-----BEGIN PGP SIGNED MESSAGE-----
Hash: SHA1

Hi Ed,


> only prepared one sample, so if on that particular instance I
> picked up 4.8ul and not 5.0ul, this will translate into
> systematically

I don't share your opinion about a single measurement translating into
a systematic error. I would call it a poorly designed experiment in
case you were actually iterested in how accurately you determined the
protein concentration.

Best,
Tim


On 03/11/2013 04:46 PM, Ed Pozharski wrote:
> Salve,
> 
> I would like to solicit opinions on a certain question about the 
> relationship between statistical and systematic error. Please read
> and consider the following in its entirety before commenting.
> 
> Statistical error (experiment precision) is determined by the
> degree to which experimental measurement is reproducible. It is
> derived from variance of the data when an experiment is repeated
> multiple times under otherwise identical conditions. Statistical
> error is by its very nature irremovable and originates from various
> sources of random noise, which can be reduced but not entirely
> eliminated.
> 
> Systematic error (experiment accuracy) reflects degree to which
> precise average deviates from a true value. Theoretically,
> corrections can be introduced to the experimental method that
> eliminate various sources of bias. Systematic error refers to some
> disconnect between the quantities one tries to determine and what
> is actually measured.
> 
> The issue is whether the classification of various sources of error
> into the two types depends on procedure. Let me explain using an
> example.
> 
> To determine the concentration of a protein stock, I derive
> extinction coefficient from its sequence, dilute it 20x to and take
> OD measurement. The OD value is then divided by extinction
> coefficient and inflated 20 times to calculate concentration.
> 
> So what is the statistical error of this when I am at the 
> spectrophotometer? I can cycle sample cuvette in and out of the
> holder to correct for reproducibility of its position and
> instrument noise. This gives me the estimated statistical error of
> the OD measurement. Scaled by extinction coefficient and dilution
> factor, this number corresponds to the statistical error
> (precision) of the protein concentration.
> 
> There are two sources of the systematic error originating from the
> two factors used to convert OD to concentration. First is
> irremovable inaccuracy of the extinction coefficient.
> 
> Second: dilution factor. Here main contribution to the systematic
> error is pipetting. Importantly, this includes both systematic
> (pipettor calibration) and statistical (pipetting precision) error.
> Notice that I only prepared one sample, so if on that particular
> instance I picked up 4.8ul and not 5.0ul, this will translate into
> systematically underestimating protein concentration, even though
> it could have equally likely been 5.2ul.
> 
> So if pipetting error could have contributed ~4% into the overall 
> systematic error while the spectrophotometer measures with 0.1% 
> precision, it makes sense to consider how this systematic error can
> be eliminated. The experiment can be modified to include multiple
> samples prepared for OD determination from the same protein stock.
> 
> An interesting thing happens when I do that. What used to be a 
> systematic error of pipetting now becomes statistical error,
> because my experiment now includes reproducing dilution of the
> stock. In a nutshell,
> 
> Whether a particular source of error contributes to accuracy or 
> precision of an experiment depends on how experiment is conducted.
> 
> 
> And one more thing. No need to waste precious protein on
> evaluating error of pipetting. I can determine that from a separate
> calibration experiment using lysozyme solution of comparable
> concentration/surface tension. Technically, a single measurement
> has accuracy of said 4% (padded by whatever is error in extinction
> coefficient). But one can also project that with actual dilution
> repeats, the precision would be this same 4% (assuming that this is
> a dominant source of error).
> 
> So, is there anything wrong with this? Naturally, the question
> really is not about extinction coefficients, but rather about
> semantics of what is accuracy and what is precision and whether
> certain source of experimental error is rigidly assigned to one of
> the two categories. There is, of course, the wikipedia article on
> accuracy vs precision, and section 3.1 from Ian's paper (ActaD
> 68:454) can be used as a point of reference.
> 
> Cheers,
> 
> Ed.
> 

- -- 
- --
Dr Tim Gruene
Institut fuer anorganische Chemie
Tammannstr. 4
D-37077 Goettingen

GPG Key ID = A46BEE1A

-----BEGIN PGP SIGNATURE-----
Version: GnuPG v1.4.12 (GNU/Linux)
Comment: Using GnuPG with Mozilla - http://enigmail.mozdev.org/

iD8DBQFRPhm5UxlJ7aRr7hoRAuoSAJwN9zAJj2qbZBNMlF0cJ0goszaqWQCg2hFp
9u+slrVyYEYbCf2D2/SOVTg=
=UACi
-----END PGP SIGNATURE-----

Reply via email to