Hi,

In response to John Allen's comment, my understanding would be that the
medical electrical systems standard (60601-1-1) doesn't really help you
here.

This standard covers the situation where you have a system of either
interconnected electrical medical equipment or a mixed system of
interconnected electrical medical equipment and other equipment (e.g. IT).

The standard outlines additional measures that may be required for the
system to ensure that the patient safety is not impaired by the fact that
the electrical medical equipment (meeting 60601-1) is connected to
non-medical equipment (meeting e.g. 60950, 61010) or by the fact that the
non-medical equipment may be within the patient environment.

This is separate to the issue of whether an individual item of equipment is
classed as medical. The standard could be relevant though if your equipment
were required to be connected to medical equipment.

Hope that helps,

Regards,


-- 
Dr George Brimlow,
Principal Clinical Scientist
Medical Physics Dept
Queen's Medical Centre
Nottingham, NG7 2UH
UK
Tel. +44(0)115 924 9924 ext 44889
Fax  +44(0)115 942 2745
[email protected]
 


 > 
> Hi Folks

> 
> The basic answers to the questions of which items of "medical" equipment
> are, and are not, required to be compliant to 60601 are given in EN
> 60601-1-1 "Medical electrical equipment Part 1-1: General requirements for
> safety. Collateral standard: Safety requirements for medical electrical
> systems"
> 
> This describes the requirements and has diagrams to illustrate the various
> combinations of 60601 and non-60601 equipments.
> 
> Regards
> 
> John Allen
> -----Original Message-----

> From: Han, Delphina [mailto:[email protected]]
> 
> Hi
> 
> I am trying to find out safety requirements for devices that control and
> monitor equipment in a hospital lab (used for pathology). Does it fall under
> the IEC 61010 standard? If so, are there any requirements for use of
> isolation transformers in that standard?
> 
> Thanks in advance for your response!
> 
> -Delphina
> 



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
     [email protected]
with the single line:
     unsubscribe emc-pstc

For help, send mail to the list administrators:
     Ron Pickard:              [email protected]
     Dave Heald:               [email protected]

For policy questions, send mail to:
     Richard Nute:           [email protected]
     Jim Bacher:             [email protected]

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
    http://www.ieeecommunities.org/emc-pstc

Reply via email to