On Thu, Dec 8, 2016 at 1:40 PM, Glasser, Matthew <[email protected]> wrote:

> What phase encoding directions are you using?
>
> Peace,
>
> Matt.
>
> From: neuroimage analyst <[email protected]>
> Date: Thursday, December 8, 2016 at 1:37 PM
> To: "Harms, Michael" <[email protected]>
> Cc: Matt Glasser <[email protected]>, "[email protected]" <
> [email protected]>
> Subject: Re: [HCP-Users] FW: Skyra - BOLD and diffusion MRI protocols
>
> Hi,
>
> Just an update:
>
> 1) BOLD: I was able to run the sequence with all the parameters as
> mentioned but with the resolution of 2.5mm3 and Partial fourier 7/8. I was
> unable to decrease the resolution with that Partial Fourier factor of
> anything or resolution of 2.5mm3 with no partial fourier. I will scan a
> subject today and see how the data looks.
>
> 2) dMRI: MB factor = 3. RF pulse clipping issue was resolved. TR is 5600
> and TE is 100 ms , 2 b-values and 71 diffusion encoding directions running
> under free mode. Resolution of 1.5mm3; scan time of 15 mins. I added a PA
> phase encoding with 2 averages of b0. Again, I will test the sequence with
> a subject and different resolutions to see if the data is acceptable.
>
> Thank you for all your help and advice/suggestions.
>
> Regards
>
> --VM
>
> On Wed, Dec 7, 2016 at 12:31 PM, Harms, Michael <[email protected]> wrote:
>
>>
>> Oh.  I read too quickly and conflated the problem you were having for the
>> BOLD with the dMRI.  So, you can set up a BOLD scan with those parameters,
>> and it only fails at run-time?  Can you achieve a higher MB factor if you
>> relax the spatial resolution to say 2.4 mm.  (Although that shouldn’t be
>> necessary, because it is possible to do 2.0 mm, MB=8 BOLD on a Trio).  You
>> might have to ask the CMRR folks via their github issues page:
>> https://github.com/CMRR-C2P/MB/issues
>>
>> Regarding the dMRI:  What MB factor were you using there?   And did you
>> resolve the issue with the RF pulse clipping?   (Either by lengthening the
>> pulses further (within reason), or you could try using the “MB RF phase
>> scramble” option; and if that isn’t sufficient, try the “Time-shifted MB
>> RF” option as well).
>>
>> cheers,
>> -MH
>>
>> --
>> Michael Harms, Ph.D.
>> -----------------------------------------------------------
>> Conte Center for the Neuroscience of Mental Disorders
>> Washington University School of Medicine
>> Department of Psychiatry, Box 8134
>> 660 South Euclid Ave.Tel: 314-747-6173 <(314)%20747-6173>
>> St. Louis, MO  63110Email: [email protected]
>>
>> From: neuroimage analyst <[email protected]>
>> Date: Wednesday, December 7, 2016 at 1:54 PM
>> To: Michael Harms <[email protected]>
>> Cc: "Glasser, Matthew" <[email protected]>, "
>> [email protected]" <[email protected]>
>> Subject: Re: [HCP-Users] FW: Skyra - BOLD and diffusion MRI protocols
>>
>> I apologize for not making the question very clear. I think I have sort
>> of optimized the diffusion protocol the way I wanted it to be. I just need
>> to do a human scan and compare SNR between different resolutions to see the
>> limit where i can get acceptable data.
>>
>> The issue that I am unable to solve is this:
>> BOLD: We were hoping to get a TR in the range of 750-900ms, res = 2mm3,
>> echo spacing (ES) approx 0.65. We were able to get to echo spacing of 0.69
>> and all the other parameters with MB = 8. However, the sequence doesn't run
>> and it gives us "Max amplitude overflow on gradient z axis" after MB factor
>> exceeds 2.  It appears to me that then we have to sacrifice TR and only run
>> with MB = 2. We are using 32 channel head coil.
>>
>> Thank you for any suggestions.
>>
>> Regards
>>
>> Virendra
>>
>> On Wed, Dec 7, 2016 at 11:22 AM, Harms, Michael <[email protected]> wrote:
>>
>>>
>>> I wouldn’t think that increasing the MB factor would increase the
>>> gradient requirements that much, so I’m wondering if trying to run with
>>> b_max = 2500 isn’t at the very edge of what can be achieved on the Skyra
>>> for that particular gradient direction set.  What happens if you lower the
>>> b_max to say 1500?  Can you then increase the MB factor without getting
>>> that error?
>>>
>>> --
>>> Michael Harms, Ph.D.
