Re: Rietveld: U,V,W

2008-12-01 Thread Jim Cline

Hi,

Self Citation: NIST Standard Reference Materials for 
Characterization of Instrument Performance Industrial Applications 
of X-ray Diffraction Ed by F. H. Chung  D. K. Smith, pp 903-917.  A 
pdf is available; email to request.


In this I outline how I use SRM 660 (it was 660 in 2000, then it was 
660a until ~July '07, we are working on 660b) for qualifying and 
characterizing instrument performance with both profile fitting and 
the Rietveld method, via GSAS.  I suggest obtaining a set of TCH 
parameters with LaB6.  The instrument specific ones thereafter 
remained fixed while the other are used as a floor for refined 
values.  With this approach I was able to obtain stable and 
meaningful results from laboratory equipment.


Regards,

Jim

At 08:25 PM 11/30/2008, you wrote:

To all:

What is the correct procedure for refining U,V,W?  It is my 
understanding that those parameters are a function of instrument 
geometry.  Does one use a standard material to determine U,V,W and 
then fix their values for the instrument you're using?or do the 
values of U,V,W change depending on the sample being examined?  If 
so, why do the values change?


Thanks in advance.

Frank May



From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED]
Sent: Sun 11/30/2008 1:32 PM
To: rietveld_l@ill.fr
Subject: Re: I am a newcome, how can I begin my rietveld refinement analysis



Hi, Li:

To me, the most wonderful tool to determine initial peakshape parameters is
CMPR.
CMPR is especially oriented to GSAS and gives you GU, GV, GW etc. And when
you use EXPGUI for GSAS, you can also try Graphs-widplt to see how FWHM
develops when parameters are tuned.
When you prefer Fullprof, you should take a factor to get U, V, W etc. i am
not sure about the facor exactly, maybe GX~100X(X=U, V, W), er..? Anyway,
just go ahead and make a try.

Faithfully
Jun Lu
--
Lst. Prof. Lijie Qiao
Department of Materials Physics and Chemistry
University of Science and Technology Beijing
100083 Beijing
P.R. China
http://www.instrument.com.cn/ilog/handsomeland/


James P. Cline
Ceramics Division
National Institute of Standards and Technology
100 Bureau Dr. stop 8520 [ B113 / Bldg 217 ]
Gaithersburg, MD 20899-8523USA
[EMAIL PROTECTED]
(301) 975 5793
FAX (301) 975 5334 

Re: % Crystallinity

2008-02-28 Thread Jim Cline


Hi all,
Please prepare for a shameless plug for the newly certified NIST SRM
676a:
In order to measure amorphous content (via non-PDF diffraction methods)
one needs two things: 1) An unbiased diffraction method for measurement
of phase abundance (laughter please) and 2) A standard of known phase
purity. A quantitative Rietveld analysis of data from a divergent
beam powder diffractometer yields results that are consistently biased by
a few percent. The nature of the bias appears dependent on the
phases being measured and I don't really know of the origin. But by
process of elimination, I think it is due to an over simplistic model for
the Lorentz/polarization factor. Therefore, you are better off
using nonconventional sources, or accepting that your results may be
inaccurate to a few percent. For 10+ years I/we have been pursuing
an experimental design to measure absolute phase purity. It has
taken several iterations to obtain results I/we cared to talk
about. SRM 676 was recertified ~2 years ago for phase purity,
however, stocks of it were exhausted (the recertification is
retroactive).  SRM 676a was released for sale a week or two ago and
offers improvements in both the phase purity (99%) and the error bounds
on the certified value.
See:

https://srmors.nist.gov/view_detail.cfm?srm=676A
I haven't written the work up yet (I'm working on it); but the
experimental design is described in the certificate.
Regards,
Jim
At 06:58 AM 2/27/2008, you wrote:
Dear Users,
I'm trying to determine the percentage crystallinity in a
crystalline/amorphous mixture, could someone point me in a foolproof
direction.

Thanks,
John

Free upgrade for your Windows Live Messenger!
Click here! 

James P. Cline
Ceramics
Division


National Institute of Standards and Technology
100 Bureau Dr. stop 8520 [ B113 / Bldg 217 ]
Gaithersburg, MD 20899-8523 USA
[EMAIL PROTECTED]
(301) 975 5793
FAX (301) 975 5334




Opportunity at NIST

2008-02-20 Thread Jim Cline

Hi all,

We have a newly created position for an individual to work with us on 
the certification of NIST Standard Reference Materials for powder and 
thin film diffraction and reflectometry methods.  The desired 
background is one with theoretical leanings and competence in 
programing and data analysis methods.


Please don't hesitate to contact Don Windover, 301 975 6102, or 
myself with any questions.


Regards,

Jim

Official Solicitation:

CERAMICS DIVISION
MATERIALS SCIENCE AND ENGINEERING LABORATORY (MSEL)
NATIONAL INSTITUTE OF STANDARDS AND TECHNOLOGY (NIST)

ANALYST, X-RAY METROLOGY AND REFLECTROMETRY
STRUCTURE DETERMINATION METHODS GROUP

Applications are invited from qualified candidates for a research 
position involving analytical methods in X-ray reflectometry and 
reflectivity (XRR), high-resolution X-ray diffraction (HRXRD), and 
powder diffraction (XRPD) within the Structure Determination Methods 
Group, Ceramics Division, at NIST in Gaithersburg, Maryland.  The 
position is specific to the X-ray Metrology (XRPD and HRXRD) and XRR 
projects in the Structure Determination Methods Group, one of four 
groups in the Ceramics Division.  These projects provide the 
measurement science, Standard Reference Materials (SRMs), and 
advanced measurement methods needed by U.S. industry to apply XRR, 
HRXRD, and XRPD techniques across the entire spectrum of materials 
science and engineering. The Ceramics Division has built SI-traceable 
X-ray metrology instrumentation featuring advanced X-ray 
monochromator optics, optical angle encoders, and autocollimation techniques.


More information about other activities in the Ceramics Division may 
be found on our website, 
http://www.ceramics.nist.gov/http://www.ceramics.nist.gov/.


The research position will involve work with experimental data and 
physical and statistical theories of thin film analysis by XRR and 
HRXRD, and microstructure (size and strain) analysis by X-ray and 
neutron powder diffraction. Use of sophisticated statistical methods 
and theory will be required for parameter estimation and model 
comparison, e.g., Bayesian and Monte Carlo approaches.  Estimating 
uncertainties in parameters of physical models (particularly 
uncertainties across various models) from measurements pertaining to 
SRMs is of paramount interest.  The position will require research as 
part of a project team and participation in efforts with project, 
Group, and Division leadership to define new measurement directions.


Applicants should have experience in implementing data analysis or 
physical models in mainstream languages for scientific computing 
(e.g., Python, C, Fortran), preferably with code available for 
review. It is highly desirable that applicants have a Ph.D. in 
applied mathematics, computational physics, scientific computing, 
statistics, physics, materials science, or other related disciplines. 
Demonstrated completion of projects resulting in publication outputs 
is required. Experience with specialized environments for instrument 
control or data analysis (especially Labview, but also Mathematica, 
Matlab, R) is also valuable. A background in thin films, materials 
science, or dynamical diffraction or scattering would also be relevant.


