I apologize for not making the question very clear. I think I have sort of
optimized the diffusion protocol the way I wanted it to be. I just need to
do a human scan and compare SNR between different resolutions to see the
limit where i can get acceptable data.

The issue that I am unable to solve is this:
BOLD: We were hoping to get a TR in the range of 750-900ms, res = 2mm3,
echo spacing (ES) approx 0.65. We were able to get to echo spacing of 0.69
and all the other parameters with MB = 8. However, the sequence doesn't run
and it gives us "Max amplitude overflow on gradient z axis" after MB factor
exceeds 2.  It appears to me that then we have to sacrifice TR and only run
with MB = 2. We are using 32 channel head coil.

Thank you for any suggestions.

Regards

Virendra

On Wed, Dec 7, 2016 at 11:22 AM, Harms, Michael <[email protected]> wrote:

>
> I wouldn’t think that increasing the MB factor would increase the gradient
> requirements that much, so I’m wondering if trying to run with b_max = 2500
> isn’t at the very edge of what can be achieved on the Skyra for that
> particular gradient direction set.  What happens if you lower the b_max to
> say 1500?  Can you then increase the MB factor without getting that error?
>
> --
> Michael Harms, Ph.D.
> -----------------------------------------------------------
> Conte Center for the Neuroscience of Mental Disorders
> Washington University School of Medicine
> Department of Psychiatry, Box 8134
> 660 South Euclid Ave. Tel: 314-747-6173 <(314)%20747-6173>
> St. Louis, MO  63110 Email: [email protected]
>
> From: neuroimage analyst <[email protected]>
> Date: Wednesday, December 7, 2016 at 1:03 PM
> To: Michael Harms <[email protected]>
> Cc: "Glasser, Matthew" <[email protected]>, "
> [email protected]" <[email protected]>
> Subject: Re: [HCP-Users] FW: Skyra - BOLD and diffusion MRI protocols
>
> Thanks for that reply, Micahel. Exactly, what I plan to do for diffusion
> protocol.
>
> Just wondering though on your or anyone else's suggestion on running BOLD
> with MB>2.
>
> Thanks
>
> Regards
>
> --VM
>
>
> On Wed, Dec 7, 2016 at 10:50 AM, Harms, Michael <[email protected]> wrote:
>
>>
>> Hi,
>> It appears that you are trying to replicate the basics of one of our
>> Lifespan pilot protocols on a Skyra — i.e., 1.5 mm voxels with 20 total min
>> of dMRI scanning.  But due to the longer TE and longer TR (resulting in
>> fewer total directions), you are definitely going to have worse overall SNR
>> than what we achieved.  We highly suggest that you collect some pilot data
>> to assess the quality of your proposed dMRI protocol with your end-measures
>> of interest, including possibly a comparison against a protocol with larger
>> voxels (e.g., 1.7-1.8 mm voxels) that would recover some SNR.
>>
>> cheers,
>> -MH
>>
>> --
>> Michael Harms, Ph.D.
>> -----------------------------------------------------------
>> Conte Center for the Neuroscience of Mental Disorders
>> Washington University School of Medicine
>> Department of Psychiatry, Box 8134
>> 660 South Euclid Ave.Tel: 314-747-6173 <(314)%20747-6173>
>> St. Louis, MO  63110Email: [email protected]
>>
>> From: <[email protected]> on behalf of "Glasser,
>> Matthew" <[email protected]>
>> Date: Wednesday, December 7, 2016 at 11:49 AM
>> To: "[email protected]" <[email protected]>
>> Subject: [HCP-Users] FW: Skyra - BOLD and diffusion MRI protocols
>>
>>
>>
>> From: Matt Glasser <[email protected]>
>> Date: Wednesday, December 7, 2016 at 10:14 AM
>> To: neuroimage analyst <[email protected]>
>> Subject: Re: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>>
>> I wouldn’t acquire multiple repetitions of the same directions, but
>> rather more unique directions.  If necessary you can acquire the DWIs with
>> only a single phase encoding direction in an effort to get more unique
>> diffusion directions, but it is important that you get phase reversed b0
>> images.
>>
>> As for the MB error, I don’t know how to fix it, but you should
>> definitely be able to do better than MB=2 with a 32 channel head coil.
>> Hopefully someone else on the list knows how to fix this.
>>
>> Peace,
>>
>> Matt.
>>
>> From: neuroimage analyst <[email protected]>
>> Date: Tuesday, December 6, 2016 at 8:54 PM
>> To: Matt Glasser <[email protected]>
>> Subject: Re: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>>
>> Hi Matt,
>>
>> 32 channel head coil.
>>
>> I was able to run the diffusion sequence on a phantom when TR/TE was
>> 7000/106 ms. Since the scan time with this tr is already 20 minutes, I was
>> thinking to acquire just b0 with PA phase encoding and the other 64
>> directions 2b data+b0  with AP. will that be okay? my concern is snr at te
>> will not be very good and data may not be too great and may be atleast 2
>> reps are desired?
>>
