Re: [ccp4bb] relationship between B factors and Koff

2010-11-22 Thread Herman . Schreuder
Although Bfactors and occupancies are correlated, occupancy refinement is quite 
feasible for the following reasons:

1) Modern maximum likelyhood programs like Buster do a decent job in separating 
occupancies and Bfactors. It shows up as shells of negative difference density 
e.g. around partially occupied metal ions which have been refined with full 
occupancy and inflated Bfactors.
2) A ligand is either completely there or it is completely gone, i.e. it is not 
that one could refine a group occupancy, one should refine a group occupancy 
and one has only one parameter to refine for say a ligand of 10+ atoms. In my 
experience, refining group occupancies even at moderate resolution, is stable.

I now refine by default (group) occupancies for all ligands. These include the 
inhibitor soaked in, but also items like bound sulfate and phosphate ions, 
Tris, Hepes molecules and bound metal ions like zinc or cadmium. Only waters I 
refine at full occupancy and let the Bfactors soak up any partial occupancy and 
other errors since here the data do not allow occupancy refinement. 

Best,
Herman

 

-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of James 
Holton
Sent: Friday, November 19, 2010 9:20 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] relationship between B factors and Koff

I don't think there is any relationship between rate constants and B factors.  
Yes, there is the hand-wavy argument of disorder begets disorder (and people 
almost always LITERALLY wave their hands when they propose this), but you have 
to be much more careful than that when it comes to thermodynamics.  Yes, 
disorder and entropy are related, but just because a ligand is disordered 
does not mean that the delta-S term of the binding delta-G is higher.

Now, there is a relationship between equilibrium constants and occupancies, 
since the occupancy is really just the ratio of the concentration of 
protein-bound-to-ligand to the total protein concentration, and an 
equilibrium exists between these two species.  NB all the usual caveats of how 
crystal packing could change binding constants, etc.  You could logically 
extend this to B factors by invoking a property of refinement:  Specifically, 
if the true 
occupancy is less than 1, but modeled as 1.00, your refinement program will 
give you a B factor that is larger than the true atomic B factor.  However, 
if you try to make this claim, then the obvious cantankerous reviewer 
suggestion would be to refine the occupancy.  
Problem is, refining both occupancy and B at the same time is usually unstable 
at moderate resolution.  In general, it is hard to distinguish between 
something that is flopping around (high B factor) and something that is simply 
not there part of the time (low occupancy).

I know it is tempting to try and relate B factors directly to entropy, but the 
disorder that leads to large B factors has a lot more to do with crystals 
than it has to do with proteins.  For example, there are plenty of 
tightly-bound complexes that don't diffract well at all.  If you refine these 
structures, you will get big B factors (roughly, B =
4*d^2+12 where d is the resolution in Angstrom).  You may even have several 
crystal forms of the same thing with different Wilson B factors, but that is in 
no way evidence that the proteins in the two crystal forms somehow have 
different binding constants or rate constants.

On the bright side, in your case it sounds like you have an entire protomer 
that is disorered relative to the rest of the crystal lattice.  We have seen 
a few cases now like this where dehydrating the crystal (with an FMS or similar 
procedure) causes the unit cell to shrink and this locks the wobbly molecule 
into place.  I think this is the principle mechanism of improved diffraction 
from dehydration.  No, it does not work very often!  But sometimes it does.

-James Holton
MAD Scientist

On 11/19/2010 4:58 AM, Sebastiano Pasqualato wrote:
 Hi all,
 I have a crystallographical/biochemical problem, and maybe some of you guys 
 can help me out.

 We have recently crystallized a protein:protein complex, whose Kd has been 
 measured being ca. 10 uM (both by fluorescence polarization and surface 
 plasmon resonance).
 Despite the 'decent' affinity, we couldn't purify an homogeneous complex in 
 size exclusion chromatography, even mixing the protein at concentrations up 
 to 80-100 uM each.
 We explained this behavior by assuming that extremely high Kon/Koff values 
 combine to give this 10 uM affinity, and the high Koff value would account 
 for the dissociation going on during size exclusion chromatography. We have 
 partial evidence for this from the SPR curves, although we haven't actually 
 measured the Kon/Koff values.

 We eventually managed to solve the crystal structure of the complex by mixing 
 the two proteins (we had to add an excess of one of them to get good 
 diffraction data).
 Once solved the structure

Re: [ccp4bb] relationship between B factors and Koff

2010-11-22 Thread George M. Sheldrick

I agree that refining a single occupancy parameter for the ligand is 
a good approach. Because of the correlation with the B-values, it is 
essential to refine occupancy and isotropic or anisotropic B-values
together, not in separate steps. SHELX has been able to do this
correctly for the last 40 years. One factor that can affect the
results is that if the ligand isn't there, something else (e.g.
water) probably is.

George

Prof. George M. Sheldrick FRS
Dept. Structural Chemistry,
University of Goettingen,
Tammannstr. 4,
D37077 Goettingen, Germany
Tel. +49-551-39-3021 or -3068
Fax. +49-551-39-22582


On Mon, 22 Nov 2010, herman.schreu...@sanofi-aventis.com wrote:

 Although Bfactors and occupancies are correlated, occupancy refinement is 
 quite feasible for the following reasons:
 
 1) Modern maximum likelyhood programs like Buster do a decent job in 
 separating occupancies and Bfactors. It shows up as shells of negative 
 difference density e.g. around partially occupied metal ions which have been 
 refined with full occupancy and inflated Bfactors.
 2) A ligand is either completely there or it is completely gone, i.e. it is 
 not that one could refine a group occupancy, one should refine a group 
 occupancy and one has only one parameter to refine for say a ligand of 10+ 
 atoms. In my experience, refining group occupancies even at moderate 
 resolution, is stable.
 
 I now refine by default (group) occupancies for all ligands. These include 
 the inhibitor soaked in, but also items like bound sulfate and phosphate 
 ions, Tris, Hepes molecules and bound metal ions like zinc or cadmium. Only 
 waters I refine at full occupancy and let the Bfactors soak up any partial 
 occupancy and other errors since here the data do not allow occupancy 
 refinement. 
 
 Best,
 Herman
 
  
 
 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of James 
 Holton
 Sent: Friday, November 19, 2010 9:20 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] relationship between B factors and Koff
 
 I don't think there is any relationship between rate constants and B factors. 
  Yes, there is the hand-wavy argument of disorder begets disorder (and 
 people almost always LITERALLY wave their hands when they propose this), but 
 you have to be much more careful than that when it comes to thermodynamics.  
 Yes, disorder and entropy are related, but just because a ligand is 
 disordered does not mean that the delta-S term of the binding delta-G is 
 higher.
 
 Now, there is a relationship between equilibrium constants and occupancies, 
 since the occupancy is really just the ratio of the concentration of 
 protein-bound-to-ligand to the total protein concentration, and an 
 equilibrium exists between these two species.  NB all the usual caveats of 
 how crystal packing could change binding constants, etc.  You could logically 
 extend this to B factors by invoking a property of refinement:  Specifically, 
 if the true 
 occupancy is less than 1, but modeled as 1.00, your refinement program will 
 give you a B factor that is larger than the true atomic B factor.  However, 
 if you try to make this claim, then the obvious cantankerous reviewer 
 suggestion would be to refine the occupancy.  
 Problem is, refining both occupancy and B at the same time is usually 
 unstable at moderate resolution.  In general, it is hard to distinguish 
 between something that is flopping around (high B factor) and something that 
 is simply not there part of the time (low occupancy).
 
 I know it is tempting to try and relate B factors directly to entropy, but 
 the disorder that leads to large B factors has a lot more to do with 
 crystals than it has to do with proteins.  For example, there are plenty of 
 tightly-bound complexes that don't diffract well at all.  If you refine these 
 structures, you will get big B factors (roughly, B =
 4*d^2+12 where d is the resolution in Angstrom).  You may even have several 
 crystal forms of the same thing with different Wilson B factors, but that is 
 in no way evidence that the proteins in the two crystal forms somehow have 
 different binding constants or rate constants.
 
 On the bright side, in your case it sounds like you have an entire protomer 
 that is disorered relative to the rest of the crystal lattice.  We have 
 seen a few cases now like this where dehydrating the crystal (with an FMS or 
 similar procedure) causes the unit cell to shrink and this locks the wobbly 
 molecule into place.  I think this is the principle mechanism of improved 
 diffraction from dehydration.  No, it does not work very often!  But 
 sometimes it does.
 
