Dear Markus

Please find a response from one of our specialists - feel free to contact him 
for more information (his contact details are at the bottom of this email as he 
is not on the bulletin board).

The primary factors I would consider are the type of material in the column and 
also the pore size/ effective molecular weight range. The pore size will 
dictate the selectivity of the separation, you mention you are looking at large 
proteins (MDa) so a column with larger pores will give you a higher effective 
molecular weight range and better resolution. The molecular weight range for 
each column is published by each manufacturer, I have listed these below: 

SEPAX UNIX-C 5,000-1,250,000 (https://sepax-tech.com/Unix-C.php) 
BioZen SEC-3 10,000-700,000 (http://www.phenomenex.com/products/detail/biozen) 
Waters BEH450 or XBridge450  SEC 100,000-1,500,000 
(http://www.waters.com/waters/partDetail.htm?partNumber=186006852) 

The SEPAX and BioZen are both silica materials so high efficiency and good 
pressure tolerance but also with the potential for proteins to stick requiring 
high salt. The Xbridge and BEH are hybrid particles (hybrid of a silica and 
polymer) giving reduced secondary interactions but retaining the high 
efficiency and pressure tolerance of silica. so may be a nice middle ground 
between silica and polymer.   

I was also wondering if you working with a HPLC system or a UPLC system? The 
extra column volume has a big impact on resolution of peaks in SEC if you are 
working on HPLC 7.8mm i.d. column will help to absorb some of the system 
dispersion, if your system is a well optimised UPLC (1290, ACQUITY etc) 4.6mm 
will be fine. 

Richard Robinson 
Biopharm Specialist
UK and Ireland  

[T] +442082386100   
[M] +447824326822   
[W] www.waters.com 
[E] [email protected] 


-----Original Message-----
From: CCP4 bulletin board <[email protected]> On Behalf Of Markus Heckmann
Sent: 26 April 2018 17:04
To: [email protected]
Subject: [ccp4bb] size exclusion columns

Dear all,

We are looking for a size exclusion chromatography column
(silica-based) for protein purification prior to a MALS-detector. We looked for 
(www.waters.com BEH-450), Sepax (Unix-C 300) and Phenomex (BioZen SEC-3).  Any 
'column' tips or recommendations when dealing with large proteins (MDa)?

Many thanks
Markus