Mr Jenson - I agree with Bill's input. When I do in house validation I will
usually go to 10kV (or K) - 12kV for a test level. The standard we use is EN
60601-1-2 (medical electrical equipment) which states 3kV contact discharge
and 8kV air discharge. These values are way too conservative for our
Minnesota winters, when the air is dry and one can easily generate voltages
up into the 10k - 12k range. I've witnessed many of our older products that
would pass the requirements for the DoC but would fail once placed in the
field if we followed the standards test level requirements. I highly
recommend to test to as many "user contact" test points as deemed
appropriate, use higher test levels than the standards specify and if
questionable results - test at more than 10 discharges per test point.
Assure that all interface cables are connected to the EUT and opposite end
terminated if required. Cables with a marginal shielding design can have a
big impact on the results. Also assure that you perform ESD discharges to
the horizontal and vertical planes. I have seen devices pass the contact
discharge ESD events but fail when performing the vertical coupling plane
tests. 
Good Luck! 

-----Original Message-----
From: Jacowleff, Bill [mailto:[email protected]]
Sent: Monday, January 17, 2000 7:43 AM
To: '[email protected]'; [email protected]
Subject: RE: EN 61000-4-2



Dear Mr. Kim Boll Jensen:

Well the answer to your question is 10 discharges to each point per
polarity.  A few other things to consider. 

It is also important to step through the voltages and not skip any.

Is your internal test procedure going to be used as a supporting test for a
DOC or is it for a DV or PV?  If it is a Design validation it is quite
possible to want to exceed the specs per the standard, it all depends on
what you expect to learn from this type of testing.   The Design Validations
I have composed for ESD have exceeded the standard specifications.

Best Regards,

Bill Jacowleff
VDO Control Systems
Airpax Instruments
150 Knotter Drive
Cheshire, CT 06410
Phone: 203 271-6394
FAX:    203 271-6200
[email protected]

 

-----Original Message-----
From: [email protected] [mailto:[email protected]]
Sent: Monday, January 17, 2000 5:41 AM
To: [email protected]
Subject: EN 61000-4-2



Dear all

I'm writing on a internal testprocedure for our products, and have now come
to
EN 61000-4-2 ESD. What confuse me is that the standard (as I read it) only
refere to 10 discharges at each point (and each polarity), but I have been
told
from several different persons that the right number is 50 at each points
(but
without any reference to any standard).

Do anynoe know what the right number is and where I can find it

Best regards,

Mr. Kim Boll Jensen
i-data international
Denmark



---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).


---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).


---------
This message is coming from the emc-pstc discussion list.
To cancel your subscription, send mail to [email protected]
with the single line: "unsubscribe emc-pstc" (without the
quotes).  For help, send mail to [email protected],
[email protected], [email protected], or
[email protected] (the list administrators).

Reply via email to