On 12/1/09 9:41 AM, Leonid Solovyov l_solov...@yahoo.com wrote: Dear Alexandra, The 1/4 divergence and 1/2 anti-scatter slits are narrow enough to give a reasonably low background after 5 degrees 2Theta. First I would check that the background is not related to the sample itself. It may be done by measuring an empty sample holder. The background may also be due to scattering from sample holder edges if the holder diameter is small and the primary beam mask is wide or/and the Soller slit divergence is big. The automated sample changer can not produce background, but I am not sure if the beam knife can be installed together with the changer. The beam knife cuts the air scattering which is essential in the low-angle region. The standard beam knife supplied by PANal is rather thick and if it is installed too close to the sample surface it may attenuate intensity at high angles. With fixed 1/4 divergence slit the knife should be put about 3-4 mm under the sample surface which must be safe for measurements up to 100 degrees 2Theta. To check if the knife attenuates high angle reflections it is necessary to measure the same sample with and without knife and then compare the intensities. On my X’Pert I simply replaced the standard knife by a thin (ca. 1 mm) stainless steel screen, and installed it 3 mm above the sample surface which allows preserving intensity up to 150 degrees 2Theta even with the 1/2 divergence primary slit and have a reasonable low-angle background. Of course, the knife should also be properly centered on the diffractometer axis and aligned vertically. Best regards, Leonid *** Leonid A. Solovyov Institute of Chemistry and Chemical Technology 660049, K. Marx 42, Krasnoyarsk, Russia www.icct.ru/eng/content/persons/Sol_LA sites.google.com/site/solovyovleonid *** --- On Mon, 11/30/09, Alexandra Seclaman seclaman_alexan...@yahoo.com wrote: From: Alexandra Seclaman seclaman_alexan...@yahoo.com Subject: RE: Configuration of PANalytical X'pert Pro To: rietveld_l@ill.fr Date: Monday, November 30, 2009, 2:05 PM Thank you for all of your answers, it was more than I hoped for. I did a Rietveld refinement on stibnite (Sb2S3). And I mesured it from 5 2theta to 90. But? from 5 to 20 2Θ i had a serious background noise and general trend similar to that of glass. This trend and the background noise are there no matter what I measure. I can't get to the lab until Wednesday, but from what I recall, from the X ray source to the X'celerator detector: there are 1/4 +1/2 fixed divergence,the widthest mask and the slit (i'm not shure if that is the proper name) in front of the X'celerator is also the widthest. The automated system of handling the sample is installed, should i changed it? And I forgot to mention, the system is MPD. I can give you more details on Wednesday.? I'm sorry for my not technical terms, but is very hard for me to speak about diffraction in English. Alexandra Seclaman seclaman_alexan...@yahoo.com a.c.secla...@gmail.com
RE: Configuration of PANalytical X'pert Pro
Thank you for all of your answers, it was more than I hoped for. I did a Rietveld refinement on stibnite (Sb2S3). And I mesured it from 5 2theta to 90. But from 5 to 20 2Θ i had a serious background noise and general trend similar to that of glass. This trend and the background noise are there no matter what I measure. I can't get to the lab until Wednesday, but from what I recall, from the X ray source to the X'celerator detector: there are 1/4 +1/2 fixed divergence,the widthest mask and the slit (i'm not shure if that is the proper name) in front of the X'celerator is also the widthest. The automated system of handling the sample is installed, should i changed it? And I forgot to mention, the system is MPD. I can give you more details on Wednesday. I'm sorry for my not technical terms, but is very hard for me to speak about diffraction in English. Alexandra Seclaman seclaman_alexan...@yahoo.com a.c.secla...@gmail.com --- On Sun, 11/29/09, Tony Raftery a.raft...@qut.edu.au wrote: From: Tony Raftery a.raft...@qut.edu.au Subject: RE: Configuration of PANalytical X'pert Pro To: Alexandra Seclaman seclaman_alexan...