Dear Jose,

I used both instruments for a number of years, first the VP ITC and later on the ITC 200. I personally find the ITC 200 much more demanding to operate. I think Microcal managed to improve the sensitivity of the ITC 200 about 2-3 fold compared to the VP ITC, but they decided to decrease the volumes by a factor of 6-7. This means that you have to work at higher sample concentrations in the cell and in the syringe to get decent signals. In addition the instrument is rather fragile (especially the glass syringe) and more difficult to clean and maintain. I decided to go back to the VP ITC. I cannot comment on other instruments.

best wishes
Michael


On 01/17/2012 05:02 PM, Jose Artur Brito wrote:
Dear All,
sorry for this off-topic questions but I would like to have some feed-back from you on Isothermal Titration Calorimetry (ITC) equipments. We have a very nice quotation for an iTC200 from GE Healthcare. We wanted this one because it uses ~200uL sample per measurement (really nice when your dealing with "precious" samples, ie., proteins with low expression yields). However, I was told that, although consuming much less sample, is not as good (sensitivity, mixing issues, bubbles, ...), as the VP-ITC (it uses ~1.4mL per measurement, seven times more than the iTC200). Does anyone has experience with these two equipments? Would you prefer one over the other (please state your reasons)? Would you suggest another equipment/brand for the ITC (like the NanoITC from TA Instruments)?
Thanks in advance,
Jose Artur Brito