All,

Thanks all for the responses.  The power supply I am using is an external power 
supply (similar to an external PC laptop supply).  From the manufacturers CB 
report, I can see that it was tested beyond the requirements of the original 
EN60950 requirements, and does meet the requirements of the EN61010:2010 
humidity soaking.

All that said, it appears that it still valid to continue using in our product. 
 Thanks all for the very helpful information, and pointing out the respective 
sections of the EN61010 standard.  It also helps to now have the latest 
EN61010-1 standard in hand.


Thanks,

Gary


From: Doug Powell [mailto:[email protected]]
Sent: Tuesday, September 24, 2013 10:02 AM
To: [email protected]
Subject: Re: [PSES] Using an IEC60950 power supply to power an EN61010 equipment

Gary,

One other concern you should be aware of. With 60950, the majority of products 
are evaluated to Overvoltage Category II, while it is possible that your 
61010-1 product is evaluated to OV Cat III. This will have a significant effect 
on determination of your minimum  clearance requirements as well as your surge 
immunity levels when applying EN 61326-1 for your EMC evaluation. Your 60950-1 
approved power supply may not be adequate to the task.  Best to double 
check.the data sheet and if necessary call the manufacturer. Also, I always 
like to get a full list of the conditions of use for the power supply.

Thanks -doug


Douglas Powell
http://www.linkedin.com/in/dougp01




Sent from my BlackBerry 10 smartphone on the Verizon Wireless 4G LTE network.
From: Kunde, Brian
Sent: Tuesday, September 24, 2013 6:52 AM
To: [email protected]<mailto:[email protected]>
Reply To: Kunde, Brian
Subject: Re: [PSES] Using an IEC60950 power supply to power an EN61010
equipment


I know this doesn’t answer your questions specifically, but thought they may be 
helpful.
The Other Brian

EN 61010-1:2010 section 1.1.3 says,
NOTE Computing devices and similar equipment within the scope of IEC 60950 and 
conforming to its requirements are considered to be suitable for use with 
equipment within the scope of this standard. However, some of the requirements 
of IEC 60950 for resistance to moisture and liquids are less stringent than 
those in this standard (see 5.4.4 second paragraph).

Section 5.4.4 says,
If equipment conforming to IEC 60950 is used with equipment conforming to this 
standard, and if there is a HAZARD due to moisture or liquids, the instructions 
for use shall specify any additional precautions necessary.

Also, read section 14 regarding the use of components and subassemblies “such 
as power supplies”…

From: Darnel, Gary [mailto:[email protected]]
Sent: Monday, September 23, 2013 5:01 PM
To: [email protected]<mailto:[email protected]>
Subject: [PSES] Using an IEC60950 power supply to power an EN61010 equipment

Dear All,

I have a question about insulation requirements of the EN61010:2010 standards 
(clause 6).

Are the insulation requirements of EN61010 / IEC61010:2010 identical to those 
found in the latest IEC60950 (2009 I believe, including all amendments)?

Basically I need to know if I use a third party certified power supply marked 
to EN60950:2009, will it meet the insulation requirements of EN61010 
(clearance, creepage, protective bonding…etc)

Maybe in a nutshell, I am wondering where the EN61010 insulation requirements 
were derived from.  Are they the same source as the EN60950 requiements.


Thanks,

Gary Darnel

-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<http://www.ieee-pses.org/list.html>
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]<mailto:[email protected]>>
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>

________________________________
LECO Corporation Notice: This communication may contain confidential 
information intended for the named recipient(s) only. If you received this by 
mistake, please destroy it and notify us of the error. Thank you.
-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<http://www.ieee-pses.org/list.html>
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]<mailto:[email protected]>>
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>


-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://www.ieee-pses.org/list.html (including how to 
unsubscribe)<http://www.ieee-pses.org/list.html>
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]<mailto:[email protected]>>
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to