 -James Holton
 MAD Scientist
 
 On 11/19/2010 4:58 AM, Sebastiano Pasqualato wrote:
  Hi all,
  I have a crystallographical/biochemical problem, and maybe some of you guys 
  can help me out.
 
  We have recently crystallized a protein:protein complex, whose Kd has been 
  measured being ca. 10 uM (both by fluorescence

Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Vellieux Frederic
I can direct you to PDB entry 1EWY, where the average isotropic 
temperature factor for the major component of the complex is ca. 47 A**2 
and that for the smaller component is ca. 69 A**2. Similar values than 
the ones you are reporting. I am assuming some sort of disorder, or if 
you prefer, wobbling of the smaller component at the lever of the 
binding site.


Fred.

Sebastiano Pasqualato wrote:

Hi all,
I have a crystallographical/biochemical problem, and maybe some of you guys can 
help me out.

We have recently crystallized a protein:protein complex, whose Kd has been 
measured being ca. 10 uM (both by fluorescence polarization and surface plasmon 
resonance).
Despite the 'decent' affinity, we couldn't purify an homogeneous complex in 
size exclusion chromatography, even mixing the protein at concentrations up to 
80-100 uM each.
We explained this behavior by assuming that extremely high Kon/Koff values 
combine to give this 10 uM affinity, and the high Koff value would account for 
the dissociation going on during size exclusion chromatography. We have partial 
evidence for this from the SPR curves, although we haven't actually measured 
the Kon/Koff values.

We eventually managed to solve the crystal structure of the complex by mixing 
the two proteins (we had to add an excess of one of them to get good 
diffraction data).
Once solved the structure (which makes perfect biological sense and has been 
validated), we get mean B factors for one of the component (the larger) much 
lower than those of the other component (the smaller one, which we had in 
excess). We're talking about 48 Å^2 vs. 75 Å^2.

I was wondering if anybody has had some similar cases, or has any hint on the 
possible relationship it might (or might not) exist between high a Koff value 
and high B factors (a relationship we are tempted to draw).

Thanks in advance,
best regards,
ciao
s


  


Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Justin Hall

Hi Sebastiano,

I have had some experience with protein:protein complexes with KD ~  
10-1 uM, kinetic characterization and trying to purify a complex of  
these proteins using SEC. While I would say that if you have reliable  
evidence from SPR that you have a fast on (high Kon), then you must  
have a fast off (high Koff) because by definition KD = 10 E-6 =  
Koff/Kon. However, I have observed several systems where you have a KD  
~ 10-1 uM, but the kinetics are not fast on/fast off. In my  
experience, I have never seen anything in the crystal structures of  
the weak affinity complexes I have solved that would coorelate  
B-factors to Kon/Koff, and while it might be tempting for you to draw  
this comparison in your structure, I would warn that this is too large  
a leap without further (non-anecdotal) evidence.


As a further note, during SEC purification of complexes, I have  
observed that you generally have to have the complexes at at least 5  
to 10-fold higher initial concentration if you want to purify the  
complex, which you are only pushing with your 80-100 uM high end  
concentration. A colleague of mine once told me this is due to a 5 to  
10-fold dilution effect upon addition to the column, but I have never  
verified this nor read any primary source that validated this so I  
cannot supply a reference (others might be able to help here). Good  
luck and cheers~


~Justin



Quoting Sebastiano Pasqualato sebastiano.pasqual...@ifom-ieo-campus.it:


Hi all,
I have a crystallographical/biochemical problem, and maybe some of  
you guys can help me out.


We have recently crystallized a protein:protein complex, whose Kd  
has been measured being ca. 10 uM (both by fluorescence polarization  
and surface plasmon resonance).
Despite the 'decent' affinity, we couldn't purify an homogeneous  
complex in size exclusion chromatography, even mixing the protein at  
concentrations up to 80-100 uM each.
We explained this behavior by assuming that extremely high Kon/Koff  
values combine to give this 10 uM affinity, and the high Koff value  
would account for the dissociation going on during size exclusion  
chromatography. We have partial evidence for this from the SPR  
curves, although we haven't actually measured the Kon/Koff values.


We eventually managed to solve the crystal structure of the complex  
by mixing the two proteins (we had to add an excess of one of them  
to get good diffraction data).
Once solved the structure (which makes perfect biological sense and  
has been validated), we get mean B factors for one of the component  
(the larger) much lower than those of the other component (the  
smaller one, which we had in excess). We're talking about 48 Å^2 vs.  
75 Å^2.


I was wondering if anybody has had some similar cases, or has any  
hint on the possible relationship it might (or might not) exist  
between high a Koff value and high B factors (a relationship we are  
tempted to draw).


Thanks in advance,
best regards,
ciao
s


--
Sebastiano Pasqualato, PhD
IFOM-IEO Campus
Dipartimento di Oncologia Sperimentale
Istituto Europeo di Oncologia
via Adamello, 16
20139 - Milano
Italy

tel +39 02 9437 5094
fax +39 02 9437 5990




Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Jacob Keller
A suggestion for purifying the complex: let's say there is a 5mL gap between 
the complex and one of its (smaller)constituents A. You can pre-load the 
column with, say, 5mL of A at 1uM, then inject the complex at 80-100uM, to 
be injected right after the pre-load. This should provide approximately 
equilibrium conditions, so that the complex should be basically 1:1 when it 
comes out, even with a high Koff. (Alternatively, for true equilibrium 
conditions, just equilibrate the entire column in A, then inject the 
complex.)


JPK

- Original Message - 
From: Justin Hall hallj...@onid.orst.edu

To: CCP4BB@JISCMAIL.AC.UK
Sent: Friday, November 19, 2010 7:32 AM
Subject: Re: [ccp4bb] relationship between B factors and Koff


Hi Sebastiano,

I have had some experience with protein:protein complexes with KD ~
10-1 uM, kinetic characterization and trying to purify a complex of
these proteins using SEC. While I would say that if you have reliable
evidence from SPR that you have a fast on (high Kon), then you must
have a fast off (high Koff) because by definition KD = 10 E-6 =
Koff/Kon. However, I have observed several systems where you have a KD
~ 10-1 uM, but the kinetics are not fast on/fast off. In my
experience, I have never seen anything in the crystal structures of
the weak affinity complexes I have solved that would coorelate
B-factors to Kon/Koff, and while it might be tempting for you to draw
this comparison in your structure, I would warn that this is too large
a leap without further (non-anecdotal) evidence.

As a further note, during SEC purification of complexes, I have
observed that you generally have to have the complexes at at least 5
to 10-fold higher initial concentration if you want to purify the
complex, which you are only pushing with your 80-100 uM high end
concentration. A colleague of mine once told me this is due to a 5 to
10-fold dilution effect upon addition to the column, but I have never
verified this nor read any primary source that validated this so I
cannot supply a reference (others might be able to help here). Good
luck and cheers~

~Justin



Quoting Sebastiano Pasqualato sebastiano.pasqual...@ifom-ieo-campus.it:


Hi all,
I have a crystallographical/biochemical problem, and maybe some of  you 
guys can help me out.


We have recently crystallized a protein:protein complex, whose Kd  has 
been measured being ca. 10 uM (both by fluorescence polarization  and 
surface plasmon resonance).
Despite the 'decent' affinity, we couldn't purify an homogeneous  complex 
in size exclusion chromatography, even mixing the protein at 
concentrations up to 80-100 uM each.
We explained this behavior by assuming that extremely high Kon/Koff 
values combine to give this 10 uM affinity, and the high Koff value  would 
account for the dissociation going on during size exclusion 
chromatography. We have partial evidence for this from the SPR  curves, 
although we haven't actually measured the Kon/Koff values.


We eventually managed to solve the crystal structure of the complex  by 
mixing the two proteins (we had to add an excess of one of them  to get 
good diffraction data).
Once solved the structure (which makes perfect biological sense and  has 
been validated), we get mean B factors for one of the component  (the 
larger) much lower than those of the other component (the  smaller one, 
which we had in excess). We're talking about 48 Å^2 vs.  75 Å^2.


I was wondering if anybody has had some similar cases, or has any  hint on 
the possible relationship it might (or might not) exist  between high a 
Koff value and high B factors (a relationship we are  tempted to draw).