>> Thanks
>>
>> Regards
>>
>> -VM
>>
>> On Dec 6, 2016 6:46 PM, "Glasser, Matthew" <[email protected]> wrote:
>>
>>> What coil are you using?
>>>
>>> Peace,
>>>
>>> Matt.
>>>
>>> From: <[email protected]> on behalf of neuroimage
>>> analyst <[email protected]>
>>> Date: Tuesday, December 6, 2016 at 11:56 AM
>>> To: "[email protected]" <[email protected]>
>>> Subject: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>>>
>>> Hi, HCP Users and Developers.
>>>
>>> We have been trying to build a protocol for BOLD and DWI on our Skyra
>>> (VE11C) using the latest release of HCP pulse sequences. But we
>>> haven't been completely successful with what we wanted to achieve.
>>>
>>> A) BOLD: We were hoping to get a TR in the range of 750ms, res = 2mm3,
>>> echo spacing (ES) approx 0.65. We were able to get to echo spacing of 0.69
>>> and all the other parameters with MB = 8. However, the sequence doesn't run
>>> and it gives us "Max amplitude overflow on gradient z axis" after MB factor
>>> exceeds 2.  It appears to me that then we have to sacrifice TR and only run
>>> with MB = 2. Is there anybody with a Skyra who has able to achieve what we
>>> are hoping for and willing to share the protocol with us? OR if somebody
>>> could guide us to resolve the error of amplitude overflow.
>>>
>>> B) DWI: The idea was to have res = 1.5mm3, 2b-values at 1000 and 2500,
>>> 64 directions with the best BW and ES achievable. The sequence runs for 7
>>> minutes out of total 13 minutes and gives gradient power amplifier error.
>>> On Michael Harms's suggestion, I adjusted flip angle to 78/160 at a TR of
>>> 5500 ms with excite/refocus pulse duration of 3840 and 7680. there was a
>>> pop up warning that RF is clipped and the maximum refocusing angle was 142
>>> instead of 160. I placed a 64 directions diffusion vector and ran the
>>> sequence under "free" mode. Again, I will appreciate if  there is anybody
>>> with a Skyra who has able to achieve what we are hoping for and willing to
>>> share the protocol with us, along with the diffusion vector sets? OR if
>>> somebody could guide us to resolve the error.
>>>
>>> Thank you.
>>>
>>> Regards
>>>
>>> --VM
>>>
>>> _______________________________________________
>>> HCP-Users mailing list
>>> [email protected]
>>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>>
>>>
>>> ------------------------------
>>>
>>> The materials in this message are private and may contain Protected
>>> Healthcare Information or other information of a sensitive nature. If you
>>> are not the intended recipient, be advised that any unauthorized use,
>>> disclosure, copying or the taking of any action in reliance on the contents
>>> of this information is strictly prohibited. If you have received this email
>>> in error, please immediately notify the sender via telephone or return mail.
>>>
>>
>> ------------------------------
>>
>> The materials in this message are private and may contain Protected
>> Healthcare Information or other information of a sensitive nature. If you
>> are not the intended recipient, be advised that any unauthorized use,
>> disclosure, copying or the taking of any action in reliance on the contents
>> of this information is strictly prohibited. If you have received this email
>> in error, please immediately notify the sender via telephone or return mail.
>>
>> _______________________________________________
>> HCP-Users mailing list
>> [email protected]
>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>
>>
>> ------------------------------
>>
>> The materials in this message are private and may contain Protected
>> Healthcare Information or other information of a sensitive nature. If you
>> are not the intended recipient, be advised that any unauthorized use,
>> disclosure, copying or the taking of any action in reliance on the contents
>> of this information is strictly prohibited. If you have received this email
>> in error, please immediately notify the sender via telephone or return mail.
>>
>> _______________________________________________
>> HCP-Users mailing list
>> [email protected]
>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>
>
>
> ------------------------------
>
> The materials in this message are private and may contain Protected
> Healthcare Information or other information of a sensitive nature. If you
> are not the intended recipient, be advised that any unauthorized use,
> disclosure, copying or the taking of any action in reliance on the contents
> of this information is strictly prohibited. If you have received this email
> in error, please immediately notify the sender via telephone or return mail.
>

_______________________________________________
HCP-Users mailing list
[email protected]
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

Reply via email to