Hi Gordon (and John)

This is a good example of a recent standard created by
IEC with stupid mistakes in it and voted by CENELEC without thinking.
While for conducted immunity the frequency boundaries are
defined (6.2 4th par) , for radiated immunity they are not.
One may argue that the this way one should comply with a defined range
in EN 61000-4-3 (most recentversion : as undated in table 1)
and automatically cover extended frequency ranges as the standard updates,
but
EN 6100-4-3 is not harmonized (basic standard)!
It's the product standard (so EN 61326) that should list frequency ranges,
and
corresponding limits. Limits without frequencies are without meaning.

(By absence of defined boundaries one should fall back to the Directives
text, and the EMC directive covers the range 0 - 400 Ghz.
So now we should take the measurement set up from 6100-4-3
and measure up to 400 GHz.  Ridiculous)

This standard was voted upon and no-one noticed the omission.
I must admit that it has passed my desk, without me noticing either.

I suggest that you use your common sense, and perform the
test as defined in the EN 50082-1:1997 (generic standard; harmonized)

UNLESS you have the feeling that your product might fail in
proximity of DEC and GSM signal at 1800 Mhz or other
suitable signal sources above 1 Gig.
Due diligence: you know ?

Another stupid point:

While stating that the product must remain safe as a result of
applying the standard, it does not require that it is safe
during the application of the test. (page 23)

In additon to that a new environment has been defined (page 35)
Controlled EM environment.
Stupid as  i am I was thinking that it means that some shielding
might be present. No, it means that cellulars are prohibited.
Who knows such a test lab ? And who would comply !
Several limits have been relaxed because engineers are used to
interpret deviant behavior of equipment when using the cellular !
No word about safety: engineers are used to dangerously operating
test equipment and are resistant to electrical shocks and other
hazards. Table B1 is allowed for any laboratory, including f.a. biomedical
where no-one has read the electrical code...!
To my so humble opinion, this table is there just as an
escape route for non compliant equipment.
Who cares about immunity anyway !

Gert Gremmen
ce-test


-----Original Message-----
From: [email protected]
[mailto:[email protected]]On Behalf Of Gordon,Ian
Sent: Friday, September 27, 2002 1:47 PM
To: 'IEEE EMC & SAFETY PSTC'
Subject: RADIATED IMMUNITY TESTING ABOVE 1GHz



Everybody
Does anyone know whether it is necessary to perform radiated immunity tests
above 1GHz when applying BS EN61326 which is the product standard titled
"Electrical equipment for measurement, control and laboratory use - EMC
requirements"? The products concerned are not telecom or IT devices.
The standard refers to IEC 61000-4-3 for this testing which includes tests
above 1GHz at various mobile telephone frequencies.  I believe the generic
standard EN61000-6-2 refers to testing using pulsed modulation at 900MHz
only.

Thanks
Ian Gordon

_____________________________________________________________________
This e-mail has been scanned for viruses by the WorldCom Internet Managed
Scanning Service - powered by MessageLabs. For further information visit
http://www.worldcom.com

-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://ieeepstc.mindcruiser.com/
    Click on "browse" and then "emc-pstc mailing list"




-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

All emc-pstc postings are archived and searchable on the web at:
    http://ieeepstc.mindcruiser.com/
    Click on "browse" and then "emc-pstc mailing list"

Reply via email to