Thanks in advance,
best regards,
ciao
s


--
Sebastiano Pasqualato, PhD
IFOM-IEO Campus
Dipartimento di Oncologia Sperimentale
Istituto Europeo di Oncologia
via Adamello, 16
20139 - Milano
Italy

tel +39 02 9437 5094
fax +39 02 9437 5990





***
Jacob Pearson Keller
Northwestern University
Medical Scientist Training Program
Dallos Laboratory
F. Searle 1-240
2240 Campus Drive
Evanston IL 60208
lab: 847.491.2438
cel: 773.608.9185
email: j-kell...@northwestern.edu
***


Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Herman . Schreuder
Dear Jacob,
The SEC is generally run to separate the complex from the unbound components. 
If run the way your propose, the peak of unbound preinjected smaller component 
coincides with the peak of the complex and the final stochiometry is not better 
than by just mixing the components without SEC. 

Best, Herman 

-Original Message-
From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of Jacob 
Keller
Sent: Friday, November 19, 2010 3:01 PM
To: CCP4BB@JISCMAIL.AC.UK
Subject: Re: [ccp4bb] relationship between B factors and Koff

A suggestion for purifying the complex: let's say there is a 5mL gap between 
the complex and one of its (smaller)constituents A. You can pre-load the column 
with, say, 5mL of A at 1uM, then inject the complex at 80-100uM, to be injected 
right after the pre-load. This should provide approximately equilibrium 
conditions, so that the complex should be basically 1:1 when it comes out, even 
with a high Koff. (Alternatively, for true equilibrium conditions, just 
equilibrate the entire column in A, then inject the
complex.)

JPK

- Original Message -
From: Justin Hall hallj...@onid.orst.edu
To: CCP4BB@JISCMAIL.AC.UK
Sent: Friday, November 19, 2010 7:32 AM
Subject: Re: [ccp4bb] relationship between B factors and Koff


Hi Sebastiano,

I have had some experience with protein:protein complexes with KD ~
10-1 uM, kinetic characterization and trying to purify a complex of
these proteins using SEC. While I would say that if you have reliable
evidence from SPR that you have a fast on (high Kon), then you must
have a fast off (high Koff) because by definition KD = 10 E-6 =
Koff/Kon. However, I have observed several systems where you have a KD
~ 10-1 uM, but the kinetics are not fast on/fast off. In my
experience, I have never seen anything in the crystal structures of
the weak affinity complexes I have solved that would coorelate
B-factors to Kon/Koff, and while it might be tempting for you to draw
this comparison in your structure, I would warn that this is too large
a leap without further (non-anecdotal) evidence.

As a further note, during SEC purification of complexes, I have
observed that you generally have to have the complexes at at least 5
to 10-fold higher initial concentration if you want to purify the
complex, which you are only pushing with your 80-100 uM high end
concentration. A colleague of mine once told me this is due to a 5 to
10-fold dilution effect upon addition to the column, but I have never
verified this nor read any primary source that validated this so I
cannot supply a reference (others might be able to help here). Good
luck and cheers~

~Justin



Quoting Sebastiano Pasqualato sebastiano.pasqual...@ifom-ieo-campus.it:

 Hi all,
 I have a crystallographical/biochemical problem, and maybe some of  you 
 guys can help me out.

 We have recently crystallized a protein:protein complex, whose Kd  has 
 been measured being ca. 10 uM (both by fluorescence polarization  and 
 surface plasmon resonance).
 Despite the 'decent' affinity, we couldn't purify an homogeneous  complex 
 in size exclusion chromatography, even mixing the protein at 
 concentrations up to 80-100 uM each.
 We explained this behavior by assuming that extremely high Kon/Koff 
 values combine to give this 10 uM affinity, and the high Koff value  would 
 account for the dissociation going on during size exclusion 
 chromatography. We have partial evidence for this from the SPR  curves, 
 although we haven't actually measured the Kon/Koff values.

 We eventually managed to solve the crystal structure of the complex  by 
 mixing the two proteins (we had to add an excess of one of them  to get 
 good diffraction data).
 Once solved the structure (which makes perfect biological sense and  has 
 been validated), we get mean B factors for one of the component  (the 
 larger) much lower than those of the other component (the  smaller one, 
 which we had in excess). We're talking about 48 Å^2 vs.  75 Å^2.

 I was wondering if anybody has had some similar cases, or has any  hint on 
 the possible relationship it might (or might not) exist  between high a 
 Koff value and high B factors (a relationship we are  tempted to draw).

 Thanks in advance,
 best regards,
 ciao
 s


 --
 Sebastiano Pasqualato, PhD
 IFOM-IEO Campus
 Dipartimento di Oncologia Sperimentale
 Istituto Europeo di Oncologia
 via Adamello, 16
 20139 - Milano
 Italy

 tel +39 02 9437 5094
 fax +39 02 9437 5990




***
Jacob Pearson Keller
Northwestern University
Medical Scientist Training Program
Dallos Laboratory
F. Searle 1-240
2240 Campus Drive
Evanston IL 60208
lab: 847.491.2438
cel: 773.608.9185
email: j-kell...@northwestern.edu
***


Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Jacob Keller
Well, mixing together has a much bigger pipetting error (to get, say,
500uM, you have to add 1000uM proteins A and B together, so with a
pipetting error of ~1% even (and concentrated protein solutions seem
to be tricky to pipette accurately), there would be an error of 10uM.
Also, there are the errors associated with concentration
determination, which are probably not trivial, especially with low EC.
If, however, one of the components A is preloaded at low concentration
(1-5uM, say), as I have recommended, the excess of that component with
be exactly 1-5uM, assuming the complex was loaded with a slight excess
of A. And this, as a percentage of the total 500uM, is much less than
the various errors involved in mixing the two together.

So this would be the procedure:

Mix A with B at a calculated stoichiometric ratio of 1.1:1.0
Preload the column with a low level of A just before injecting the complex
Inject the complex
Collect fractions, solve structure, publish in your favorite venue

JPK


On Fri, Nov 19, 2010 at 8:30 AM,  herman.schreu...@sanofi-aventis.com wrote:
 Dear Jacob,
 The SEC is generally run to separate the complex from the unbound components. 
 If run the way your propose, the peak of unbound preinjected smaller 
 component coincides with the peak of the complex and the final stochiometry 
 is not better than by just mixing the components without SEC.

 Best, Herman

 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of Jacob 
 Keller
 Sent: Friday, November 19, 2010 3:01 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] relationship between B factors and Koff

 A suggestion for purifying the complex: let's say there is a 5mL gap between 
 the complex and one of its (smaller)constituents A. You can pre-load the 
 column with, say, 5mL of A at 1uM, then inject the complex at 80-100uM, to be 
 injected right after the pre-load. This should provide approximately 
 equilibrium conditions, so that the complex should be basically 1:1 when it 
 comes out, even with a high Koff. (Alternatively, for true equilibrium 
 conditions, just equilibrate the entire column in A, then inject the
 complex.)

 JPK

 - Original Message -
 From: Justin Hall hallj...@onid.orst.edu
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Friday, November 19, 2010 7:32 AM
 Subject: Re: [ccp4bb] relationship between B factors and Koff


 Hi Sebastiano,

 I have had some experience with protein:protein complexes with KD ~
 10-1 uM, kinetic characterization and trying to purify a complex of
 these proteins using SEC. While I would say that if you have reliable
 evidence from SPR that you have a fast on (high Kon), then you must
 have a fast off (high Koff) because by definition KD = 10 E-6 =
 Koff/Kon. However, I have observed several systems where you have a KD
 ~ 10-1 uM, but the kinetics are not fast on/fast off. In my
 experience, I have never seen anything in the crystal structures of
 the weak affinity complexes I have solved that would coorelate
 B-factors to Kon/Koff, and while it might be tempting for you to draw
 this comparison in your structure, I would warn that this is too large
 a leap without further (non-anecdotal) evidence.

 As a further note, during SEC purification of complexes, I have
 observed that you generally have to have the complexes at at least 5
 to 10-fold higher initial concentration if you want to purify the
 complex, which you are only pushing with your 80-100 uM high end
 concentration. A colleague of mine once told me this is due to a 5 to
 10-fold dilution effect upon addition to the column, but I have never
 verified this nor read any primary source that validated this so I
 cannot supply a reference (others might be able to help here). Good
 luck and cheers~

 ~Justin



 Quoting Sebastiano Pasqualato sebastiano.pasqual...@ifom-ieo-campus.it:

 Hi all,
 I have a crystallographical/biochemical problem, and maybe some of  you
 guys can help me out.