@yahoo.com Date: Sunday, November 29, 2009, 11:57 PM Alexandra, while there is no such thing as a standard configuration, choices can be made to help in the modelling. Another corresponant (to the Rietveld list) made some suggestions that I only partly agree with. A lot depends on you goals ans available instrumentation - for instance if you have an X'cellerator detector and a monochromator. If you are interested in the low angle region (as I think you are) or not. If you were looking at some phases or phases where the peaks began after 25 deg 2theta, you could afford to go to larger divergence than if the peaks started at 5 deg 2theta. I see little use for beam knifes except over restricted low angle ranges (they always interfere at high angle, where they interfere depends on how they are set). It see little use for varable divergence slits for all sorts of reasons (firstly as there is no rock solid conversion from automatic to fixed intensities, secondly the resolution changes with angle with variable slits that can't be modelled - unless a more sofisticated model than that of Highcore Plus is used). What I agree with is that you need to use and properly adjust the anti-scatter slits. Fro memory (I am away from the lab for a few weeks) my conditions are for scans 3.5 deg to 140 deg 2theata - a general scan, well crystalled phases fixed divergence 0.5 deg, a-scatter (incident) 1 deg, sollers (incident) 0.04 rad 3.4mm (diffracted) a-scatter, mono fot X'cellerator, X'cellerator set full )about 2 deg) step 0.02 deg (or what X'cellerator allows, in my case 0.0167 deg), step-time to give about 10,000 counts full scale From: Alexandra Seclaman [seclaman_alexan...@yahoo.com] Sent: Saturday, 28 November 2009 3:09 AM To: rietveld_l@ill.fr Subject: Configuration of PANalytical X'pert Pro Hello, I have done a Rietveld refinement using a PANalytical X'pert Pro but the standard configuration is giving me a lot of troubles. This configuration is somewhat blind to the small 2 theta angles and it introduces (because of the poorly focused X ray beam) a heavy background. I've manage to remove that background by refinement but I'm sure that the quality of the acquisition can be improved. I am no engineer, so the inner workings of that machine are somewhat a mystery to me. I am asking if someone else has used a PANalytical for Rietveld and if you can give me a better configuration of slots, masks etc. I have tried different configurations with no success. Please take into consideration, while writing your reply, that I'm still a student and I have yet a lot to learn. Thank you! Alexandra Seclaman seclaman_alexan...@yahoo.com a.c.secla...@gmail.com
Configuration of PANalytical X'pert Pro
Hello, I have done a Rietveld refinement using a PANalytical X'pert Pro but the standard configuration is giving me a lot of troubles. This configuration is somewhat blind to the small 2 theta angles and it introduces (because of the poorly focused X ray beam) a heavy background. I've manage to remove that background by refinement but I'm sure that the quality of the acquisition can be improved. I am no engineer, so the inner workings of that machine are somewhat a mystery to me. I am asking if someone else has used a PANalytical for Rietveld and if you can give me a better configuration of slots, masks etc. I have tried different configurations with no success. Please take into consideration, while writing your reply, that I'm still a student and I have yet a lot to learn. Thank you! Alexandra Seclaman seclaman_alexan...@yahoo.com a.c.secla...@gmail.com
Re: Configuration of PANalytical X'pert Pro
Dear Alexandra, The term “standard configuration” is rather indefinite since PANal has a lot of various configurations and accessories. The most straightforward option for reducing the low-angle background is the programmable divergence and anti-scatter slits. For a fixed slit system the low-angle background may be reduced by primary slits and mask of lower divergence and a proper adjustment of the beam knife (above the sample) that cuts the air scattering. The alignment of the sample holder in the primary beam is also important for low-angle measurements. Best regards, Leonid *** Leonid A. Solovyov Institute of Chemistry and Chemical Technology 660049, K. Marx 42, Krasnoyarsk, Russia www.icct.ru/eng/content/persons/Sol_LA sites.google.com/site/solovyovleonid *** --- On Fri, 11/27/09, Alexandra Seclaman seclaman_alexan...@yahoo.com wrote: From: Alexandra Seclaman seclaman_alexan...@yahoo.com Subject: Configuration of PANalytical X'pert Pro To: rietveld_l@ill.fr rietveld_l@ill.fr Date: Friday, November 27, 2009, 5:09 PM Hello, I have done a Rietveld refinement using a PANalytical X'pert Pro but the standard configuration is giving me a lot of troubles. This configuration is somewhat blind to the small 2 theta angles and it introduces (because of the poorly focused X ray beam) a heavy background. I've manage to remove that background by refinement but I'm sure that the quality of the acquisition can be improved. I am no engineer, so the inner workings of that machine are somewhat a mystery to me. I am asking if someone else has used a PANalytical for Rietveld and if you can give me a better configuration of slots, masks etc. I have tried different configurations with no success. Please take into consideration, while writing your reply, that I'm still a student and I have yet a lot to learn. Thank you! Alexandra Seclaman seclaman_alexan...@yahoo.com a.c.secla...@gmail.com