I did the replacement degausser you found the zip file for. I use it for
my 5061s quite successfully.
It fairly closely duplicates the time and current profiles used by the
original.
If you have any questions, please ask.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
Here's a site that has some interesting analyses of some of the various
topics covered in this wonderful thread:
http://www.st-andrews.ac.uk/~www_pa/Scots_Guide/audio/Analog.html
In particular, the advanced topics about halfway down the page have a
detailed analysis of the real impact of skin
If I remember correctly, the 1PPS pulse width is only about 100 ns.
There are two possibilities:
1) you can't see the pulse height properly at the very low sweep rate
you're using. Trigger on it at a much higher sweep rate, crank up the
intensity on your scope, and look carefully.
2) the
My understanding is that the precise cesium hyperfine transition energy,
and hence microwave frequency, is dependent upon the applied C-field,
and the field setting to select a particular transition is the same for
all clocks (which is why these are primary standards). The C-field is
set by
Thanks all. I reviewed some physics on this, and now it makes sense.
Corby, my 5061A is an oldie, it has the early synthesizer that has the
UTC offset correction, which I've set to match atomic time (L2095 on the
switches).
The tube was replaced in '82. So, I need to verify that the C-field
I got one without the outer cover, which isn't quite as useful. I
removed all the junk to get down to the orginal non-double-ovened osc.
One thing I noticed - I wonder where these were actually deployed? The
10811B itself has a fair amount of corrosion on it, lots of pitting. It
for the outer oven and a new cover of
some sort. Is there any reason not to do that?
Thanks,
Ed
Date: Sat, 19 Jul 2008 17:33:07 -0400
From: wje [1][EMAIL PROTECTED]
Subject: Re: [time-nuts] Double ovened 10811-60158 on ebay
To: Discussion of precise time and frequency measurement
[2]time-nuts
Simply put, stability is a measure of how much something varies from a
constant value.
A stable value doesn't mean an accurate value; accuracy is a measure of
how close to some standard value a device emits (in this case) or
measures, for a meter.
So, you could have something that's perfectly
MMIC = Monolithic Microwave Integrated Circuit
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
J. L. Trantham wrote:
I have been reading this thread with interest since I enjoy resurrecting
'dead' instruments. Two questions though:
1. Since I have been out of the
I was faced with this problem a few years ago, too.
I essentially followed John's solution, but used the 1pps GPS pulse as
the PLL reference.
I then used the 32.768K vco output to injection-lock the 32.768K
crystal on the LCD clock.
Much to my surprise, it worked quite well. And,
Can someone provide a sanity check on my S/N and FOM measurements? I've
used the procedure in my manual, but there is one rather vague
calculation involving 2.3 e 13 x V403b. I assume this is actually 2.3 e
-13 x (noise voltage in mv), because otherwise the other calculations
result in absurd
The cpu card in my 5371A seems to have died. It's asserting reset
constantly, with lights 1 and 6 on.
Unfortunately, my service manual doesn't include any schematics for that
card. Does anyone have any suggestions? (other than doing a lot of
tedious tracing)
--
Bill Ezell
--
They
Thanks. I don't know how I missed those manuals on the Agilent site; I
had looked there a while ago.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
Bruce Griffiths wrote:
wje wrote:
The cpu card in my 5371A seems to have died. It's asserting reset
constantly
Thanks, all. In this case, it was a 74ALS1035 OC hex buffer that gave up
the ghost.
I've temporarily replaced it with 6 diodes, which works like a charm.
After all, what's an OC buffer but a diode plus a little 1x gain?
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
My favorite for many uses is the Freescale MC68HC908QT4, or others in
that series. Freescale provides a complete C development environment
for free. The QT4 is an 8-pin package, with up to 6 I/O pins. I've used
it for everything from a 555 timer replacement to the controller for an
As the one who made the first comment about not liking the PIC, I'll
give you my reasons. Yes, they are philosophical, even religious. I'm
also distinguishing between microprocessors (this discussion) and other
variants, such as DSPs, FPLDs, etc.
First, I've used a very large number of
down. And yes, I use vi. :)
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
Scott Newell wrote:
At 07:36 AM 8/16/2008, wje wrote:
I have both EE and CS degrees, and I work in both worlds. In my humble
(but completely accurate and stable) opinion, Basic
??? (gazes at 11/70 backplane on wall...)
dave
On Sat, Aug 16, 2008 at 10:10 AM, wje [1][EMAIL PROTECTED] wrote:
You certainly don't need formal training to be a good programmer; I've
seen plenty of code from CS grads that's terrible, and very nice code
from art majors.
In my book, a good
If you install the second unit, your 5061A still won't work. +20v at the
sensor output will signal the ion current monitor to shut down the
electron multiplier supply.
If you have some high-value resistors around, can you make up a test
200M resistor string? If so, see what happens when that's
and thus the S/N output gets poor?