 We have recently crystallized a protein:protein complex, whose Kd  has
 been measured being ca. 10 uM (both by fluorescence polarization  and
 surface plasmon resonance).
 Despite the 'decent' affinity, we couldn't purify an homogeneous  complex
 in size exclusion chromatography, even mixing the protein at
 concentrations up to 80-100 uM each.
 We explained this behavior by assuming that extremely high Kon/Koff
 values combine to give this 10 uM affinity, and the high Koff value  would
 account for the dissociation going on during size exclusion
 chromatography. We have partial evidence for this from the SPR  curves,
 although we haven't actually measured the Kon/Koff values.

 We eventually managed to solve the crystal structure of the complex  by
 mixing the two proteins (we had to add an excess of one of them  to get
 good diffraction data).
 Once solved the structure (which makes perfect biological sense and  has
 been validated), we get mean B factors for one of the component  (the
 larger) much

Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread VAN RAAIJ , MARK JOHAN

Hi Sebastiano, 
I don't see how the k-off would influence this, given the timescale of growing 
crystals. 
An explanation in terms of high Kd and relative lack of crystal contacts for 
the component with higher temperature factors would sound more convincing to 
me. 
Mark 

Quoting Vellieux Frederic:

 I can direct you to PDB entry 1EWY, where the average isotropic 
 temperature factor for the major component of the complex is ca. 47 
 A**2 and that for the smaller component is ca. 69 A**2. Similar 
 values than the ones you are reporting. I am assuming some sort of 
 disorder, or if you prefer, wobbling of the smaller component at 
 the lever of the binding site.

 Fred.

 Sebastiano Pasqualato wrote:
 Hi all,
 I have a crystallographical/biochemical problem, and maybe some of 
 you guys can help me out.

 We have recently crystallized a protein:protein complex, whose Kd 
 has been measured being ca. 10 uM (both by fluorescence polarization 
 and surface plasmon resonance).
 Despite the 'decent' affinity, we couldn't purify an homogeneous 
 complex in size exclusion chromatography, even mixing the protein at 
 concentrations up to 80-100 uM each.
 We explained this behavior by assuming that extremely high Kon/Koff 
 values combine to give this 10 uM affinity, and the high Koff value 
 would account for the dissociation going on during size exclusion 
 chromatography. We have partial evidence for this from the SPR 
 curves, although we haven't actually measured the Kon/Koff values.

 We eventually managed to solve the crystal structure of the complex 
 by mixing the two proteins (we had to add an excess of one of them 
 to get good diffraction data).
 Once solved the structure (which makes perfect biological sense and 
 has been validated), we get mean B factors for one of the component 
 (the larger) much lower than those of the other component (the 
 smaller one, which we had in excess). We're talking about 48 Å^2 vs. 
 75 Å^2.

 I was wondering if anybody has had some similar cases, or has any 
 hint on the possible relationship it might (or might not) exist 
 between high a Koff value and high B factors (a relationship we are 
 tempted to draw).

 Thanks in advance,
 best regards,
 ciao
 s





Mark J van Raaij
Laboratorio M-4
Dpto de Estructura de Macromoléculas
Centro Nacional de Biotecnología - CSIC
c/Darwin 3, Campus Cantoblanco
28049 Madrid
tel. 91 585 4616
email: mjvanra...@cnb.csic.es

Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Jacob Keller
I should add that this procedure is really only advantageous for high
Koff complexes. If the complex does not dissociate appreciably in the
time required for SEC, I agree that there is no great benefit for
doing it my way. I have been working recently, however, on a high Koff
complex, so have been thinking about how to get exactly the right
ratio (other suggestions welcome!)

Jacob

On Fri, Nov 19, 2010 at 8:45 AM, Jacob Keller
j-kell...@fsm.northwestern.edu wrote:
 Well, mixing together has a much bigger pipetting error (to get, say,
 500uM, you have to add 1000uM proteins A and B together, so with a
 pipetting error of ~1% even (and concentrated protein solutions seem
 to be tricky to pipette accurately), there would be an error of 10uM.
 Also, there are the errors associated with concentration
 determination, which are probably not trivial, especially with low EC.
 If, however, one of the components A is preloaded at low concentration
 (1-5uM, say), as I have recommended, the excess of that component with
 be exactly 1-5uM, assuming the complex was loaded with a slight excess
 of A. And this, as a percentage of the total 500uM, is much less than
 the various errors involved in mixing the two together.

 So this would be the procedure:

 Mix A with B at a calculated stoichiometric ratio of 1.1:1.0
 Preload the column with a low level of A just before injecting the complex
 Inject the complex
 Collect fractions, solve structure, publish in your favorite venue

 JPK


 On Fri, Nov 19, 2010 at 8:30 AM,  herman.schreu...@sanofi-aventis.com wrote:
 Dear Jacob,
 The SEC is generally run to separate the complex from the unbound 
 components. If run the way your propose, the peak of unbound preinjected 
 smaller component coincides with the peak of the complex and the final 
 stochiometry is not better than by just mixing the components without SEC.

 Best, Herman

 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of Jacob 
 Keller
 Sent: Friday, November 19, 2010 3:01 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] relationship between B factors and Koff

 A suggestion for purifying the complex: let's say there is a 5mL gap between 
 the complex and one of its (smaller)constituents A. You can pre-load the 
 column with, say, 5mL of A at 1uM, then inject the complex at 80-100uM, to 
 be injected right after the pre-load. This should provide approximately 
 equilibrium conditions, so that the complex should be basically 1:1 when it 
 comes out, even with a high Koff. (Alternatively, for true equilibrium 
 conditions, just equilibrate the entire column in A, then inject the
 complex.)

 JPK

 - Original Message -
 From: Justin Hall hallj...@onid.orst.edu
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Friday, November 19, 2010 7:32 AM
 Subject: Re: [ccp4bb] relationship between B factors and Koff


 Hi Sebastiano,

 I have had some experience with protein:protein complexes with KD ~
 10-1 uM, kinetic characterization and trying to purify a complex of
 these proteins using SEC. While I would say that if you have reliable
 evidence from SPR that you have a fast on (high Kon), then you must
 have a fast off (high Koff) because by definition KD = 10 E-6 =
 Koff/Kon. However, I have observed several systems where you have a KD
 ~ 10-1 uM, but the kinetics are not fast on/fast off. In my
 experience, I have never seen anything in the crystal structures of
 the weak affinity complexes I have solved that would coorelate
 B-factors to Kon/Koff, and while it might be tempting for you to draw
 this comparison in your structure, I would warn that this is too large
 a leap without further (non-anecdotal) evidence.

 As a further note, during SEC purification of complexes, I have
 observed that you generally have to have the complexes at at least 5
 to 10-fold higher initial concentration if you want to purify the
 complex, which you are only pushing with your 80-100 uM high end
 concentration. A colleague of mine once told me this is due to a 5 to
 10-fold dilution effect upon addition to the column, but I have never
 verified this nor read any primary source that validated this so I
 cannot supply a reference (others might be able to help here). Good
 luck and cheers~

 ~Justin



 Quoting Sebastiano Pasqualato sebastiano.pasqual...@ifom-ieo-campus.it:

 Hi all,
 I have a crystallographical/biochemical problem, and maybe some of  you
 guys can help me out.

 We have recently crystallized a protein:protein complex, whose Kd  has
 been measured being ca. 10 uM (both by fluorescence polarization  and
 surface plasmon resonance).
 Despite the 'decent' affinity, we couldn't purify an homogeneous  complex
 in size exclusion chromatography, even mixing the protein at
 concentrations up to 80-100 uM each.
 We explained this behavior by assuming that extremely high Kon/Koff
 values combine to give this 10 uM affinity, and the high Koff value  would
 account for the dissociation going

Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Herman . Schreuder
The amount of small component one looses from the complex depends on the Koff 
and the local concentrations during the SEC run, so I doubt if one could 
estimate those losses with an error less than the 1% pipetting error you quote. 
What I would do is to pipet dilute solutions, and concentrate afterwards. If 
there are conditions which give bad crystals, one could also harvest these 
crystals, redissolve them and set up crystallizations with those.

