The term "monitor" in the Instrumentation context does not always infers a
display device. "Monitor", in this context, could refer to a data acquisition
device. 

I have not read anything in 601-x-x standards or the med directive, that would
make me believe that the intended end-use of an instrument by a physician, in
a lab enviroment only, requires compliance with 601-1. I would appreciate
standard/clause reference if I have missed this..

thanx, 
Brian 



My opinion is that unless there are unusual circumstances, an ITE (IEC60950) 
certified monitor should satisfy ordinary lab needs. Any dissent? 

Mike Harris/Teccom Co./Hayward CA 
----- Original Message ----- 
From: "Ronald R. Wellman" <[email protected]> 
To: "Han, Delphina" <[email protected]>; <[email protected]> 
Sent: Wednesday, February 19, 2003 5:27 PM 
Subject: Re: IEC 61010 requirements 
> 
> Hello Delphina, 
> 
> Laboratory equipment can be considered medical devices if you advertise 
> them as such or claim that they can be used be physicians to form medical 
> opinions. However, most laboratory equipment manufacturers get around this 
> by claiming that their products are intended for research use only so they 
> don't have to make pre-market approvals to regulatory agencies. How a 
> hospital lab uses non-approved devices is not governed by pre-market 
> approval regulations because the regulations are not intended to tell 
> physicians how to practice medicine. 
> 
> Therefore, to answer your question concerning the use of an isolation 
> transformer, this will depend on whether you need to conform to the 
leakage 
> current requirements of IEC 61010-1 or 60101-1. I suggest that you 
purchase 
> copies of both standards and understand their differences. However, based 
> on your equipment description it appears that 61010-1 will be the safety 
> standard you will need to use. 
> 
> Best regards, 
> Ron Wellman 
> 
> At 09:10 AM 2/19/2003 -0800, Han, Delphina wrote: 
> 
> >Hi 
> > 
> >I am trying to find out safety requirements for devices that control and 
> >monitor equipment in a hospital lab (used for pathology). Does it fall 
under 
> >the IEC 61010 standard? If so, are there any requirements for use of 
> >isolation transformers in that standard? 
> > 


Reply via email to