Had a 200 Watt power supply last week that simply shut-off (fold back)
at 10V/m . It was applied in a laboratory equipment for creating
expensive bio-cultures for vaccination ... These cultures may cost up
to a million dollar. This was not a cheap unknown manufacturer, but one
With a quality system implemented. Design update without re-test ?
Maybe it had an obscure fault. One of the interesting cases I had some
years ago was an all-analogue hi-fi preamplifier which was radiating at
17 MHz. One section of a quad op-amp, used for producing a stable DC
level, was faulty. If it had been in the signal chain, the fault would
have been discovered. But you don't test analogue audio preamps for
high-frequency emissions!
--
OOO - Own Opinions Only. See www.jmwa.demon.co.uk
The longer it takes to make a point, the more obtuse it proves to be.
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK
-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc discussion
list. To post a message to the list, send your e-mail to <[email protected]>
All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html
Attachments are not permitted but the IEEE PSES Online Communities site at
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used
formats), large files, etc.
Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html
For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>
For policy questions, send mail to:
Jim Bacher: <[email protected]>
David Heald: <[email protected]>