We talked to the PSU manufacturer and they say that the 60601-1 is more
severe than 60950-1 for PSU, so that's why they only do CB test and cert
according to 60601-1.
Maybe only 80% truth?  :-)

#Amund



-----Opprinnelig melding-----
Fra: Richard Nute [mailto:[email protected]] 
Sendt: 14. september 2010 23:08
Til: amund
Kopi: [email protected]
Emne: Re: IEC 60950-1 vs IEC 60601-1




> Is it possible to make use of a CB certificated (IEC 60601-1) PSU, into a
> ITE which shall be approved according to IEC60950-1?

Yes.

> Well, I assume that's OK, but should we expect additional tests on the PSU
/
> ITE since it's not CB acc to IEC60950-1?

This will be a function of the certification engineer
assigned to your project.

You can help this out by filling out a 60950-1 CB
report using the data from the 60601-1 CB report.
This will tell you EXACTLY where the differences are.
And, it will give you negotiating power to avoid a
complete re-test (that a lazy cert engineer will
impose upon you).

Of course, end-product tests such as the temperature
tests will need to be carried out in your end-product
as they will surely be different than in the medical
equipment.

If you are buying the PSU from another party, I
would recommend that you ask the PSU manufacturer
to seek out complementary certification to 60950-1.


Richard Nute
Product Safety Consultant
San Diego, California, USA

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to