My two cents,
Herman 

-Original Message-
From: Jacob Keller [mailto:j-kell...@fsm.northwestern.edu] 
Sent: Friday, November 19, 2010 3:57 PM
To: Schreuder, Herman RD/DE
Cc: CCP4BB@jiscmail.ac.uk
Subject: Re: [ccp4bb] relationship between B factors and Koff

I should add that this procedure is really only advantageous for high Koff 
complexes. If the complex does not dissociate appreciably in the time required 
for SEC, I agree that there is no great benefit for doing it my way. I have 
been working recently, however, on a high Koff complex, so have been thinking 
about how to get exactly the right ratio (other suggestions welcome!)

Jacob

On Fri, Nov 19, 2010 at 8:45 AM, Jacob Keller j-kell...@fsm.northwestern.edu 
wrote:
 Well, mixing together has a much bigger pipetting error (to get, say, 
 500uM, you have to add 1000uM proteins A and B together, so with a 
 pipetting error of ~1% even (and concentrated protein solutions seem 
 to be tricky to pipette accurately), there would be an error of 10uM.
 Also, there are the errors associated with concentration 
 determination, which are probably not trivial, especially with low EC.
 If, however, one of the components A is preloaded at low concentration 
 (1-5uM, say), as I have recommended, the excess of that component with 
 be exactly 1-5uM, assuming the complex was loaded with a slight excess 
 of A. And this, as a percentage of the total 500uM, is much less than 
 the various errors involved in mixing the two together.

 So this would be the procedure:

 Mix A with B at a calculated stoichiometric ratio of 1.1:1.0 Preload 
 the column with a low level of A just before injecting the complex 
 Inject the complex Collect fractions, solve structure, publish in your 
 favorite venue

 JPK


 On Fri, Nov 19, 2010 at 8:30 AM,  herman.schreu...@sanofi-aventis.com wrote:
 Dear Jacob,
 The SEC is generally run to separate the complex from the unbound 
 components. If run the way your propose, the peak of unbound preinjected 
 smaller component coincides with the peak of the complex and the final 
 stochiometry is not better than by just mixing the components without SEC.

 Best, Herman

 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of 
 Jacob Keller
 Sent: Friday, November 19, 2010 3:01 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] relationship between B factors and Koff

 A suggestion for purifying the complex: let's say there is a 5mL gap 
 between the complex and one of its (smaller)constituents A. You can 
 pre-load the column with, say, 5mL of A at 1uM, then inject the 
 complex at 80-100uM, to be injected right after the pre-load. This 
 should provide approximately equilibrium conditions, so that the 
 complex should be basically 1:1 when it comes out, even with a high 
 Koff. (Alternatively, for true equilibrium conditions, just 
 equilibrate the entire column in A, then inject the
 complex.)

 JPK

 - Original Message -
 From: Justin Hall hallj...@onid.orst.edu
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Friday, November 19, 2010 7:32 AM
 Subject: Re: [ccp4bb] relationship between B factors and Koff


 Hi Sebastiano,

 I have had some experience with protein:protein complexes with KD ~
 10-1 uM, kinetic characterization and trying to purify a complex of 
 these proteins using SEC. While I would say that if you have reliable 
 evidence from SPR that you have a fast on (high Kon), then you must 
 have a fast off (high Koff) because by definition KD = 10 E-6 = 
 Koff/Kon. However, I have observed several systems where you have a 
 KD ~ 10-1 uM, but the kinetics are not fast on/fast off. In my 
 experience, I have never seen anything in the crystal structures of 
 the weak affinity complexes I have solved that would coorelate 
 B-factors to Kon/Koff, and while it might be tempting for you to draw 
 this comparison in your structure, I would warn that this is too 
 large a leap without further (non-anecdotal) evidence.

 As a further note, during SEC purification of complexes, I have 
 observed that you generally have to have the complexes at at least 5 
 to 10-fold higher initial concentration if you want to purify the 
 complex, which you are only pushing with your 80-100 uM high end 
 concentration. A colleague of mine once told me this is due to a 5 to 
 10-fold dilution effect upon addition to the column, but I have never 
 verified this nor read any primary source that validated this so I 
 cannot supply a reference (others might be able to help

Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Jacob Keller
Well, if one of the components is a small peptide and the solutions
are dilute, it is hard to concentrate the complex, as the free peptide
will go through the concentrator! It's a tricky problem.

JPK

On Fri, Nov 19, 2010 at 9:38 AM,  herman.schreu...@sanofi-aventis.com wrote:
 The amount of small component one looses from the complex depends on the Koff 
 and the local concentrations during the SEC run, so I doubt if one could 
 estimate those losses with an error less than the 1% pipetting error you 
 quote. What I would do is to pipet dilute solutions, and concentrate 
 afterwards. If there are conditions which give bad crystals, one could also 
 harvest these crystals, redissolve them and set up crystallizations with 
 those.

 My two cents,
 Herman

 -Original Message-
 From: Jacob Keller [mailto:j-kell...@fsm.northwestern.edu]
 Sent: Friday, November 19, 2010 3:57 PM
 To: Schreuder, Herman RD/DE
 Cc: CCP4BB@jiscmail.ac.uk
 Subject: Re: [ccp4bb] relationship between B factors and Koff

 I should add that this procedure is really only advantageous for high Koff 
 complexes. If the complex does not dissociate appreciably in the time 
 required for SEC, I agree that there is no great benefit for doing it my way. 
 I have been working recently, however, on a high Koff complex, so have been 
 thinking about how to get exactly the right ratio (other suggestions welcome!)

 Jacob

 On Fri, Nov 19, 2010 at 8:45 AM, Jacob Keller 
 j-kell...@fsm.northwestern.edu wrote:
 Well, mixing together has a much bigger pipetting error (to get, say,
 500uM, you have to add 1000uM proteins A and B together, so with a
 pipetting error of ~1% even (and concentrated protein solutions seem
 to be tricky to pipette accurately), there would be an error of 10uM.
 Also, there are the errors associated with concentration
 determination, which are probably not trivial, especially with low EC.
 If, however, one of the components A is preloaded at low concentration
 (1-5uM, say), as I have recommended, the excess of that component with
 be exactly 1-5uM, assuming the complex was loaded with a slight excess
 of A. And this, as a percentage of the total 500uM, is much less than
 the various errors involved in mixing the two together.

 So this would be the procedure:

 Mix A with B at a calculated stoichiometric ratio of 1.1:1.0 Preload
 the column with a low level of A just before injecting the complex
 Inject the complex Collect fractions, solve structure, publish in your
 favorite venue

 JPK


 On Fri, Nov 19, 2010 at 8:30 AM,  herman.schreu...@sanofi-aventis.com 
 wrote:
 Dear Jacob,
 The SEC is generally run to separate the complex from the unbound 
 components. If run the way your propose, the peak of unbound preinjected 
 smaller component coincides with the peak of the complex and the final 
 stochiometry is not better than by just mixing the components without SEC.

 Best, Herman

 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf Of
 Jacob Keller
 Sent: Friday, November 19, 2010 3:01 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] relationship between B factors and Koff

 A suggestion for purifying the complex: let's say there is a 5mL gap
 between the complex and one of its (smaller)constituents A. You can
 pre-load the column with, say, 5mL of A at 1uM, then inject the
 complex at 80-100uM, to be injected right after the pre-load. This
 should provide approximately equilibrium conditions, so that the
 complex should be basically 1:1 when it comes out, even with a high
 Koff. (Alternatively, for true equilibrium conditions, just
 equilibrate the entire column in A, then inject the
 complex.)

 JPK

 - Original Message -
 From: Justin Hall hallj...@onid.orst.edu
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Friday, November 19, 2010 7:32 AM
 Subject: Re: [ccp4bb] relationship between B factors and Koff


 Hi Sebastiano,

 I have had some experience with protein:protein complexes with KD ~
 10-1 uM, kinetic characterization and trying to purify a complex of
 these proteins using SEC. While I would say that if you have reliable
 evidence from SPR that you have a fast on (high Kon), then you must
 have a fast off (high Koff) because by definition KD = 10 E-6 =
 Koff/Kon. However, I have observed several systems where you have a
 KD ~ 10-1 uM, but the kinetics are not fast on/fast off. In my
 experience, I have never seen anything in the crystal structures of
 the weak affinity complexes I have solved that would coorelate
 B-factors to Kon/Koff, and while it might be tempting for you to draw
 this comparison in your structure, I would warn that this is too
 large a leap without further (non-anecdotal) evidence.