-Brian, WA1ZMS
-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED]
Behalf Of wje
Sent: Monday, August 18, 2008 1:43 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5061A problem
If you
Rsync works quite well on both Linux and Windows. I use it to back up
my lab computer (XP), my work computer (Vista, unfortunately), several
family computers (XP), and my primary Linux system. There's an
excellent bit of freeware, Delta Copy, that provides rsync for the
Windows
I just did an oven rebuild on mine. It can be straightforward or not,
depending upon what has failed.
However, if your unit locks and stays locked, the oven is most likely at
least partially working. The VCO loop capture range is very small and it
doesn't take a great deal of temperature
There's a small hole on the side of the case.
This is the trimmer access, at least on mine.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
EMMANOUIL MANTZARAS wrote:
Hi
Is there any one that knows where is the c-field trimer in the FE-5680A board
??
Thanks
sv1bke
and they do not have C field adjustments.
Brian - KD4FM
wje wrote:
Unless you can measure the frequency to at least a few parts in 10e10,
you won't see any change.
These devices are far more precise than standard frequency counters.
You either need something like an HP 5370 or 5371
I just got one of the new networked GPIB controllers, and I've been
having some issues. I'm not sure if it's the Prologix or the
instruments, or both.
With the 5370A, I can get samples for some period of time ranging from
about 2 to 4 hours before the GPIB controller stops responding. I
can send that to you offline.
Regards,
Abdul
-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of wje
Sent: Wednesday, August 27, 2008 3:12 PM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] Prologix GPIB-NET, HP 5370A
If the 5370A problem is actually the 5370A gpib controller bug, then I
don't think it can be reproduced with the 5370B; I'm pretty sure that
was fixed in the B model.
However, if you want to give it a try, here's what I do (you'll have to
interpret, but it should be pretty clear):
--
They said 'Windows or better'
so I used Linux.
Prologix wrote:
You could try explicitly triggering with ++trg followed by ++read 10.
-Original Message-
From: [EMAIL PROTECTED] [[2]mailto:[EMAIL PROTECTED] On
Behalf Of wje
Sent: Wednesday, August 27, 2008 4:14 PM
To: Discussion
-t will trigger
a new measurement, which is what I would expect.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
wje wrote:
That does the trick, thanks. Interestingly, the HP manual doesn't
discuss group-execute-trigger at all. It's possible that the HP-85
I have a problem with my HP 5371A. Fairly often, when I display samples
in one of the graph modes, the graph is displayed properly, but then a
number of horizontal lines show up composed of semi-regularly-spaced
dots. Typically the rows will be spaced about 1/5 of the vertical size
apart.
The dot are aligned horizontally, only a single scan line per set of
dots. However, this only shows up in graph mode, not in any other
display. The dots are actually in memory. If I plot the graph, they show
up in the plot also.
Bill Ezell
--
They said 'Windows or better'
so I used
,
depending upon the ram design).
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
WB6BNQ wrote:
Bill,
Are you sure that they are not suppose to be there ? In otherwords like a mino
r axis/division indicator ?
Just a thought.
BillWB6BNQ
wje wrote:
The dot are aligned
I'm interested in a related item - ROM images for a later version of
the firmware than I have in my 5371A. Quite a few features were added
in the late releases; I have an early set of ROMs in my
otherwise-pristine unit.
If anyone has uploaded them in any of the common programmer
I've been looking for one for ages. They're out there, but at outrageous
prices.
ManualsPlus sells a very nice reprint manual with schematics.
If you're patient, I just started building a replacement using most of
the original circuitry. I'd be happy to make board files available, or a
board
And this is another reason you should pay attention to hot and neutral.
The hot side is usually where the fuse is, as is the case in the 5061.
If you swap them, you no longer have any protection against an internal
hot-to-ground AC fault.
Bill Ezell
--
They said 'Windows or
Just a nitpick - the reading on the meter isn't any particular scale,
it's essentially arbitrary.
The actual beam current is typically in the picoamp range. My measured
beam current is only about 400 pA.
I'm not sure what you would be seeing from induction; the beam current
is a result of
I've finished my degausser replacement. I've put a zip file with
schematics, pcb layout, a picture of the finished unit, etc. on my ftp site.
Note that you have to use ftp to get to it, not a web browser:
ftp ftp.quackers.net
login: [EMAIL PROTECTED]
Use the login name exactly as given. Just
Artekmedia has a very nicely scanned manual for about $10.
There's also a copy floating around on one of the member's sites, but I
don't remember which one offhand. A Google search and some persistence
will locate it.
Along these lines, none of the manuals I've found cover the late-model
DC
I'm in Southern NH. I have GPS and Rubidium 10 Mhz sources, and a 5 Mhz
Cesium source. Le me know if you'd like to drop in, I'm near Peterborough.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
Lee Reynolds wrote:
Hi, people.