The position is in the Scientific and Engineering Career Path (ZP) 
and is initially a two-year appointment, with potential for a 
permanent appointment. The base salary ranges from $70,000 to 
$110,000, depending on qualifications and experience.  United States 
citizenship is required.


Interested applicants should send a curriculum vita and three 
references (names and contact information only) via post or e-mail to 
Dr. Terrell Vanderah, Acting Leader, Structure Determination Methods 
Group, Ceramics Division, NIST, 100 Bureau Drive, Stop 8520, 
Gaithersburg, MD 20899, [EMAIL PROTECTED] Applications must 
be received by April 15, 2008.


NIST, a bureau of the U.S. Department of Commerce, is an Equal 
Opportunity Employer.



James P. Cline
Ceramics Division
National Institute of Standards and Technology
100 Bureau Dr. stop 8520 [ B113 / Bldg 217 ]
Gaithersburg, MD 20899-8523USA
[EMAIL PROTECTED]
(301) 975 5793
FAX (301) 975 5334  





Re: Sources of standard reference materials

2007-11-16 Thread Jim Cline

Hi all,

We expect to have SRMs 640d  660b available by no later then November 2008.

SRM 676a will available in a ~month.

I regret the difficulty that is being caused by this lapse; we are 
taking measures to prevent them from recurring.


Regards,

Jim

At 09:50 PM 11/15/2007, you wrote:

Hi All,

Does anyone know an appropriate source of standard reference materials for a
line profile standard, like NIST SRM 660a (LaB6) - other than NIST, as they
are out of stock and ...is expected to become available by November 2009.

Thanks,

Ross


James P. Cline
Ceramics Division
National Institute of Standards and Technology
100 Bureau Dr. stop 8520 [ B113 / Bldg 217 ]
Gaithersburg, MD 20899-8523USA
[EMAIL PROTECTED]
(301) 975 5793
FAX (301) 975 5334 





Re: SRM676a

2007-08-26 Thread Jim Cline

Hi All,  David,

SRM 676a will be available this fall.  It is of similar 
microstructure to SRM 676 and is certified with respect to phase 
purity.  It is, however, at 99%, of much higher purity than SRM 676.


Regards,

Jim

At 02:48 PM 8/21/2007, you wrote:

Hi All,

I'm looking for a source for NIST 676 alumina powder for use in
measuring RIRs for some
polymers.The NIST website states that they are out of stock.
Does anyone know of
other sources for this standard?

Thanks,
David Lee


James P. Cline
Ceramics Division
National Institute of Standards and Technology
100 Bureau Dr. stop 8520 [ B113 / Bldg 217 ]
Gaithersburg, MD 20899-8523USA
[EMAIL PROTECTED]
(301) 975 5793
FAX (301) 975 5334 


Re: RIET: NIST Si 640c cell axis as a function of temperature?

2005-07-05 Thread Jim Cline

Hi,

The temperature issue is one that I have been made to suffer on numerous 
occasions.  My operation was moved to the new NIST Advanced Measurement 
Laboratory that, among other things, features temperature control, in May 
of 2004.  The parallel beam diffractometer, from which future certification 
measurements will be obtained, is now located in a 0.01C space.  We 
measure the variation at 0.02C.  A divergent beam diffractometer is located 
in the vestibule for the 0.01C space that is controlled to 0.1C.  We 
haven't measured the variation in this space yet.  We are still in the 
process of commissioning this (largely custom) equipment; however, the 
parallel beam diffractometer is presently operational in a high-resolution 
configuration.  The first work on powders with the parallel beam 
diffractometer will include the certification of SRMs 640d, 660b and 1976a 
in 12-18 months.


At 10:52 PM 7/2/2005, you wrote:


Is there an official table of NIST 640c Silicon cell axes (with ESD)
for when NIST 640c is not exactly measured at 22.5 °C (say anywhere
between 5°C to 40°C)


Nope.


Using the certificate reference via:
 https://srmors.nist.gov/view_cert.cfm?srm=640C
 The certified lattice parameter for a temperature of 22.5 °C is
   0.54311946 nm ± 0.0092 nm

The NIST certificate mentions the thermal expansion coefficient which
was applied to the measured peaks.


The temperature variation during the data collection for SRMs 640c and 660a 
was +/- 1.5C.  The data, therefore, embody both a simple thermal expansion 
effect from the specimen itself, and a second, more complex one from the 
equipment.  The separation of the two is non-trivial.  We applied a simple 
thermal expansion correction to the data of the silicon, and, to a first 
approximation, did not see a instrument contribution.



For those who like their XRD calibrations and/or wavelengths backed up
all the way with a bit of NIST Onion Paper, it would be nice to have
such a table (or the range in which the stated thermal expansion
coefficient in the certificate (and certificate ESD) is valid).

(Some labs have different air-conditioning temperatures / or lack of
room temperature control).


Measurements at specific temperatures would require the construction of a 
furnace assembly that would have to embody some non-trivial 
engineering:  There could be no temperature differential between it and the 
goniometer assembly; thermal transfer between it and the atmosphere would 
have to be minimal ( 10 watts ); volumetric, homogeneous temperature 
control to ~.05C; and a sample spinner.  I'll think about it in earnest 
after the certification of the next series of line position SRMs (and 
perhaps a low angle line position SRM).


Best,

Jim



James P. Cline
Ceramics Division
National Institute of Standards and Technology
100 Bureau Dr. stop 8520 [ B113 / Bldg 217 ]
Gaithersburg, MD 20899-8523USA
[EMAIL PROTECTED]
(301) 975 5793
FAX (301) 975 5334  





Re: fundamental parameters approach X'celerator

2005-05-02 Thread Jim Cline
Arie,
Bob Cheary developed the required description/equations for the use of a 
psd.  But I don't know that they have been implemented in any of the 
available FPA codes, except perhaps Topas in launch mode.

Alan???  Are you listening?
Regards,
Jim
At 06:48 AM 5/2/2005, you wrote:
Dear colleagues,
Is there anybody who has experience with describing peak shapes
using the fundamental parameters approach for a typical X'pert-pro
diffractometer setup with an X'celerator detector (fixed incident
slits, 0.02 rad sollers, no monochromator)?
Thanks in advance, Arie
***
A. van der Lee
Institut Européen des Membranes (UMR 5635)
Université de Montpellier II - cc 047
Place E. Bataillon
34095 Montpellier Cedex 5
FRANCE
visiting address: 300 Av. Prof. E. Jeanbrau
tél.: 00-33-(0)4.67.14.91.35
FAX.: 00-33-(0)4.67.14.91.19
***
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA 




Re: fundamental parameters approach X'celerator

2005-05-02 Thread Jim Cline
Indeed...
A situation not without some complications.
Jim
At 08:00 AM 5/2/2005, you wrote:
Alan???  Are you listening?
Bruker providing help to analyze Panalytical data ?-).
Armel

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA 



Re: fundamental parameters approach X'celerator

2005-05-02 Thread Jim Cline
Laurel,
At 09:13 AM 5/2/2005, you wrote:
Hi,
Alan Coelho helped us get our instrument parameter file set up for Topas 
Academic.  We use an X'celerator detector with cobalt radiation (no 
monochromator) for most of our work.

