Actually it is the best method to use for 61000 – 4 – 3. Leveling with modulation off does take more time, but only by a bit (a couple of seconds per step). The advantage is a more accurate leveling regardless of your modulation For medical devices where the modulation is at 2 Hz, it is almost impossible for a power meter to actually level correctly during the playback process.
Best regards, Brent DeWitt PS my apologies for any misspellings or bad grammar. I have a collarbone and I am using Dragon NaturallySpeaking to dictate this. From: Deniz Demirci [mailto:[email protected]] Sent: Tuesday, September 27, 2011 8:44 PM To: [email protected] Subject: Re: [PSES] BS EN61000-4-3 and calibrated fields. The option a is also acceptable in some standards such as GR-1089-CORE. For each frequency step; The modulation has to be turned off, adjust the level, apply modulation… It is quite time consuming Again, this method is not for EN 61000-4-3 Deniz On Tue, Sep 27, 2011 at 3:59 AM, Ian White (UK) <[email protected]> wrote: Could you please help: Could you explain what is the correct procedure when using an anechoic chamber which is calibrated and operating to EN61000-4-3. During calibration of the chamber, a calibration file is made and it is assessed according to the guidelines laid down under BS EN61000-4-3. During calibration the chamber is empty. During a normal test run a UUT is put into the chamber. The questions is : a) Do you use the calibration file as the starting point and adjust the field strength which may have changed due to the UUT being in the chamber. b) Just take the calibration file and run it and except the filed strengths would have changed due to the UUT being in the chamber. Thanks Ian White Compliance and Reliability. - ---------------------------------------------------------------- This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <[email protected]> All emc-pstc postings are archived and searchable on the web at http://product-compliance.oc.ieee.org/ Graphics (in well-used formats), large files, etc. can be posted to that URL. Website: http://www.ieee-pses.org/ Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <[email protected]> Mike Cantwell <[email protected]> For policy questions, send mail to: Jim Bacher <[email protected]> David Heald <[email protected]> - ---------------------------------------------------------------- This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <[email protected]> All emc-pstc postings are archived and searchable on the web at http://product-compliance.oc.ieee.org/ Graphics (in well-used formats), large files, etc. can be posted to that URL. Website: http://www.ieee-pses.org/ Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <[email protected]> Mike Cantwell <[email protected]> For policy questions, send mail to: Jim Bacher <[email protected]> David Heald <[email protected]> - ---------------------------------------------------------------- This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <[email protected]> All emc-pstc postings are archived and searchable on the web at http://product-compliance.oc.ieee.org/ Graphics (in well-used formats), large files, etc. can be posted to that URL. Website: http://www.ieee-pses.org/ Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <[email protected]> Mike Cantwell <[email protected]> For policy questions, send mail to: Jim Bacher <[email protected]> David Heald <[email protected]>