Because the field around the EUT is heavily distorted during the test.

It might work for small EUT.

 

THE best method is calibration without EUT using a field probe, and monitoring

forward power during radiation of the EUT.

If you are worried about the antenna, monitor the reflected power also.

If that deviates SUBSTANTIALLY  , you have a problem.

Small deviations are to be expected due to the properties of your room

changing when EUT is in place. (think close field coupling between EUT and
antenna!!!)

 

A field probe can also be used, but should NOT be used to adjust levels.

 

Alternately, if you use the same antenna as during receive, do a check

on receive quality using a small portable reference source  (also before

installing EUT as EUT will have impact on the transfer) .

If the result is within a defined margin the  measurement chain is OK.

 

 

We have a small dipole installed at the back of the room (behind EUT)

and verify the transfer from that antenna (connected to a generator)

to the measuring receiver at a number of frequencies before

each emission measurement and before each immunity

measurement. Antenna, receiver and amplifier are automatically

coupled using coaxial switches so it’s a fast procedure.

If the majority of the frequencies received are within a certain margin

(+/- 2dB ) the receive chain (and so the transmit chain) is OK.



Better accuracy cannot be obtained, unless the EUT is removed.

Even worse, at some frequencies EUT radiates , or the transfer

is influenced by the EUT metal work / cabling.

That is why we use a number of frequencies, and no they

are of course not related harmonically, nor popular clock frequencies.

 

We also use the procedure to verify proper antenna orientation.

 

The results are logged with each measurement and in a generic

log file.

 

As we write our own test software that was not a big deal.

 

 

Regards,

Ing.  Gert Gremmen, BSc

 

 

 

[email protected] <mailto:[email protected]> 

www.cetest.nl


Kiotoweg 363

3047 BG Rotterdam

T 31(0)104152426
F 31(0)104154953

Before printing, think about the environment. 

 

 

Van: [email protected] [mailto:[email protected]] Namens Gärdin Petter
Verzonden: Wednesday, September 28, 2011 5:14 PM
Aan: [email protected]
Onderwerp: RE: [PSES] BS EN61000-4-3 and calibrated fields.

 

According to the standard, 

8.2 Execution of the test

…

Before testing the intensity of the calibrated field strength should be
checked to verify that the test equipment/system is operating properly.

…

 

It says should, and the easiest way is to check field strength is with a field
probe, so why not use it during the whole test to make sure nothing goes wrong
during testing. It's usually much more expensive to rerun test than to make
sure you do it correctly.

 

Petter

 

From: Grasso, Charles [mailto:[email protected]] 
Sent: den 28 september 2011 17:00
To: [email protected]
Subject: Re: [PSES] BS EN61000-4-3 and calibrated fields.

 

But is that REQUIRED in 61000-4-3?

 

Best Regards
Charles Grasso
Compliance Engineer
Echostar Communications
(w) 303-706-5467
(c) 303-204-2974
(t) [email protected] <mailto:[email protected]> 
(e) [email protected] <mailto:[email protected]> 

(e2) [email protected] <mailto:[email protected]> 

 

From: [email protected] [mailto:[email protected]] On Behalf Of ce-test,
qualified testing bv - Gert Gremmen
Sent: Wednesday, September 28, 2011 12:30 AM
To: Brent G DeWitt; Kunde, Brian; [email protected]
Subject: RE: [PSES] BS EN61000-4-3 and calibrated fields.

 

EUT has big impact on local field. 

Field NEEDS to be leveled without EUT.

 

Regards,

Ing.  Gert Gremmen, BSc

 

 

 

[email protected]

www.cetest.nl


Kiotoweg 363

3047 BG Rotterdam

T 31(0)104152426
F 31(0)104154953

 

Before printing, think about the environment. 

 

 

Van: [email protected] [mailto:[email protected]] Namens Brent G DeWitt
Verzonden: Wednesday, September 28, 2011 1:21 AM
Aan: 'Kunde, Brian'; [email protected]
Onderwerp: RE: [PSES] BS EN61000-4-3 and calibrated fields.

 

We have done this at several labs I work for, but it was always just a sanity
check to make sure that neither the amplifier nor cables to the antenna had
failed.  It is totally useless as a function for leveling for commercial
immunity measurements.

 

Best regards

Brent

 

From: Kunde, Brian [mailto:[email protected]] 
Sent: Tuesday, September 27, 2011 8:19 PM
To: [email protected]
Subject: Re: [PSES] BS EN61000-4-3 and calibrated fields.

 

Not to hijack this thread, but why do some test labs setup the Field Probe in
the chamber next to the EUT during the test?  The measurement with the EUT in
the field should be all over the place and would have little value. Am I
missing something or are some labs leveling off the Field Probe during
testing?  

 

From: [email protected] [mailto:[email protected]] On Behalf Of Ian White (UK)
Sent: Tuesday, September 27, 2011 6:00 AM
To: IEEE Forum ([email protected])
Subject: BS EN61000-4-3 and calibrated fields.

 

Could you please  help:

 

Could you explain what is  the correct procedure when using an anechoic
chamber which is  calibrated and operating to EN61000-4-3.

 

During calibration of the chamber,   a calibration file is made and it is
assessed according to the guidelines laid down under BS EN61000-4-3. During
calibration the chamber is empty.

 

During a normal test run a UUT is put into the chamber. The questions is :

 

a)      Do you use the calibration file as the starting point and adjust the
field strength which may have changed due to the UUT being in the chamber.

b)      Just take the calibration file and run it and except the filed
strengths would have changed due to the UUT being in the chamber.

 

 

Thanks

 

Ian White

Compliance and Reliability.

 

 

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://product-compliance.oc.ieee.org/
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 

________________________________


LECO Corporation Notice: This communication may contain confidential
information intended for the named recipient(s) only. If you received this by
mistake, please destroy it and notify us of the error. Thank you. 


-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://product-compliance.oc.ieee.org/
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the
list, send your e-mail to <[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://product-compliance.oc.ieee.org/
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://product-compliance.oc.ieee.org/
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://product-compliance.oc.ieee.org/
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://product-compliance.oc.ieee.org/
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at
http://product-compliance.oc.ieee.org/
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 


Reply via email to