>>> -----------------------------------------------------------
>>> Conte Center for the Neuroscience of Mental Disorders
>>> Washington University School of Medicine
>>> Department of Psychiatry, Box 8134
>>> 660 South Euclid Ave.Tel: 314-747-6173 <(314)%20747-6173>
>>> St. Louis, MO  63110Email: [email protected]
>>>
>>> From: neuroimage analyst <[email protected]>
>>> Date: Wednesday, December 7, 2016 at 1:03 PM
>>> To: Michael Harms <[email protected]>
>>> Cc: "Glasser, Matthew" <[email protected]>, "
>>> [email protected]" <[email protected]>
>>> Subject: Re: [HCP-Users] FW: Skyra - BOLD and diffusion MRI protocols
>>>
>>> Thanks for that reply, Micahel. Exactly, what I plan to do for diffusion
>>> protocol.
>>>
>>> Just wondering though on your or anyone else's suggestion on running
>>> BOLD with MB>2.
>>>
>>> Thanks
>>>
>>> Regards
>>>
>>> --VM
>>>
>>>
>>> On Wed, Dec 7, 2016 at 10:50 AM, Harms, Michael <[email protected]>
>>> wrote:
>>>
>>>>
>>>> Hi,
>>>> It appears that you are trying to replicate the basics of one of our
>>>> Lifespan pilot protocols on a Skyra — i.e., 1.5 mm voxels with 20 total min
>>>> of dMRI scanning.  But due to the longer TE and longer TR (resulting in
>>>> fewer total directions), you are definitely going to have worse overall SNR
>>>> than what we achieved.  We highly suggest that you collect some pilot data
>>>> to assess the quality of your proposed dMRI protocol with your end-measures
>>>> of interest, including possibly a comparison against a protocol with larger
>>>> voxels (e.g., 1.7-1.8 mm voxels) that would recover some SNR.
>>>>
>>>> cheers,
>>>> -MH
>>>>
>>>> --
>>>> Michael Harms, Ph.D.
>>>> -----------------------------------------------------------
>>>> Conte Center for the Neuroscience of Mental Disorders
>>>> Washington University School of Medicine
>>>> Department of Psychiatry, Box 8134
>>>> 660 South Euclid Ave.Tel: 314-747-6173 <(314)%20747-6173>
>>>> St. Louis, MO  63110Email: [email protected]
>>>>
>>>> From: <[email protected]> on behalf of "Glasser,
>>>> Matthew" <[email protected]>
>>>> Date: Wednesday, December 7, 2016 at 11:49 AM
>>>> To: "[email protected]" <[email protected]>
>>>> Subject: [HCP-Users] FW: Skyra - BOLD and diffusion MRI protocols
>>>>
>>>>
>>>>
>>>> From: Matt Glasser <[email protected]>
>>>> Date: Wednesday, December 7, 2016 at 10:14 AM
>>>> To: neuroimage analyst <[email protected]>
>>>> Subject: Re: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>>>>
>>>> I wouldn’t acquire multiple repetitions of the same directions, but
>>>> rather more unique directions.  If necessary you can acquire the DWIs with
>>>> only a single phase encoding direction in an effort to get more unique
>>>> diffusion directions, but it is important that you get phase reversed b0
>>>> images.
>>>>
>>>> As for the MB error, I don’t know how to fix it, but you should
>>>> definitely be able to do better than MB=2 with a 32 channel head coil.
>>>> Hopefully someone else on the list knows how to fix this.
>>>>
>>>> Peace,
>>>>
>>>> Matt.
>>>>
>>>> From: neuroimage analyst <[email protected]>
>>>> Date: Tuesday, December 6, 2016 at 8:54 PM
>>>> To: Matt Glasser <[email protected]>
>>>> Subject: Re: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>>>>
>>>> Hi Matt,
>>>>
>>>> 32 channel head coil.
>>>>
>>>> I was able to run the diffusion sequence on a phantom when TR/TE was
>>>> 7000/106 ms. Since the scan time with this tr is already 20 minutes, I was
>>>> thinking to acquire just b0 with PA phase encoding and the other 64
>>>> directions 2b data+b0  with AP. will that be okay? my concern is snr at te
>>>> will not be very good and data may not be too great and may be atleast 2
>>>> reps are desired?
>>>>
>>>> Thanks
>>>>
>>>> Regards
>>>>
>>>> -VM
>>>>
>>>> On Dec 6, 2016 6:46 PM, "Glasser, Matthew" <[email protected]> wrote:
>>>>
>>>>> What coil are you using?
>>>>>
>>>>> Peace,
>>>>>
>>>>> Matt.
>>>>>
>>>>> From: <[email protected]> on behalf of neuroimage
>>>>> analyst <[email protected]>
>>>>> Date: Tuesday, December 6, 2016 at 11:56 AM
>>>>> To: "[email protected]" <[email protected]>
>>>>> Subject: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>>>>>
>>>>> Hi, HCP Users and Developers.
>>>>>
>>>>> We have been trying to build a protocol for BOLD and DWI on our Skyra
>>>>> (VE11C) using the latest release of HCP pulse sequences. But we
>>>>> haven't been completely successful with what we wanted to achieve.
