Hello Grace,

I do not believe that there is a specific separation requirement between
the EFT generator and the coupling clamp.  I have a system from Keytek that
came with a 0.5 m cable to connect the generator and the clamp.  If
anything, I find supplied cable to be too short for my needs.  From what I
know, the capacitive clamp should not cause interference with the EFT
generator even if they are in close proximity.

The coupling clamp should not create significant radiated noise.  In this I
mean that the noise radiated by the clamp will be far less than that
coupled into the EUT through the cable under test.  There should not be a
significant problem with the test if the EUT is less than 0.5 m from the
clamp as long as you have the 0.5 m length of exposed cable that is
specified.

I will admit that I am not an expert in EMC measurements.  I have been
working with susceptibility testing for about eight years, but it only a
fraction of the work that I perform.  I am posting my opinion (and it is
only an opinion) because I have not yet seen a response to your inquiry.  I
would like to encourage the more experienced EMC professionals to let me
know if my though process is sound on this issue.  If I am wrong, let me
know; I am eager to learn more about the rationale behind the test
procedure.

Ted Eckert
American Power Conversion/MGE
http://www.apc.com/

The items contained in this e-mail reflect the personal opinions of the
writer and are only provided for the assistance of the reader. The writer
is not speaking in an official capacity for APC-MGE or Schneider Electric.
The speaker does not represent APC-MGE's or Schneider Electric's official
position on any matter.


                                                                           
             "Grace Lin"                                                   
             <gracelinnj@gmail                                             
             .com>                                                      To 
             Sent by:                  [email protected]                   
             [email protected]                                          cc 
                                                                           
                                                                   Subject 
             10/29/2007 02:24          EFT Setup                           
             PM                                                            
                                                                           
                                                                           
                                                                           
                                                                           
                                                                           




Dear Members,

I have two questions and appreciate your inputs:

1. Is there any clearance distance/requirement between an EFT generator and
a capacitive coupling clamp (ccc)?

2. Is there any clearance distance/requirement between EUT and a capacitive
coupling clamp (CCC)?  From Figure 1 of IEC 61000-4-4, the cable length
between the capacitive coupling clamp and the EUT "should be 0.5m +/- 0.05
m".  Does this mean the distance between EUT and CCC should also be 0.5m?

We built a 4ft x 8ft table with the same size of metal sheet (reference to
ground reference) on top of it.  I would like to place an EFT generator,
CCC, and EUT on top of the table.

Thank you for your time and look forward to your help.

Best regards,
Grace Lin
- ---------------------------------------------------------------- This
message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/


To post a message to the list, send your e-mail to [email protected]


Instructions: http://listserv.ieee.org/request/user-guide.html


List rules: http://www.ieee-pses.org/listrules.html


For help, send mail to the list administrators:


Scott Douglas [email protected] Mike Cantwell [email protected]


For policy questions, send mail to:


Jim Bacher: [email protected] David Heald: [email protected]


All emc-pstc postings are archived and searchable on the web at:


http://www.ieeecommunities.org/emc-pstc

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Reply via email to