 As a further note, during SEC purification of complexes, I have
 observed that you generally have to have the complexes at at least 5
 to 10-fold higher initial concentration if you want to purify the
 complex, which you are only pushing with your 80-100 uM high end

Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Herman . Schreuder
If the peptide is really small compared to the protein, I would just add excess 
peptide and not worry about the stochiometry.

Best,
Herman 

-Original Message-
From: Jacob Keller [mailto:j-kell...@fsm.northwestern.edu] 
Sent: Friday, November 19, 2010 5:03 PM
To: Schreuder, Herman RD/DE
Cc: CCP4BB@jiscmail.ac.uk
Subject: Re: [ccp4bb] relationship between B factors and Koff

Well, if one of the components is a small peptide and the solutions are dilute, 
it is hard to concentrate the complex, as the free peptide will go through the 
concentrator! It's a tricky problem.

JPK

On Fri, Nov 19, 2010 at 9:38 AM,  herman.schreu...@sanofi-aventis.com wrote:
 The amount of small component one looses from the complex depends on the Koff 
 and the local concentrations during the SEC run, so I doubt if one could 
 estimate those losses with an error less than the 1% pipetting error you 
 quote. What I would do is to pipet dilute solutions, and concentrate 
 afterwards. If there are conditions which give bad crystals, one could also 
 harvest these crystals, redissolve them and set up crystallizations with 
 those.

 My two cents,
 Herman

 -Original Message-
 From: Jacob Keller [mailto:j-kell...@fsm.northwestern.edu]
 Sent: Friday, November 19, 2010 3:57 PM
 To: Schreuder, Herman RD/DE
 Cc: CCP4BB@jiscmail.ac.uk
 Subject: Re: [ccp4bb] relationship between B factors and Koff

 I should add that this procedure is really only advantageous for high 
 Koff complexes. If the complex does not dissociate appreciably in the 
 time required for SEC, I agree that there is no great benefit for 
 doing it my way. I have been working recently, however, on a high Koff 
 complex, so have been thinking about how to get exactly the right 
 ratio (other suggestions welcome!)

 Jacob

 On Fri, Nov 19, 2010 at 8:45 AM, Jacob Keller 
 j-kell...@fsm.northwestern.edu wrote:
 Well, mixing together has a much bigger pipetting error (to get, say, 
 500uM, you have to add 1000uM proteins A and B together, so with a 
 pipetting error of ~1% even (and concentrated protein solutions seem 
 to be tricky to pipette accurately), there would be an error of 10uM.
 Also, there are the errors associated with concentration 
 determination, which are probably not trivial, especially with low EC.
 If, however, one of the components A is preloaded at low 
 concentration (1-5uM, say), as I have recommended, the excess of that 
 component with be exactly 1-5uM, assuming the complex was loaded with 
 a slight excess of A. And this, as a percentage of the total 500uM, 
 is much less than the various errors involved in mixing the two together.

 So this would be the procedure:

 Mix A with B at a calculated stoichiometric ratio of 1.1:1.0 Preload 
 the column with a low level of A just before injecting the complex 
 Inject the complex Collect fractions, solve structure, publish in 
 your favorite venue

 JPK


 On Fri, Nov 19, 2010 at 8:30 AM,  herman.schreu...@sanofi-aventis.com 
 wrote:
 Dear Jacob,
 The SEC is generally run to separate the complex from the unbound 
 components. If run the way your propose, the peak of unbound preinjected 
 smaller component coincides with the peak of the complex and the final 
 stochiometry is not better than by just mixing the components without SEC.

 Best, Herman

 -Original Message-
 From: CCP4 bulletin board [mailto:ccp...@jiscmail.ac.uk] On Behalf 
 Of Jacob Keller
 Sent: Friday, November 19, 2010 3:01 PM
 To: CCP4BB@JISCMAIL.AC.UK
 Subject: Re: [ccp4bb] relationship between B factors and Koff

 A suggestion for purifying the complex: let's say there is a 5mL gap 
 between the complex and one of its (smaller)constituents A. You can 
 pre-load the column with, say, 5mL of A at 1uM, then inject the 
 complex at 80-100uM, to be injected right after the pre-load. This 
 should provide approximately equilibrium conditions, so that the 
 complex should be basically 1:1 when it comes out, even with a high 
 Koff. (Alternatively, for true equilibrium conditions, just 
 equilibrate the entire column in A, then inject the
 complex.)

 JPK

 - Original Message -
 From: Justin Hall hallj...@onid.orst.edu
 To: CCP4BB@JISCMAIL.AC.UK
 Sent: Friday, November 19, 2010 7:32 AM
 Subject: Re: [ccp4bb] relationship between B factors and Koff


 Hi Sebastiano,

 I have had some experience with protein:protein complexes with KD ~
 10-1 uM, kinetic characterization and trying to purify a complex of 
 these proteins using SEC. While I would say that if you have 
 reliable evidence from SPR that you have a fast on (high Kon), then 
 you must have a fast off (high Koff) because by definition KD = 10 
 E-6 = Koff/Kon. However, I have observed several systems where you 
 have a KD ~ 10-1 uM, but the kinetics are not fast on/fast off. In 
 my experience, I have never seen anything in the crystal structures 
 of the weak affinity complexes I have solved that would coorelate 
 B-factors to Kon/Koff, and while it might

Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Tanner, John J.
This doesn't directly address your question, but since the subject of analyzing 
protein-protein interactions with gel filtration is raised on this bb 
occasionally, I thought I would mention that there are cases in which 
conventional gel filtration chromatography fails to provide evidence of a known 
protein:protein interaction.  In such cases, the Hummel-Dreyer gel filtration 
method is sometimes used.  It involves supplementing the running buffer with 
one of the proteins, so you need a lot of protein.  Here are two references:

Proc. Natl. Acad. Sci. USA 88 (1991)
Biochemisrry (1993) 32, 11124-11131


On 11/19/10 6:58 AM, Sebastiano Pasqualato 
sebastiano.pasqual...@ifom-ieo-campus.it wrote:

Hi all,
I have a crystallographical/biochemical problem, and maybe some of you guys can 
help me out.

We have recently crystallized a protein:protein complex, whose Kd has been 
measured being ca. 10 uM (both by fluorescence polarization and surface plasmon 
resonance).
Despite the 'decent' affinity, we couldn't purify an homogeneous complex in 
size exclusion chromatography, even mixing the protein at concentrations up to 
80-100 uM each.
We explained this behavior by assuming that extremely high Kon/Koff values 
combine to give this 10 uM affinity, and the high Koff value would account for 
the dissociation going on during size exclusion chromatography. We have partial 
evidence for this from the SPR curves, although we haven't actually measured 
the Kon/Koff values.

We eventually managed to solve the crystal structure of the complex by mixing 
the two proteins (we had to add an excess of one of them to get good 
diffraction data).
Once solved the structure (which makes perfect biological sense and has been 
validated), we get mean B factors for one of the component (the larger) much 
lower than those of the other component (the smaller one, which we had in 
excess). We're talking about 48 Å^2 vs. 75 Å^2.

I was wondering if anybody has had some similar cases, or has any hint on the 
possible relationship it might (or might not) exist between high a Koff value 
and high B factors (a relationship we are tempted to draw).

Thanks in advance,
best regards,
ciao
s


--
Sebastiano Pasqualato, PhD
IFOM-IEO Campus
Dipartimento di Oncologia Sperimentale
Istituto Europeo di Oncologia
via Adamello, 16
20139 - Milano
Italy

tel +39 02 9437 5094
fax +39 02 9437 5990


Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Jacob Keller
Yes, this is where I heard of the technique, and my suggestion was a 
modification of that, but my references were:


Hummel JP  Dreyer WJ (1962) Measurement of protein-binding phenomena by gel 
filtration. Biochim Biophys Acta 63:530-532.


Ratner D (1974) The interaction bacterial and phage proteins with 
immobilized Escherichia coli RNA polymerase. J Mol Biol 88(2):373-383.



