> Is it possible to make use of a CB certificated (IEC 60601-1) PSU, into a > ITE which shall be approved according to IEC60950-1?
Yes. > Well, I assume that's OK, but should we expect additional tests on the PSU / > ITE since it's not CB acc to IEC60950-1? This will be a function of the certification engineer assigned to your project. You can help this out by filling out a 60950-1 CB report using the data from the 60601-1 CB report. This will tell you EXACTLY where the differences are. And, it will give you negotiating power to avoid a complete re-test (that a lazy cert engineer will impose upon you). Of course, end-product tests such as the temperature tests will need to be carried out in your end-product as they will surely be different than in the medical equipment. If you are buying the PSU from another party, I would recommend that you ask the PSU manufacturer to seek out complementary certification to 60950-1. Richard Nute Product Safety Consultant San Diego, California, USA - ---------------------------------------------------------------- This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <[email protected]> All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc Graphics (in well-used formats), large files, etc. can be posted to that URL. Website: http://www.ieee-pses.org/ Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <[email protected]> Mike Cantwell <[email protected]> For policy questions, send mail to: Jim Bacher: <[email protected]> David Heald: <[email protected]>