Here is an example file:
xdd test.dat
CoKa7_Holzer(0.001)
Radius(240)
LP_Factor(17)
axial_conv
filament_length 12
sample_length 15
receiving_slit_length 12
primary_soller_angle @  5
This is curious: it is the basic criteria one would use for the full axial 
model of Topas.  It has nothing to do with a PSD.

Jim
Hope this helps.
Laurel Basciano
At 06:48 AM 5/2/2005, you wrote:
Dear colleagues,
Is there anybody who has experience with describing peak shapes
using the fundamental parameters approach for a typical X'pert-pro
diffractometer setup with an X'celerator detector (fixed incident
slits, 0.02 rad sollers, no monochromator)?
Thanks in advance, Arie
***
A. van der Lee
Institut Européen des Membranes (UMR 5635)
Université de Montpellier II - cc 047
Place E. Bataillon
34095 Montpellier Cedex 5
FRANCE
visiting address: 300 Av. Prof. E. Jeanbrau
tél.: 00-33-(0)4.67.14.91.35
FAX.: 00-33-(0)4.67.14.91.19
***
--
Laurel Basciano
Department of Geological Sciences and Geological Engineering
Queen's University
[EMAIL PROTECTED]
(613)484-1911

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA 




Re: fundamental parameters approach X'celerator

2005-05-02 Thread Jim Cline
Arie,
Yes; Bruker has an interest in this matter.
Jim
At 09:37 AM 5/2/2005, you wrote:
Bruker's Vantex detector is rather similar to PanAnalytical's
X'celerator detector from a fundamental parameters point of view,
not? Maybe I should have formulated my question differently
Is there anybody who knows how to model peak shapes using the
fundamental parameters approach for diffractometer setups with
the newly developped PSD's such as the Vantex and X'celerator?
Nobody will be vexed in this way, I hope at least 
Arie
ps: and thanks, Laurel, for the suggestion, I'll try that.
 Indeed...

 A situation not without some complications.

 Jim

 At 08:00 AM 5/2/2005, you wrote:

 Alan???  Are you listening?
 
 Bruker providing help to analyze Panalytical data ?-).
 
 Armel
 
 

***
A. van der Lee
Institut Européen des Membranes (UMR 5635)
Université de Montpellier II - cc 047
Place E. Bataillon
34095 Montpellier Cedex 5
FRANCE
visiting address: 300 Av. Prof. E. Jeanbrau
tél.: 00-33-(0)4.67.14.91.35
FAX.: 00-33-(0)4.67.14.91.19
***
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA 




Re: request

2005-04-20 Thread Jim Cline
Venkat,
Care to count up the number of requests you have made of the members of 
this Rietveld Listserv?  Is this an online library/database?

Jim
At 07:10 PM 4/20/2005 +0630, you wrote:
Thanks very much for bothering
+++
M Venkata Kamalakar
Junior Research Fellow,
S.N.Bose.National Centre for Basic Sciences,
Block-JD, Sector-3, Salt Lake,
Kolkata, Pin: 700 098.
Phone no: 033 23355705/6/7/8 Extn: 404, 104, 301.
+++
-- Original Message ---
From: Jim Cline [EMAIL PROTECTED]
To: rietveld_l@ill.fr
Sent: Wed, 20 Apr 2005 07:31:57 -0400
 Venkat,

 What journals does your institute subscribe to?

 Jim

 At 05:25 PM 4/20/2005 +0630, you wrote:
 Dear all,
 
 Can anyone please send me the following paper.
 
 Ferroelectrics
 Publisher:   Taylor  Francis
 Volume:   Volume 306 / 2004
 Pages:   95 - 109
 URL:   Linking Options
 DOI:   10.1080/00150190490458437
 
 Glass­Ceramic­Metal Nanocomposites Containing a Ferroelectric Phase
 
 
 A. DAN A1, T. K. KUNDU A2, B. SATPATI A3, P. V. SATYAM A3, D. CHAKRAVORTY
A4
 
 It very essential and our institute doesn't subscribe it.
 
 Thankyou
 
 venkat
 
 
 +++
 M Venkata Kamalakar
 Junior Research Fellow,
 S.N.Bose.National Centre for Basic Sciences,
 Block-JD, Sector-3, Salt Lake,
 Kolkata, Pin: 700 098.
 Phone no: 033 23355705/6/7/8 Extn: 404, 104, 301.
 +++

 James P. Cline  [EMAIL PROTECTED]
 Ceramics Division   Voice (301) 975
 5793
 National Institute of Standards and Technology  FAX (301) 975 5334
 100 Bureau Dr. stop 8520
 Gaithersburg, MD 20899-8523USA
--- End of Original Message ---
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA




Re: Size Strain in GSAS

2005-04-15 Thread Jim Cline

Hi,
In response to some of this post:
There was a move by a bunch of us in the ICDD to hold a profile fitting
round robin ( which I think would by quite useful ). But it died
when we realized the prodigious level of resources that would be required
to make sense of the rather large matrix of data that would
arrive.
But with regards to a round robin on this question: seems to me some
qualified individual could simply do the work and publish a nice paper on
it.
Regards,
Jim
At 12:30 PM 4/15/2005 +0200, you wrote:
urn:schemas-microsoft-com:office:office

Nicolae, Nick, Bob, Leonid,

I have looked at many patterns (recorded by others)
and a few cases have shown profiles that are sharper that a Lorentzian;
whereby sharper means that the integral breadth is smaller than that for
a unit area lorentzian. 

To put a figure on it would be difficult but at a
guess I would say  3% of patterns fall into this category in a
noticeable manner. 

I have no doubt that the work of Nick Armstrong and
co. is mathematically sound but a simulated data round robin as
suggested by Leonid Solovyov may be useful and I am not generally a fan
of round robins but this s different as the data is simulated. 

A pure peak fitting approach shows that two pV s (or
two Voigts) when added with different FWHMs and integrated intensities
but similar peak positions and eta values can almost exactly fit Pearsons
II functions that are sharper that Lorentzians. This is not surprising as
both profiles comprise 6 parameters. 

Thus from my observations two pVs added together can
fit a bimodal distributions quite easily. In fact my guess is that two
pVs can fit a large range of crystallite size distributions. 

Thus distinguishing whether a distribution is not
monomodal is of course possible especially if the two pV approach is
taken. 

Attempting to determine more than that however takes
some convincing as two pVs seem to fit almost anything that I have seen
that is symmetric. Thus introducing more pVs seems unnecessary.

Thus yes GSAS can determine if a distribution is not
monmodal if you were to fit two identical phases to the pattern except
for the TCH parameters. If the Rwp drops by .1% then I wont be
convinced.

Forgive me Nick but I have not yet read all of your
work and I am certain that it is sound. Outside of nano particles (and
maybe even inside) my reservation are that we may well be analyzing noise
and second order sample and instrumental effects.

Thus to show up my naive ness can you categorically
say that there are real world distributions that two pseudo Voigts cannot
fit because I have not come across such a pattern. 

Once you have done that then it would be time to
concentrate on strain, micro strain, surface roughness and then
disloactions

all the best

alan


-Original Message-
From: Nicolae Popa
[mailto:[EMAIL PROTECTED]] 
Sent: Friday, April 15, 2005 9:30 AM
To: rietveld_l@ill.fr
Subject: Re: Size Strain in GSAS

Dear Bob,

Perhaps I was not enough clear. Let me be more
explicit.
It's about one sample of CeO2 (not that from round-robin) that we
fitted in 4 ways.

