It's not only an issue of intended use, but also a question of location. With
regard to the example 
you give, if a PC for the surgeon's urgent emails is located within the
'patient vicinity', as 
defined in IEC 60601-1-1 (within about 6 feet of the patient's bed), then I
believe that it should 
comply with IEC 60601-1, even though it's not a medical device.

The reasoning behind this is that the patient could possibly touch the PC
(that's the logic behind 
the 'patient vicinity' concept), and close a leakage current loop, together
with other medical 
instruments with which he is in contact. This could be particularly nasty if
there is a fault in 
the PC's earth wire.

Regards,

Jon Griver
http://www.601help.com
The Medical Device Developers Guide to IEC 60601-1


On 20 Feb 2003 at 12:38, Han, Delphina wrote:

> 
> Hi all,
> 
> Thanks for your responses. When I originally asked the question, I was sure
> that the particular equipment I was asking about does not fall under the
> scope IEC60601, but I wanted to confirm if IEC61010 is the relevant
> standard. As you may notice, I have a copy of the IEC60601 standard but not
> IEC61010! I also needed to find out if there were special requirements for
> isolation transformers in IEC61010. 
> 
> >From the discussion, it seems that which standard to comply with is
> dependent on the intended use, and not where the equipment is placed. So, if
> we put a PC in the hospital operating room, but it is just for the surgeon
> to let's say, check email (not practical...but this is just an example),
> would 60950 be the relevant standard? I doubt it'd have to meet 60601
> requirements.
> 
> How about EMC standards? (maybe I should start a new thread...). Will this
> PC have to meet IEC 60601-1-2 for medical? Or EN 55022 for ITE? Bear in mind
> this is in the hospital operating room. I am familiar with medical equipment
> requirements but not so much with ITE, so forgive me if I'm making mistakes
> here.
> 
> My thoughts are all from the regulatory point of view. Of course if the
> customer wants compliance with a more stringent requirement, that is what he
> gets...
> 
> I appreciate all your thoughts/comments!
> 
> -Delphina
>   
> 
>  
>  
> >
> > At 09:10 AM 2/19/2003 -0800, Han, Delphina wrote:
> >
> > >Hi
> > >
> > >I am trying to find out safety requirements for devices that control and
> > >monitor equipment in a hospital lab (used for pathology). Does it fall
> under
> > >the IEC 61010 standard? If so, are there any requirements for use of
> > >isolation transformers in that standard?
> > >
> > >Thanks in advance for your response!
> > >
> > >-Delphina
> > >
> > >-------------------------------------------
> > >This message is from the IEEE EMC Society Product Safety
> > >Technical Committee emc-pstc discussion list.
> > >
> > >Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
> > >
> > >To cancel your subscription, send mail to:
> > >      [email protected]
> > >with the single line:
> > >      unsubscribe emc-pstc
> > >
> > >For help, send mail to the list administrators:
> > >      Ron Pickard:              [email protected]
> > >      Dave Heald:               [email protected]
> > >
> > >For policy questions, send mail to:
> > >      Richard Nute:           [email protected]
> > >      Jim Bacher:             [email protected]
> > >
> > >Archive is being moved, we will announce when it is back on-line.
> > >All emc-pstc postings are archived and searchable on the web at:
> > >     http://www.ieeecommunities.org/emc-pstc
>  
> 
> -------------------------------------------
> This message is from the IEEE EMC Society Product Safety
> Technical Committee emc-pstc discussion list.
> 
> Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/
> 
> To cancel your subscription, send mail to:
>      [email protected]
> with the single line:
>      unsubscribe emc-pstc
> 
> For help, send mail to the list administrators:
>      Ron Pickard:              [email protected]
>      Dave Heald:               [email protected]
> 
> For policy questions, send mail to:
>      Richard Nute:           [email protected]
>      Jim Bacher:             [email protected]
> 
> Archive is being moved, we will announce when it is back on-line.
> All emc-pstc postings are archived and searchable on the web at:
>     http://www.ieeecommunities.org/emc-pstc
> 




This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to