>>>>>
>>>>> A) BOLD: We were hoping to get a TR in the range of 750ms, res = 2mm3,
>>>>> echo spacing (ES) approx 0.65. We were able to get to echo spacing of 0.69
>>>>> and all the other parameters with MB = 8. However, the sequence doesn't 
>>>>> run
>>>>> and it gives us "Max amplitude overflow on gradient z axis" after MB 
>>>>> factor
>>>>> exceeds 2.  It appears to me that then we have to sacrifice TR and only 
>>>>> run
>>>>> with MB = 2. Is there anybody with a Skyra who has able to achieve what we
>>>>> are hoping for and willing to share the protocol with us? OR if somebody
>>>>> could guide us to resolve the error of amplitude overflow.
>>>>>
>>>>> B) DWI: The idea was to have res = 1.5mm3, 2b-values at 1000 and 2500,
>>>>> 64 directions with the best BW and ES achievable. The sequence runs for 7
>>>>> minutes out of total 13 minutes and gives gradient power amplifier error.
>>>>> On Michael Harms's suggestion, I adjusted flip angle to 78/160 at a TR of
>>>>> 5500 ms with excite/refocus pulse duration of 3840 and 7680. there was a
>>>>> pop up warning that RF is clipped and the maximum refocusing angle was 142
>>>>> instead of 160. I placed a 64 directions diffusion vector and ran the
>>>>> sequence under "free" mode. Again, I will appreciate if  there is anybody
>>>>> with a Skyra who has able to achieve what we are hoping for and willing to
>>>>> share the protocol with us, along with the diffusion vector sets? OR if
>>>>> somebody could guide us to resolve the error.
>>>>>
>>>>> Thank you.
>>>>>
>>>>> Regards
>>>>>
>>>>> --VM
>>>>>
>>>>> _______________________________________________
>>>>> HCP-Users mailing list
>>>>> [email protected]
>>>>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>>>
>>>>>
>>>>> ------------------------------
>>>>>
>>>>> The materials in this message are private and may contain Protected
>>>>> Healthcare Information or other information of a sensitive nature. If you
>>>>> are not the intended recipient, be advised that any unauthorized use,
>>>>> disclosure, copying or the taking of any action in reliance on the 
>>>>> contents
>>>>> of this information is strictly prohibited. If you have received this 
>>>>> email
>>>>> in error, please immediately notify the sender via telephone or return 
>>>>> mail.
>>>>>
>>>>
>>>> ------------------------------
>>>>
>>>> The materials in this message are private and may contain Protected
>>>> Healthcare Information or other information of a sensitive nature. If you
>>>> are not the intended recipient, be advised that any unauthorized use,
>>>> disclosure, copying or the taking of any action in reliance on the contents
>>>> of this information is strictly prohibited. If you have received this email
>>>> in error, please immediately notify the sender via telephone or return 
>>>> mail.
>>>>
>>>> _______________________________________________
>>>> HCP-Users mailing list
>>>> [email protected]
>>>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>>
>>>>
>>>> ------------------------------
>>>>
>>>> The materials in this message are private and may contain Protected
>>>> Healthcare Information or other information of a sensitive nature. If you
>>>> are not the intended recipient, be advised that any unauthorized use,
>>>> disclosure, copying or the taking of any action in reliance on the contents
>>>> of this information is strictly prohibited. If you have received this email
>>>> in error, please immediately notify the sender via telephone or return 
>>>> mail.
>>>>
>>>> _______________________________________________
>>>> HCP-Users mailing list
>>>> [email protected]
>>>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>>
>>>
>>>
>>> ------------------------------
>>>
>>> The materials in this message are private and may contain Protected
>>> Healthcare Information or other information of a sensitive nature. If you
>>> are not the intended recipient, be advised that any unauthorized use,
>>> disclosure, copying or the taking of any action in reliance on the contents
>>> of this information is strictly prohibited. If you have received this email
>>> in error, please immediately notify the sender via telephone or return mail.
>>>
>>
>>
>> ------------------------------
>>
>> The materials in this message are private and may contain Protected
>> Healthcare Information or other information of a sensitive nature. If you
>> are not the intended recipient, be advised that any unauthorized use,
>> disclosure, copying or the taking of any action in reliance on the contents
>> of this information is strictly prohibited. If you have received this email
>> in error, please immediately notify the sender via telephone or return mail.
>>
>
>
> ------------------------------
>
> The materials in this message are private and may contain Protected
> Healthcare Information or other information of a sensitive nature. If you
> are not the intended recipient, be advised that any unauthorized use,
> disclosure, copying or the taking of any action in reliance on the contents
> of this information is strictly prohibited. If you have received this email
> in error, please immediately notify the sender via telephone or return mail.
>

_______________________________________________
HCP-Users mailing list
[email protected]
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

Reply via email to