(i) by GSAS with
TCH-pV
(ii) by another pV resulted from gamma distribution of
size
(iii) by Lorentz - (the limit of any regular pV -
eta=1)

All these 3 variants given bad fits. For example (ii): Rw=0.144,
similarly the rest. (In principle if one pV works (TCH for example) any
other kind of pV must work.)

(iv) by the profile resulted from lognormal distribution of
size; this time the fit was reasonably good: Rw=0.047. It resulted c=2.8,
that means a super Lorentzian profile (I remember that the
Lorentzian limit for lognormal is c~0.4 - JAC (2002) 35,
338).
Attention, this super Lorentzian profile is not
constructed as a pV with eta1.

Sure, such samples are rare, or, perhaps, not so rare. A Jap. group
(Ida,, Toraya, JAC (2003) 36, 1107) reported super
Lorentzian on a sample of SiC. They found c=1.37

Best wishes,
Nic Popa

Nic,
This is true for the internal math but the TCH function was assembled
to reproduce the true Voigt over the entire range of differing Lorentz
and Gauss FWHM values so it works as if the two FWHM components are
independent. As for your question, I'm not aware that anyone has actually
tried to do the fit both ways on a super Lorentzian (eta1
for old psVoigt) sample to see if a) the fit is the same and b) the
eta1 was an artifact. Any takers to settle
this?
Bob



R.B. Von Dreele

IPNS Division

Argonne National Laboratory

Argonne, IL 60439-4814


-Original Message-
From: Nicolae Popa
[mailto:[EMAIL PROTECTED]] 
Sent: Thursday, April 14, 2005 9:11 AM
To: rietveld_l@ill.fr
Subject: Re: Size Strain in GSAS

Dear Bob,

If I understand well, you say that eta1 (super Lorenzian)
appeared only because eta was free parameter, but if TCH is used super
Loreanzians do not occur?
Nevertheless, for that curious sample of cerium oxide we tried GSAS
(with TCH) and the fit was very bad.
Best wishes,
Nicolae

PS. By the way, TCH also forces 

RE: Size Strain [ In GSAS ??? ]

2005-04-14 Thread Jim Cline
Hi all,
Well, I thought I'd weigh in on this with a discussion of an aforementioned 
SRM project:

We are in the final stages of preparing an SRM for determination of 
crystallite size from line profile analysis.  Through the course of his PhD 
work and NIST postdoctoral position, Nick Armstrong has developed a MaxEnt 
/ Bayesian method specifically for the certification.  The method can 
quantify, from the quality of the raw data, the probably that a proposed 
model for the crystallite size distribution is the true one.  Thus, the 
certified values of the standard will include a valid measure of their 
uncertainty that, in our humble opinion, would not be obtainable with 
alternative methods.  Details, and results of the method as applied to the 
RR CeO2, have been published.  We are working at the production of ~kg 
quantities of strain free CeO2 and ZnO for use as the SRM feedstock; no 
small challenge.

We expect two outcomes:  1) The community will have a standard by which 
results from mortal methods may be readily tested and compared.   2) A 
high-intensity squabble will ensue as to whether or not we got the right 
answer.

 With regards to the latter issue: Nick and I have been approached about 
another round robin.  Forgive me, but: round robins don't have anything to 
do with accuracy.  They test for uniformity of measurements in the field, 
the major premise being that, as a result of mature methodology, a narrow 
distribution is expected.  Note the highly successful, Rietveld, QPA, and 
instrument sensitivity round robins.  There is, however, no mature 
methodology here.  Indeed, only a small number of operations worldwide can 
perform a credible microstructure analysis, and their methods certainly 
differ.  It is our intention to make the data collected for the 
certification available to the community.  But with regards to model 
testing for the more advanced, physical model, methods (with the use of 
simulated, bi-model data for instance), I would suggest that more of a 
collaborative effort be organized.

Regards,
Jim
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA



Re: Size Strain In GSAS

2005-03-28 Thread Jim Cline
Hi,
Nick Armstrong has advised me he will in non-email-land for a week or so.
I'm sure he'll resume this discussion when he returns...
Jim
At 03:45 PM 3/28/2005 +0400, you wrote:
Hi,
So, to resume your statements, by using Bayesian/Max.Entr. we can
distinguish between two distributions that can not be distinguished by
maximum likelihood (least square)?  Hard to swallow, once the restored peak
profiles are the same inside the noise. What other information than the
peak profile, instrumental profile and statistical noise we have that
Bayes/Max.ent. can use and the least square cannot?
prior distributions to be uniform - if I understand correctly you refer to
the distributions of  D0 and sigma of the lognormal (gamma) distribution
from which the least square chooses the solution, not to the distribution
itself (logn, gamm). Then, how is this prior distribution for Baye/Max.ent.?
Best,
Nick Popa
 Hi
 Sorry for the delay. The Bayesian results showed that the lognormal was
more probable. Yes, the problem is ill-condition which why you need to use
the Bayesian/Maximum entropy method. This method takes into account the
ill-conditioning of the problem. The idea being it determines the most
probable solutions from the set of solutions.  This solution can be shown to
be the most consistent solution or the solution with the least assumptions
given the experimental data, noise, instrument effects etc (see Skilling 
Bryan 1983; Skilling 1990; Sivia 1996). This is the role of entropy
function. There are many mathemaitcal proofs for this (see Jaynes' recent
book). The Bayesian analysis maps out the solution/model spaces.

 Also the least squares solution is simple a special case of a class of
deconvolution problems. This s well established result. It is not the least
ill-posed, since it assumes the prior distributions to be uniform (in a
Bayesian case. See Sivia and reference therein). In fact it's likely to be
the worst solution since it assumes a most ignorant state knowledge (ie.
uniform proir) and doesn't always take into consideration the surrounding
information. Moreover, it doesn't account for the underlying
physics/mathematics, that the probability distributions/line profiles are
positive  additive distributions (Skilling 1990; Sivia 1996).

 Best wishes, Nick


  Dr Nicholas Armstrong
snip
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA



Re: Size Strain In GSAS

2005-03-25 Thread Jim Cline
Hi,
I wrote an article [ that appeared in Dean Smith's book ] some time back 
that describes how to use SRM 660a, LaB6, and the TCH function of GSAS for 
characterization of the IPF, and then refine the only the microstructure 
specific terms for an estimation of the size and strain in subsequent 
analyses.  The approach has its limitations for sure, but it is accessible 
with a small effort.

Regards,
Jim
At 02:10 PM 3/25/2005 +0100, you wrote:
Dear Andreas,
I didn't said it cannot be done. Only that was not made for and so it 
is not easy as using other tools. In principle every diffraction fitting 
program can be used for size-strain.
Few questions: have you ever tried to do such analysis with GSAS the right 
way using the instrumental profile correction?
Did you use only GSAS for that or you had to use other external 
tools/computations? (this to get a feeling about your statement: Thus, on 
this level of line broadening analysis, everything necessary is contained 
in GSAS; may be level should be clarified).
I would like to know what there is inside GSAS for crystallite size and 
microstrain analysis in particular.