JPK


- Original Message - 
From: Tanner, John J. tanne...@missouri.edu

To: CCP4BB@JISCMAIL.AC.UK
Sent: Friday, November 19, 2010 11:25 AM
Subject: Re: [ccp4bb] relationship between B factors and Koff


This doesn't directly address your question, but since the subject of 
analyzing protein-protein interactions with gel filtration is raised on this 
bb occasionally, I thought I would mention that there are cases in which 
conventional gel filtration chromatography fails to provide evidence of a 
known protein:protein interaction.  In such cases, the Hummel-Dreyer gel 
filtration method is sometimes used.  It involves supplementing the running 
buffer with one of the proteins, so you need a lot of protein.  Here are two 
references:


Proc. Natl. Acad. Sci. USA 88 (1991)
Biochemisrry (1993) 32, 11124-11131


On 11/19/10 6:58 AM, Sebastiano Pasqualato 
sebastiano.pasqual...@ifom-ieo-campus.it wrote:


Hi all,
I have a crystallographical/biochemical problem, and maybe some of you guys 
can help me out.


We have recently crystallized a protein:protein complex, whose Kd has been 
measured being ca. 10 uM (both by fluorescence polarization and surface 
plasmon resonance).
Despite the 'decent' affinity, we couldn't purify an homogeneous complex in 
size exclusion chromatography, even mixing the protein at concentrations up 
to 80-100 uM each.
We explained this behavior by assuming that extremely high Kon/Koff values 
combine to give this 10 uM affinity, and the high Koff value would account 
for the dissociation going on during size exclusion chromatography. We have 
partial evidence for this from the SPR curves, although we haven't actually 
measured the Kon/Koff values.


We eventually managed to solve the crystal structure of the complex by 
mixing the two proteins (we had to add an excess of one of them to get good 
diffraction data).
Once solved the structure (which makes perfect biological sense and has been 
validated), we get mean B factors for one of the component (the larger) much 
lower than those of the other component (the smaller one, which we had in 
excess). We're talking about 48 Å^2 vs. 75 Å^2.


I was wondering if anybody has had some similar cases, or has any hint on 
the possible relationship it might (or might not) exist between high a Koff 
value and high B factors (a relationship we are tempted to draw).


Thanks in advance,
best regards,
ciao
s


--
Sebastiano Pasqualato, PhD
IFOM-IEO Campus
Dipartimento di Oncologia Sperimentale
Istituto Europeo di Oncologia
via Adamello, 16
20139 - Milano
Italy

tel +39 02 9437 5094
fax +39 02 9437 5990


***
Jacob Pearson Keller
Northwestern University
Medical Scientist Training Program
Dallos Laboratory
F. Searle 1-240
2240 Campus Drive
Evanston IL 60208
lab: 847.491.2438
cel: 773.608.9185
email: j-kell...@northwestern.edu
***


Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Narayanasami Sukumar
Sebastiano,
We observed similar phenomenon in our protein-protein complex.  This
reference will give you details (Crystal Structure of an Electron Transfer
Complex between Aromatic Amine Dehydrogenase- Azurin from Alcaligenes
faecalis, Biochemistry, 45, 13500-13510, 2006).

Sukumar

Narayanasami Sukumar
NE-CAT, Building 436E
Advanced Photon Source
Argonne National Laboratory
9700 S.Cass Ave
Argonne, IL 60439-4800
USA

Tel: 630-252-0681
Fax: 630-252-0687
e-mail: suku...@aps.anl.gov






On Fri, Nov 19, 2010 at 6:58 AM, Sebastiano Pasqualato 
sebastiano.pasqual...@ifom-ieo-campus.it wrote:

 Hi all,
 I have a crystallographical/biochemical problem, and maybe some of you guys
 can help me out.

 We have recently crystallized a protein:protein complex, whose Kd has been
 measured being ca. 10 uM (both by fluorescence polarization and surface
 plasmon resonance).
 Despite the 'decent' affinity, we couldn't purify an homogeneous complex in
 size exclusion chromatography, even mixing the protein at concentrations up
 to 80-100 uM each.
 We explained this behavior by assuming that extremely high Kon/Koff values
 combine to give this 10 uM affinity, and the high Koff value would account
 for the dissociation going on during size exclusion chromatography. We have
 partial evidence for this from the SPR curves, although we haven't actually
 measured the Kon/Koff values.

 We eventually managed to solve the crystal structure of the complex by
 mixing the two proteins (we had to add an excess of one of them to get good
 diffraction data).
 Once solved the structure (which makes perfect biological sense and has
 been validated), we get mean B factors for one of the component (the larger)
 much lower than those of the other component (the smaller one, which we had
 in excess). We're talking about 48 Å^2 vs. 75 Å^2.

 I was wondering if anybody has had some similar cases, or has any hint on
 the possible relationship it might (or might not) exist between high a Koff
 value and high B factors (a relationship we are tempted to draw).

 Thanks in advance,
 best regards,
 ciao
 s


 --
 Sebastiano Pasqualato, PhD
 IFOM-IEO Campus
 Dipartimento di Oncologia Sperimentale
 Istituto Europeo di Oncologia
 via Adamello, 16
 20139 - Milano
 Italy

 tel +39 02 9437 5094
 fax +39 02 9437 5990



Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread James Holton
I don't think there is any relationship between rate constants and B 
factors.  Yes, there is the hand-wavy argument of disorder begets 
disorder (and people almost always LITERALLY wave their hands when they 
propose this), but you have to be much more careful than that when it 
comes to thermodynamics.  Yes, disorder and entropy are related, but 
just because a ligand is disordered does not mean that the delta-S 
term of the binding delta-G is higher.


Now, there is a relationship between equilibrium constants and 
occupancies, since the occupancy is really just the ratio of the 
concentration of protein-bound-to-ligand to the total protein 
concentration, and an equilibrium exists between these two species.  NB 
all the usual caveats of how crystal packing could change binding 
constants, etc.  You could logically extend this to B factors by 
invoking a property of refinement:  Specifically, if the true 
occupancy is less than 1, but modeled as 1.00, your refinement program 
will give you a B factor that is larger than the true atomic B 
factor.  However, if you try to make this claim, then the obvious 
cantankerous reviewer suggestion would be to refine the occupancy.  
Problem is, refining both occupancy and B at the same time is usually 
unstable at moderate resolution.  In general, it is hard to distinguish 
between something that is flopping around (high B factor) and something 
that is simply not there part of the time (low occupancy).


I know it is tempting to try and relate B factors directly to entropy, 
but the disorder that leads to large B factors has a lot more to do 
with crystals than it has to do with proteins.  For example, there are 
plenty of tightly-bound complexes that don't diffract well at all.  If 
you refine these structures, you will get big B factors (roughly, B = 
4*d^2+12 where d is the resolution in Angstrom).  You may even have 
several crystal forms of the same thing with different Wilson B factors, 
but that is in no way evidence that the proteins in the two crystal 
forms somehow have different binding constants or rate constants.


On the bright side, in your case it sounds like you have an entire 
protomer that is disorered relative to the rest of the crystal 
lattice.  We have seen a few cases now like this where dehydrating the 
crystal (with an FMS or similar procedure) causes the unit cell to 
shrink and this locks the wobbly molecule into place.  I think this is 
the principle mechanism of improved diffraction from dehydration.  No, 
it does not work very often!  But sometimes it does.


-James Holton
MAD Scientist

On 11/19/2010 4:58 AM, Sebastiano Pasqualato wrote:

Hi all,
I have a crystallographical/biochemical problem, and maybe some of you guys can 
help me out.

We have recently crystallized a protein:protein complex, whose Kd has been 
measured being ca. 10 uM (both by fluorescence polarization and surface plasmon 
resonance).
Despite the 'decent' affinity, we couldn't purify an homogeneous complex in 
size exclusion chromatography, even mixing the protein at concentrations up to 
80-100 uM each.
We explained this behavior by assuming that extremely high Kon/Koff values 
combine to give this 10 uM affinity, and the high Koff value would account for 
the dissociation going on during size exclusion chromatography. We have partial 
evidence for this from the SPR curves, although we haven't actually measured 
the Kon/Koff values.

We eventually managed to solve the crystal structure of the complex by mixing 
the two proteins (we had to add an excess of one of them to get good 
diffraction data).
Once solved the structure (which makes perfect biological sense and has been 
validated), we get mean B factors for one of the component (the larger) much 
lower than those of the other component (the smaller one, which we had in 
excess). We're talking about 48 Å^2 vs. 75 Å^2.

I was wondering if anybody has had some similar cases, or has any hint on the 
possible relationship it might (or might not) exist between high a Koff value 
and high B factors (a relationship we are tempted to draw).