Best regards,
Luca Lutterotti
On Mar 25, 2005, at 12:55, Andreas Leineweber wrote:
Dear all,

I think the statement that one cannot do line-profile analysis using GSAS is
too strong. In principle it is possible to do some
size strain analysis using GSAS, if the instrumental profile is e.g.
sufficiently described previously
by the Thompson-Cox-Hastings (TCH) profile function (includes measuring
corresponding data on a suitable standard). I think, even involvement of the
Finger asymmetry correction does not introduce systematic errors. Then the
increase in the tantheta
and 1/costheta related Gaussian and Lorentzian line width components of 
the TCH
description upon Rietveld refinement on the basis of diffraction data
exhibiting physical line broadening can

in principle be associated with microstrain- and size-related quantities.
This can also be extended by involving anisotropic size and microstrain 
models.


Thus, on this level of line broadening analysis, everything neccessary is
contained in GSAS. Of course, something like microstrain and size 
distributions
cannot be obtained using GSAS.


Of course there are problems:
0. Microstrain broadening must be proportional to tan(theta). This is not
neccessarily the case. There mustn't be further line broadening contributions
like stacking faults, complicating the situation.
1. If both size and strain contributions are present, one most be aware 
of the
correlation between the tantheta and 1/costheta dependent compontents.

2. One has to be aware to which average values of the size and microstrain
distributions the increases of the line width parameters can be associated
with.
This requires for GSAS a close analysis of the GSAS manual (how are the
line-width paramerters defined!) and line broadening literature, how such
pseudo-Voigt line width parameters can be related with averages 
microstructure
parameters.


Thus it is not made easy for the user to extract something like that from the
line width parameters. But perhaps it is better that way, because 
consequently
the user is forced to to deal himself with the required theory, rather than
just refining a parameter called size and one called microstrain, 
believing in
the results and publish the values


Definitely there are much better procedures to analyse size and microstrain
than by GSAS.

So, going to the problem of Apu: If the instrumental profile is well 
described
before using TCH, refinement of (only) LX (and perhaps P) gives you 
quantities
whiuch you should be able to relate to size related quantities upon 
reading the
GSAS manual and some line broadening literature.


Best regards
Andreas Leineweber


Dear Prof. Lutterotti,

 I was also aware of the fact that GSAS is not made for Size Strain 
analysis.
I got interested to use the Size strain refinement feature of GSAS only after
going through the article :
Size-strain line broadening analysis of the ceria round-robin sample by
Prof. D. Balzar et. al. Journal of Applied Crys. 37(2004)911-924.

In that round robin results they have reported the size strain obtained from
GSAS.

I my case also when I am trying with GSAS, the diffraction pattern is 
fitting
well except the peak braodening. I think this brodening is due to small 
domain
size effect. I that case how will I obatin a good fit with GSAS.


Thanking you.

Best Regards,

Apu




/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/

Apu Sarkar

Research Fellow

Variable Energy Cyclotron Centre

Kolkata 700 064

phone: 91-33-2337-1230 (extn. 3190)

Fax:   91-33-2334-6871

INDIA

/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/


- Original Message -

From: Luca Lutterotti [EMAIL PROTECTED]

Date: Friday, March 25, 2005 3:31 pm

Dear Apu,

I know I will start up a good debate here, but size-strain

analysis

with GSAS is a 

Re: request

2005-03-23 Thread Jim Cline
Hi all,
Its getting to the point that 75% of the posts on this list are requests 
for structures and articles???

Inappropriate.
Jim
At 11:33 AM 3/23/2005 +0530, you wrote:
Dear all,
Can anyone please send me the following article.
snip
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA



RE: Enquiry of XRD standard materials (fwd)

2004-12-07 Thread Jim Cline
Hi,
At 02:09 PM 12/7/2004 +, you wrote:
 Dear Stephen,
 I have prepared XPD-standard Y2O3 by firing Y2O3 (99.999) from Aldrich
 at 1200 ºC for 72 hs. (as done in 41-1105 PDF card).
 I would also like to ask the other list members if anyone has calibrated
 the size-strain broadening of Y2O3 against LaB6, because I have no
 access to LaB6 and I've heard that yttria gives broader lines.
 Thanks
 Leopoldo
From Y2O3 annealed at around 1200 degC, and using fundamental
parameters peak fitting in XFIT (BGMN and Topas are currently
maintained programs), I get around 3300 Angstrom for crystallite
size - and near zero strain (using the size/strain modelling
defaults in XFIT).  Though don't have a comparison with LaB6.
What is the certificate crystallite size and strain value for
NIST LaB6 - via Google, the NIST website seems to be giving a
problem finding this out?
http://ois.nist.gov/srmcatalog/certificates/view_cert2pdf.cfm?certificate=660a 
Error Occurred While Processing Request
  Error Diagnostic Information
  An error has occurred.
  HTTP/1.0 404 Object Not Found

Lachlan
??? Works for me:
https://srmors.nist.gov/view_cert.cfm?srm=660A
We estimate the crystallite size of SRM 660a to be 2 microns, and could not 
find any strain broadening, from an FPA (Topas) analysis of conventional, 
lab XRD data.  We have compared SRM 660a to the super NAC at ESRF/BM16; 
but results are qualitative.  We would, of course, like to quantify the 
broadening of SRM 660a, but the project is a bit of a can-of-worms that 
will have to wait, perhaps until the certification of SRM 660b in ~2 years.

Regards,
Jim
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA




Re: back loading

2004-09-10 Thread Jim Cline

Hi,
At 08:38 AM 9/10/2004 +0200, you wrote:
Hello Andreas,

When I met Jim Cline at the Denver Conference he spoke about
a fluid to drop into a (possibly backloaded) powder sample in order to
make it solid. I will not put Jim's e-mail address on the Rietveld list
but am quite sure he is subscribed - Jim, could you advise? 

Me??? But I was having so much fun in lurk mode...
There exists a single component, moderate viscosity, low wetting angle
epoxy designed and sold for quick fixes to leaky vacuum equipment.
The stuff is supposed to seek into and fill cracks, then harden into a
semi-flexible seal. I have found that a very small quantity of it
can be used to stabilize compacts of powders as, due to the low wetting
angle, it will fully infiltrate such a compact before hardening. I
have not found that it has any impact on the measured background.
However, in order to employ it for this, back loading, application, one
would have to come up with a smooth surface to which the epoxy would not
adhere. I would certainly try Teflon, but some experiments are in
order.
Best regards, 
Jim
snip

James P.
Cline[EMAIL PROTECTED]
Ceramics
DivisionVoice
(301) 975
5793
National Institute of Standards and
TechnologyFAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523 USA



Re: Quantitive analysis

2004-05-06 Thread Jim Cline
Hi all,
This is to say that I'm am nearly finished with the re-certification of SRM 
676 for amorphous content.  The revised certificate, offering a discussion 
as to the certification method, will be on the NIST website shortly, after 
the review process is completed.  The certified phase purity of SRM 676 
(retroactive to units in the field) is 91.75% +/- 1.5%.

Regards,
Jim
At 01:44 PM 5/6/2004 -0400, you wrote:
Robert
As previously mentioned many Rietveld programs will do what you want if
you're willing to do the amorphous content calculations by hand afterwards.
Some commercial software has it built in.
I've done work in this area, and there are some pitfalls of which you must
be aware, or your results may be VERY unreliable.
1.  microabsorption.  If your phases are similar in composition and your
'spike' phase has a similar absorption then this might not be a problem.
However, there are instances in which extreme microabsorption can make a
accurate determination practically impossible (other than possibly changing
your tube to a more friendly wavelength).  There is the Brindley correction
that can sort of correct moderate microabsorption, but you need to know your
particle size, and ideally it needs to have a narrow distribution.  Reducing
the particle size of your sample to sub-micron levels helps significantly.
2. spike phase.  Many so-called crystalline phases aren't as fully
crystalline as you might think.  The SRM676 alumina had it's amorphous
content measured a while back using neutrons and came out as 1.77 +- 0.68%.
It doesn't sound alot but can make a big difference to your final results.
Many 'crystalline' phases can be significantly higher.  Microabsorption can
be an issue with the spike so choose your spike carefully, and ideally
standardise it using SRM676.  Remember to include the error in the SRM676
amorphous content if quoting absolute amorphous contents rather than
relative amounts.
3. particle statistics.  This is equally as important when looking at purely
crystalline materials, but errors due to poor particle statistics can really
mess up amorphous content work.  Ideally, micronise your sample before
running it, but if they're nanomaterials then it might not be an issue.
4.  surface roughness.  I noticed that you probably work with nanomaterials.
They tend to be fluffy, and such samples (at least on our instrument) show
noticeable surface roughness effects on low angle reflections. If this is
the case for your samples then you will either have to correct it or densify
your sample somehow (or both).
Some of this stuff was covered in the results of the quantitative analysis
round robin a while back:
Outcomes of the International Union of Crystallography Commission on Powder
Diffraction Round Robin on Quantitative Phase Analysis: samples 1a to 1h
I. C. Madsen, N. V. Y. Scarlett, L. M. D. Cranswick and T. Lwin, J. Appl.
Cryst. (2001). 34, 409-426
Outcomes of the International Union of Crystallography Commission on Powder
Diffraction Round Robin on Quantitative Phase Analysis: samples 2, 3, 4,
synthetic bauxite, natural granodiorite and pharmaceuticals
N. V. Y. Scarlett, I. C. Madsen, L. M. D. Cranswick, T. Lwin, E. Groleau, G.
Stephenson, M. Aylmore and N. Agron-Olshina
J. Appl. Cryst. (2002). 35, 383-400
It would be useful to cross-check your technique on a couple of dummy
samples if at all possible to make sure you're not being mislead.  Have
fun!!
Pam

Dr Pamela Whitfield CChem MRSC
Energy Materials Group
Institute for Chemical Process and Environmental Technology
Building M12
National Research Council Canada
1200 Montreal Road
Ottawa  ON   K1A 0R6
CANADA
Tel: (613) 998 8462 Fax: (613) 991 2384
Email: mailto:[EMAIL PROTECTED]
ICPET WWW: http://icpet-itpce.nrc-cnrc.gc.ca
-Original Message-
From: Robert Mauricot [mailto:[EMAIL PROTECTED]
Sent: May 6, 2004 10:57 AM
To: [EMAIL PROTECTED]
Dear all,
I have a powder diffraction of an amorphous multiphase material. I want to
do a quantitative analysis of the amorphous phase.  Can anyone tell me what
software does it.
Sincery
R.mauricot
Cemes-CNRS
Toulouse 31077 France
Laboratoire  NaNomat
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 
5793
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8520
Gaithersburg, MD 20899-8523USA



Re: Quantitive analysis

2004-05-06 Thread Jim Cline
Hi,
At 03:48 PM 5/6/2004 -0400, you wrote:
Jim
Excellent news!  This amorphous standard issue has been driving me nuts.
When you say phase purity do you mean purity in terms of crystalline phase
content?
Yes.
Jim
Pam
Dr Pamela Whitfield CChem MRSC
Energy Materials Group
Institute for Chemical Process and Environmental Technology
Building M12
National Research Council Canada
1200 Montreal Road
Ottawa  ON   K1A 0R6
CANADA
Tel: (613) 998 8462 Fax: (613) 991 2384
Email: mailto:[EMAIL PROTECTED]
ICPET WWW: http://icpet-itpce.nrc-cnrc.gc.ca
-Original Message-
From: Jim Cline [mailto:[EMAIL PROTECTED]
Sent: May 6, 2004 3:43 PM
To: [EMAIL PROTECTED]
Hi all,
This is to say that I'm am nearly finished with the re-certification of SRM
676 for amorphous content.  The revised certificate, offering a discussion
as to the certification method, will be on the NIST website shortly, after
the review process is completed.  The certified phase purity of SRM 676
(retroactive to units in the field) is 91.75% +/- 1.5%.
Regards,
Jim
At 01:44 PM 5/6/2004 -0400, you wrote:
Robert

As previously mentioned many Rietveld programs will do what you want if
you're willing to do the amorphous content calculations by hand afterwards.
Some commercial software has it built in.
I've done work in this area, and there are some pitfalls of which you must
be aware, or your results may be VERY unreliable.

1.  microabsorption.  If your phases are similar in composition and your
'spike' phase has a similar absorption then this might not be a problem.
However, there are instances in which extreme microabsorption can make a
accurate determination practically impossible (other than possibly changing
your tube to a more friendly wavelength).  There is the Brindley correction
that can sort of correct moderate microabsorption, but you need to know
your
particle size, and ideally it needs to have a narrow distribution.
Reducing
the particle size of your sample to sub-micron levels helps significantly.

2. spike phase.  Many so-called crystalline phases aren't as fully
crystalline as you might think.  The SRM676 alumina had it's amorphous
content measured a while back using neutrons and came out as 1.77 +- 0.68%.
It doesn't sound alot but can make a big difference to your final results.
Many 'crystalline' phases can be significantly higher.  Microabsorption can
be an issue with the spike so choose your spike carefully, and ideally
standardise it using SRM676.  Remember to include the error in the SRM676
amorphous content if quoting absolute amorphous contents rather than
relative amounts.

3. particle statistics.  This is equally as important when looking at
purely
crystalline materials, but errors due to poor particle statistics can
really
mess up amorphous content work.  Ideally, micronise your sample before
running it, but if they're nanomaterials then it might not be an issue.

4.  surface roughness.  I noticed that you probably work with
nanomaterials.
They tend to be fluffy, and such samples (at least on our instrument) show
noticeable surface roughness effects on low angle reflections. If this is
the case for your samples then you will either have to correct it or
densify
your sample somehow (or both).


Some of this stuff was covered in the results of the quantitative analysis
round robin a while back:

Outcomes of the International Union of Crystallography Commission on Powder
Diffraction Round Robin on Quantitative Phase Analysis: samples 1a to 1h
I. C. Madsen, N. V. Y. Scarlett, L. M. D. Cranswick and T. Lwin, J. Appl.
Cryst. (2001). 34, 409-426

Outcomes of the International Union of Crystallography Commission on Powder
Diffraction Round Robin on Quantitative Phase Analysis: samples 2, 3, 4,
synthetic bauxite, natural granodiorite and pharmaceuticals
N. V. Y. Scarlett, I. C. Madsen, L. M. D. Cranswick, T. Lwin, E. Groleau,
G.
Stephenson, M. Aylmore and N. Agron-Olshina
J. Appl. Cryst. (2002). 35, 383-400


It would be useful to cross-check your technique on a couple of dummy
samples if at all possible to make sure you're not being mislead.  Have
fun!!