Thanks in advance,
best regards,
ciao
s




Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Daniel Anderson
Maybe you should look at: Rotation barriers in crystals from atomic 
displacement parameters Emily Maverick and Jack Dunitz (1987) Vol 
62(2), 451-459.


They reported that the libration amplitude of one part of a crystalline 
small molecule relative to another part is strongly correlated to bond 
rotation barriers measured by other means. The discussion at the top of 
page 458 seems to be about off rates. It's not up to me if a 
multi-temperature diffraction experiment and a lot of TLS analyses will 
be worth the effort relative to purchasing equipment to measure a 
binding parameter.


-Dan



James Holton wrote:
I don't think there is any relationship between rate constants and B 
factors.  Yes, there is the hand-wavy argument of disorder begets 
disorder (and people almost always LITERALLY wave their hands when 
they propose this), but you have to be much more careful than that 
when it comes to thermodynamics.  Yes, disorder and entropy are 
related, but just because a ligand is disordered does not mean that 
the delta-S term of the binding delta-G is higher.


Now, there is a relationship between equilibrium constants and 
occupancies, since the occupancy is really just the ratio of the 
concentration of protein-bound-to-ligand to the total protein 
concentration, and an equilibrium exists between these two species.  
NB all the usual caveats of how crystal packing could change binding 
constants, etc.  You could logically extend this to B factors by 
invoking a property of refinement:  Specifically, if the true 
occupancy is less than 1, but modeled as 1.00, your refinement 
program will give you a B factor that is larger than the true atomic 
B factor.  However, if you try to make this claim, then the obvious 
cantankerous reviewer suggestion would be to refine the occupancy.  
Problem is, refining both occupancy and B at the same time is usually 
unstable at moderate resolution.  In general, it is hard to 
distinguish between something that is flopping around (high B factor) 
and something that is simply not there part of the time (low 
occupancy).


I know it is tempting to try and relate B factors directly to entropy, 
but the disorder that leads to large B factors has a lot more to do 
with crystals than it has to do with proteins.  For example, there are 
plenty of tightly-bound complexes that don't diffract well at all.  If 
you refine these structures, you will get big B factors (roughly, B = 
4*d^2+12 where d is the resolution in Angstrom).  You may even have 
several crystal forms of the same thing with different Wilson B 
factors, but that is in no way evidence that the proteins in the two 
crystal forms somehow have different binding constants or rate constants.


On the bright side, in your case it sounds like you have an entire 
protomer that is disorered relative to the rest of the crystal 
lattice.  We have seen a few cases now like this where dehydrating the 
crystal (with an FMS or similar procedure) causes the unit cell to 
shrink and this locks the wobbly molecule into place.  I think this 
is the principle mechanism of improved diffraction from 
dehydration.  No, it does not work very often!  But sometimes it does.


-James Holton
MAD Scientist

On 11/19/2010 4:58 AM, Sebastiano Pasqualato wrote:

Hi all,
I have a crystallographical/biochemical problem, and maybe some of 
you guys can help me out.


We have recently crystallized a protein:protein complex, whose Kd has 
been measured being ca. 10 uM (both by fluorescence polarization and 
surface plasmon resonance).
Despite the 'decent' affinity, we couldn't purify an homogeneous 
complex in size exclusion chromatography, even mixing the protein at 
concentrations up to 80-100 uM each.
We explained this behavior by assuming that extremely high Kon/Koff 
values combine to give this 10 uM affinity, and the high Koff value 
would account for the dissociation going on during size exclusion 
chromatography. We have partial evidence for this from the SPR 
curves, although we haven't actually measured the Kon/Koff values.


We eventually managed to solve the crystal structure of the complex 
by mixing the two proteins (we had to add an excess of one of them to 
get good diffraction data).
Once solved the structure (which makes perfect biological sense and 
has been validated), we get mean B factors for one of the component 
(the larger) much lower than those of the other component (the 
smaller one, which we had in excess). We're talking about 48 Å^2 vs. 
75 Å^2.


I was wondering if anybody has had some similar cases, or has any 
hint on the possible relationship it might (or might not) exist 
between high a Koff value and high B factors (a relationship we are 
tempted to draw).


Thanks in advance,
best regards,
ciao
s




Re: [ccp4bb] relationship between B factors and Koff

2010-11-19 Thread Daniel Anderson

The journal was Molecular Physics (1987) Vol 62(2), 451-459.

(my brain librates when I type)

Daniel Anderson wrote:
Maybe you should look at: Rotation barriers in crystals from atomic 
displacement parameters Emily Maverick and Jack Dunitz (1987) Vol 
62(2), 451-459.


They reported that the libration amplitude of one part of a 
crystalline small molecule relative to another part is strongly 
correlated to bond rotation barriers measured by other means. The 
discussion at the top of page 458 seems to be about off rates. It's 
not up to me if a multi-temperature diffraction experiment and a lot 
of TLS analyses will be worth the effort relative to purchasing 
equipment to measure a binding parameter.


-Dan



James Holton wrote:
I don't think there is any relationship between rate constants and B 
factors.  Yes, there is the hand-wavy argument of disorder begets 
disorder (and people almost always LITERALLY wave their hands when 
they propose this), but you have to be much more careful than that 
when it comes to thermodynamics.  Yes, disorder and entropy are 
related, but just because a ligand is disordered does not mean that 
the delta-S term of the binding delta-G is higher.


Now, there is a relationship between equilibrium constants and 
occupancies, since the occupancy is really just the ratio of the 
concentration of protein-bound-to-ligand to the total protein 
concentration, and an equilibrium exists between these two species.  
NB all the usual caveats of how crystal packing could change binding 
constants, etc.  You could logically extend this to B factors by 
invoking a property of refinement:  Specifically, if the true 
occupancy is less than 1, but modeled as 1.00, your refinement 
program will give you a B factor that is larger than the true 
atomic B factor.  However, if you try to make this claim, then the 
obvious cantankerous reviewer suggestion would be to refine the 
occupancy.  Problem is, refining both occupancy and B at the same 
time is usually unstable at moderate resolution.  In general, it is 
hard to distinguish between something that is flopping around (high B 
factor) and something that is simply not there part of the time 
(low occupancy).


I know it is tempting to try and relate B factors directly to 
entropy, but the disorder that leads to large B factors has a lot 
more to do with crystals than it has to do with proteins.  For 
example, there are plenty of tightly-bound complexes that don't 
diffract well at all.  If you refine these structures, you will get 
big B factors (roughly, B = 4*d^2+12 where d is the resolution in 
Angstrom).  You may even have several crystal forms of the same thing 
with different Wilson B factors, but that is in no way evidence that 
the proteins in the two crystal forms somehow have different binding 
constants or rate constants.


On the bright side, in your case it sounds like you have an entire 
protomer that is disorered relative to the rest of the crystal 
lattice.  We have seen a few cases now like this where dehydrating 
the crystal (with an FMS or similar procedure) causes the unit cell 
to shrink and this locks the wobbly molecule into place.  I think 
this is the principle mechanism of improved diffraction from 
dehydration.  No, it does not work very often!  But sometimes it does.


-James Holton
MAD Scientist

On 11/19/2010 4:58 AM, Sebastiano Pasqualato wrote:

Hi all,
I have a crystallographical/biochemical problem, and maybe some of 
you guys can help me out.


We have recently crystallized a protein:protein complex, whose Kd 
has been measured being ca. 10 uM (both by fluorescence polarization 
and surface plasmon resonance).
Despite the 'decent' affinity, we couldn't purify an homogeneous 
complex in size exclusion chromatography, even mixing the protein at 
concentrations up to 80-100 uM each.
We explained this behavior by assuming that extremely high Kon/Koff 
values combine to give this 10 uM affinity, and the high Koff value 
would account for the dissociation going on during size exclusion 
chromatography. We have partial evidence for this from the SPR 
curves, although we haven't actually measured the Kon/Koff values.


We eventually managed to solve the crystal structure of the complex 
by mixing the two proteins (we had to add an excess of one of them 
to get good diffraction data).
Once solved the structure (which makes perfect biological sense and 
has been validated), we get mean B factors for one of the component 
(the larger) much lower than those of the other component (the 
smaller one, which we had in excess). We're talking about 48 Å^2 vs. 
75 Å^2.


I was wondering if anybody has had some similar cases, or has any 
hint on the possible relationship it might (or might not) exist 
between high a Koff value and high B factors (a relationship we are 
tempted to draw).


Thanks in advance,
best regards,
ciao
s