Pam



Dr Pamela Whitfield CChem MRSC
Energy Materials Group
Institute for Chemical Process and Environmental Technology
Building M12
National Research Council Canada
1200 Montreal Road
Ottawa  ON   K1A 0R6
CANADA
Tel: (613) 998 8462 Fax: (613) 991 2384
Email: mailto:[EMAIL PROTECTED]
ICPET WWW: http://icpet-itpce.nrc-cnrc.gc.ca


-Original Message-
From: Robert Mauricot [mailto:[EMAIL PROTECTED]
Sent: May 6, 2004 10:57 AM
To: [EMAIL PROTECTED]


Dear all,

I have a powder diffraction of an amorphous multiphase material. I want to
do a quantitative analysis of the amorphous phase.  Can anyone tell me what
software does it.

Sincery

R.mauricot


Cemes-CNRS
Toulouse 31077 France
Laboratoire  NaNomat
James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975
5793
National Institute of Standards

Re: LaAlO3 structure

2001-03-12 Thread Jim Cline

Joerg,

At 07:07 AM 3/7/01 +, you wrote:
snip
We have done some investigations on the old line standard LaB6 SRM660,
it contains some minor phases (sum 5%), one was LaAlO3. The new
standard SRM660a is clean and has some other advantages, too. Now,
we favour SRM660a over Y2O3.

This is a pleasant report to hear!  Will I see you at Accuracy in Powder 
Diffraction III?

Best regards,

Jim


James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793   
 
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523USA



Re: SRMs 640c 660a

2000-09-29 Thread Jim Cline

At 04:42 PM 9/29/00 +0200, you wrote:
Jim wrote:

 The hydrostatic stress level is inversely
  proportional to the radius of curvature of the particles.  Therefore, the

Exactly, how did you determine the trend? Have you measured the cell
parameters for different granulometries?

This view of the matter came from consideration of surface tension effects 
on droplets.  The hydrostatic compressive stress on a droplet due to 
surface tension is inversely proportional it's size.  There is 
circumstantial evidence in the literature on the observance of a reduction 
in lattice parameter with a reduction in crystallite size.  But I don't 
know that anyone has actually set out to due a systematics study of this 
effect.  It does, however, follow that this purely geometric effect will 
operate in reverse, i.e., surface compression for surface tension and 
hydrostatic tension for hydrostatic compression.

Last curiosity. The old SRM 640b was having a certified cell parameter
slightly smaller than the crystal used for 640c. Do you think that in this
case it was not measured correctly the "absolute value" or it "was"
different?

Interesting question: The certification methods for the two SRMs were 
completely different (this is why SRM 640c has been so long in 
arrival).  Therefore, it is difficult to say why the two certified cell 
parameters are different.  I may look into this in the future.  I am 
building a new diffractometer for certification of SRMs and this is a 
problem to which it will be well suited.

Sorry for the question, but I personally and philosophically don't think we
can measure absolute values..

I disagree; and I enjoy my job too!

Regards,

Jim

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793   
 
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523USA



SRMs 640c 660a

2000-09-27 Thread Jim Cline

Hi all,

This is to indicate that NIST SRMs 640c, silicon powder, and 660a, 
lanthanum hexaboride powder, are available from the NIST SRM Program sales 
office.  These SRMs are certified with respect to lattice parameter at the 
part-per-million level.  Considerable effort has been expended to tailor 
the microstructure of these SRMs to exhibit a minimal level of particle 
size and micro-strain induced profile broadening.

Please see:

http://ois.nist.gov/srmcatalog/certificates/view_cert2gif.cfm?certificate=640c

and

http://ois.nist.gov/srmcatalog/certificates/view_cert2gif.cfm?certificate=660a

for views of the certificates which offer a description of the material and 
the certification method.

Regards,

Jim

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793   
 
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523USA

**
NIST Standard Reference Materials Program

Phone:  301 975 6776
FAX:301 948 3730
email:  [EMAIL PROTECTED]
Home Page:  http://www.nist.gov/srm
  Please see home page for copies of NIST SRM certificates. 
**



Re: mirror vs. Ge monochromator.

2000-07-07 Thread Jim Cline

Hi,

At 08:46 AM 7/7/00 +0100, Joe Hriljac wrote:
When we faced the same decision (well, only one machine, UK oil money does
snip

I agree with Joe: I have a D500 with a Ge monochromator and a PSD.  I don't 
like the metal wire PSD as much as the quartz wire one as it imparts noise 
to the data in excess of expected counting statistics but the machine 
offers high resolution and stability with no alpha 2 or other junk.  If the 
machine is properly aligned and samples are properly mounted than one can 
count on predicable delta d alignment curves.  I believe the jury is still 
out on high resolution / high accuracy data from mirror optics.  They do 
offer "freedom" from sample displacement errors; but there are additional 
complications.  I'll be speaking on this at Denver.

At 10:27 AM 7/7/00 +0200, Stan Gierlotka wrote:
1) Ge monochromator + PSD - no Ka2, extremely good counting statistics,
but somewhat broadened and uncertain peak shapes
2) Geobel mirror + LiF analyzer crystal - nice peak shapes down to very
low angles, very reliable peak positions but long counting times.

I disagree:  If you've got "uncertain" peak shapes from your Ge 
monochromator than you've got uncertainty somewhere in your 
mounting/alignment of it.  The Ge crystal can resolve the alpha 1 from the 
alpha 2 without distortion; however, you do need a slit at the focal point 
(line) to block scatter (the instructions supplied with the Siemens D500 
setup don't mention this but the system is designed to allow it).  The 
mirror with an LiF analyzer is something I'd like to try out.

Regards,

Jim

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793   
 
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523USA



RE: Si lattice parameter

2000-04-17 Thread Jim Cline

Hi,

At 05:26 PM 4/17/00 +0200, you wrote:
Dear All,

I would like to add the following question concerning the standard Si  powder.
Is there someone who has information on the grain size of the 640b Si 
material ?
snip

The mean particle size of SRM 640b, via laser scattering, is 7.3 
microns.  However, the distribution is fairly broad with the fines tail 
extending well into the submicron region.

Regards,

Jim

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793   
 
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523USA



Re: Kalpha2

2000-04-03 Thread Jim Cline

Hi,

At 07:31 PM 4/1/00 +0200, you wrote:
 I'm looking forward for many discussions!

Yep!

I have some naive questions :

I would like to hear exactly what is on my powder pattern:
  - Kalpha3
  - tube tails
  or both ??

The paper by G. H"olzer et al. (Phys. Rev. A 56 (1997) pp. 4554 ff.) is (at 
present) the definitive paper on emission spectra, and they do not discuss 
the presence of the "Kalphs3".  However, I believe it is referenced 
elsewhere (I'll look into it) and, as I have said, it can be readily seen 
in the experiment I described previously (anyone who wants a .pdf plot can 
email me).  I would like to see (or do) the single crystal experiment in 
order to definitively characterize the "tube tails".  I suspect that the 
answer to the Armel's question is both.

Do these effects would occur at the same angular value ?

Nope; the "Kalpha3" is observed on the low angle side of the profile while 
the "tube tails" are observed on both sides.

Regards,

Jim

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793   
 
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523USA



Re: Kalpha2

2000-04-03 Thread Jim Cline

Hi,

Verification
of tube tails with a single crystal experiment has been obtained: I
called Keith Bowen of Bede on this thread as I thought he would be
interested in this matter and I know he is knowledgeable on electron beam
focusing. [See the Bede web site and look for Microsource.]
He pointed out that the distribution of Xray sources across the anode
could be imaged with a Bartels monochromator and the scanning could be
done with a suitably small detector slit. In an hour the experiment
was complete.

From Keith:
The detector slit was 50
mm, at
750 mm from the source, and could be scanned across the beam. The
monochromator has an output divergence of 5 arc seconds (2.5 x 10-5
radians) so will give a blurring of about another 20
mm,
providing a spatial resolution overall of about 70
mm.
The tube was a Siemens ceramic K FN Cu 2K, run at full power, with
nominal focus 10 x 1 mm. It was run in vertical line mode with a takeoff
angle of 6. The maximum count was 17,500 cps and lobes were
observed at about 15 cps [as shown in figures he sent me]. The
lobes were not symmetric [they span a detector Z range of -0.4 to +0.6mm
with the central focal line from -0.15 to +0.15mm]. These lobes are
lower than reported by Joerg Bergmann in
www.bgmn.de/tubetails.html,
but also the tube is a more modern design. 


1/sin(6 [deg take off]) X 0.5 mm [detector Z] = 4.78 mm: The
origin of the X-rays of one lobe is ~5 mm from the focal line.
While the count rate is low (for the narrow spectral window of the
monochromator), this is not such good focusing. Somebody should
invent a better mousetrap!

Regards,

Jim

James P.
Cline[EMAIL PROTECTED]
Ceramics
DivisionVoice
(301) 975
5793
National Institute of Standards and
TechnologyFAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523 USA




Re: Kalpha2

2000-03-28 Thread Jim Cline

Hi,

At 04:07 PM 3/28/00 +0100, you wrote:
Hi all,

On Tue, 28 Mar 2000 16:13:11 +0200, Armel Le Bail wrote:

I have a related question about the so-called Kalpha3.
Trying to take account of it, I only obtain some slight 
increase
in the Rp and Rwp, though the fit seems to have improved,
at first glance.

Has someone a powder pattern available which shows clearly
Kalpha3, and which he was able to fit by a Rietveld code
(which one ?) and obtained some improvement in the Rp
and Rwp values ?

I have no hints about Kalpha3. But some related hint: Several years
ago,
we were believing for Kalpha3 or some related satellite lines of the
Kalpha
doublet really to disturbe our fundamental parameters lines. In the
beginning of 1999, we found out the so-called tube-tails
(which better
should be called focus tails). For example, have a look
at

www.bgmn.de/tubetails.html
---snip---

This is interesting; I've not seen it (nor have I looked for it).


I have been investigating the Kalpha3 matter in the course of
the certification of SRMs 640c (silicon) and 660a (LaB6) using the
FPA code Topas. I'll reporting on this at EPDIC7.

Regards,

Jim

James P.
Cline[EMAIL PROTECTED]
Ceramics
DivisionVoice
(301) 975
5793
National Institute of Standards and
TechnologyFAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523 USA




Re: Kalpha2

2000-03-28 Thread Jim Cline

Hello,

At 09:20 PM 3/28/00 +0200, you wrote:
On the point of view of theory, can the tube-tails and
Kalpha3 satellite explanations be reconciled ? No tube-tails
theory in the most recent X-ray powder Diffractometry book
from Jenkins and Snyder, only Kalpha1, 2, 3, 4, 5, 6.

Don't think so; "Kalpha3" is a component of the emission spectrum of copper 
while the "tube tails" appears to be an artifact of poor electron beam 
focusing within the X-ray tube.

Jim

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793   
 
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523USA



Re: Kalpha2

2000-03-28 Thread Jim Cline

Hello II,

At 09:20 PM 3/28/00 +0200, you wrote:
On the point of view of theory, can the tube-tails and
Kalpha3 satellite explanations be reconciled ? No tube-tails
theory in the most recent X-ray powder Diffractometry book
from Jenkins and Snyder, only Kalpha1, 2, 3, 4, 5, 6.

I went back to the "tube tails" web site mentioned by J"org Bergmann: Has 
anyone done a single crystal rocking curve in an attempt to see these 
tails?  I have just done one to characterize the emission spectrum (as 
transmitted through a graphite post monochromator).  I observed no "tube 
tails".  However, I wasn't looking for them; the crystal was ~27mm in size 
and scan range for the Si 400 line was 68-70.5 two-theta and the instrument 
(a D5000) was set up with the smallest divergence and receiving slits and 
both Soller collimators.  These parameters may not be suitable for 
observance of any (and all) spurious x-rays from the anode.  Some geometric 
exercises may be needed to design an experiment such that any radiation 
originating from the anode would be diffracted by the single crystal into 
an anxious detector.  The "Kalpha3" is apparent in this scan and its 
incorporation into the emission spectrum used for an FPA will address the 
low angle "oddities" shown in the profile of LaB6 (figure a) of the web 
site.  However; the high angle debris remains a problem.

Regards,

Jim

James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793   
 
National Institute of Standards and Technology  FAX (301) 975 5334
100 Bureau Dr. stop 8523
Gaithersburg, MD 20899-8523USA



Re: GSAS profile coefficients

1998-11-09 Thread Jim Cline

Hi,

At 12:27 PM 11/9/98 +1100, you wrote:
At 05:24 PM 03-11-98 -0500, you wrote:
At 08:26 PM 11/2/98 EST, you wrote:
Hi,
My questions pertain to refining the profile coefficients.  What order do
snip1
Thanks
MIZ

 Dear MIZ,

The question sounds borring, but is of great importans for everyone, using
-snip2

Sr. Petrov


Are you suggesting that one tackle the profile fit BEFORE refining atom
positions, site occupancies, and thermal parameters?
Dr Sue Kesson


I have developed and written up a strategy addressing the refinement of
profile coefficients for Rietveld analysis of conventional X-ray powder
data using GSAS.  Its basic premise is the separation of the instrumental
terms from the sample related ones and involves the use of SRMs for
verification of instrument performance (surprised?).  The manuscript is
part of an update to the workshop notes distributed at "The Design,
Alignment, Calibration and Performance Characteristics of the Conventional
Laboratory Diffractometer" presented at Denver last year by myself and Bob
Cheary.  The manuscript will also appear in a book "Industrial Applications
of X-ray Diffraction" edited by D. Smith and F. Chung.  I can send the
manuscript as a Word97 attachment upon request.  I can also sent out an
full updated a hardcopy of the workshop notes in about a month as we (Bob
and I) are working on an update at this time.

Regards,

Jim


James P. Cline  [EMAIL PROTECTED]
Ceramics Division   Voice (301) 975 5793
A256/223FAX (301) 975 5334
National Institute of Standards and Technology
Gaithersburg, MD 20899USA