I'm in need of an extremely accurate 10MHz
Another (sometimes) relatively inexpensive outstanding reference is
either a Datron 4910 or 4912. These are also LTZ1000 based. The 10 has
one reference, the 12 has four that can be averaged or used
independently. You can on rare occasions pick one up for a few hundred.
I have a pair, and yes,
Yes, I do have full schematics for the Datron refs, including the
LTZ1000 circuitry. It isn't the same as that in the LTZ1000 data sheet.
I'll dig out my manuals. If you'd like a copy, email your postal address
and I'll copy the page and send it to you:
[EMAIL PROTECTED]
I've built a few
I don't really know how rare they are, but they're reliable.
I have mine that I built many many years ago (20+), and it's still
ticking. It's been on continuously for all that time.
There's one listed on EBay now, item 160302101690. It's already up to $150.
Bill Ezell
--
They said
It's actually [1]http://www.bb-elec.com for those looking.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
Neon John wrote:
On Mon, 24 Nov 2008 17:33:25 -0800, Tom Van Baak [2][EMAIL PROTECTED] wrot
e:
Yes. I know of several commecial systems. If you only need to
It's regular Heliax that's been temperature-conditioned by repeated
temp cycling to minimize dielectric size change vs temperature. This
minimizes phase shift over length. The cycling affects the foam used
as the dielectric, which apparently eventually stops expanding and
I've uploaded some photos of the Datron 10 volt cell and reference
board, as requested by a few members.
I've kept them full-resolution so you can see fine detail. But, that
means they are about 3 mb each.
The schematics for the cell board I previously posted are also still on
the FTP site.
I've finished retracing the LTZ ref board I promised months ago. While
it somewhat resembles the LTZ1000 data sheet example, it differs
significantly in a number of details. The heater control is far more
complex, using a PWM scheme. Take the schematic with a bit of a grain of
salt... tracing
I've posted a response for you on the HP group.
I managed to buy a Solartron 7081 at auction the other day, and assuming I
can get it working (I'm assuming it won't work when it arrives), I will need
some leads for it.
Bill
--
Bill Ezell
--
They said 'Windows or better'
so I used
-of-operation pages on it.
Thanks,
Bill
wje at quackers.net
--
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time
Fluke.l (China) was selling a number of 1938's on Ebay. I snagged one
just to have a piece of HP history.
It works just fine, but I've noticed something a little strange.
Comparing the 1938 to both my cesium and GPS standards, there's a
distinct periodic 1ns phase shift every second. Seems to
A member kindly provided me with a key generated from my serial number
for my HP 3588 to enable Basic, but he wasn't exactly sure what the
proper way was to put them on a disk so the '88 would read the key.
My serial number is 3005A01033.
I tried creating a LIF file name LD1033 and put the key
This is a pull from a Lucent unit used to provide timing for cell sites.
I have a couple myself.
The board is primarily a 10Mhz to 15Mhz converter. It also has some
failover logic. Typically two independent units were installed in one
plugin chassis.
The D connector has power in and status
Much to my amazement, I did find some notes on the DB-25 15Mhz board pinout:
1 +24V
2 +24V
3 Gnd
4 Gnd
5
6 Freq Adj (this goes directly to the FRS, the board does nothing with it)
7 Gnd
8 -Enable +
9
10 Enable - (notes are unclear on 8, 10, 11, 12. See note below)
11 To 11 on other unit
12 to
I finally gave up trying to find a 50 ohm distribution amp I liked and
wanted to pay for, so I built a simple 4 channel one that uses either a
Linear LT1365 or an Analog Devices AD8044. It has an ac-coupled input
and transformer-isolated outputs. I took care in the layout to keep
trace lengths
Per request, I've put the info up on my ftp site:
ftp://f...@quackers.net:f...@ftp.quackers.net/
Pick up DistAmp.zip. Also, read the ReadMe! It explains how to actually
look at the board layout.
I'm using it to distribute my 10 Mhz GPS-based standard (Thunderbolt) to
my spectrum analyzer,
I have a manual, but it's hard-copy and LARGE. If you can narrow down
your problem area, I'd be happy to scan some schematics.
I also have two 4910's; I can probe some test points if you need some
values.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
Steve Krull
Lux, James P wrote:
I'm looking for something that I can write python or tcl or shell scripts
for and send strings to a GPIB device, and get strings back. I'm trying to
get away from the LabView .vi model. So, this looks fairly good, that way.
Jim
I have one of the LAN units also. I've
I've already purchased the eprom board. I'll try to upload them when I
get the board.
I got it as a spare, so I can compare the firmware versions. My current
board is an early version.
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
On 01/-10/-28163 02:59 PM, Bob Camp
Although I've scanned the list fairly regularly, I haven't been subscribed.
I finally decided to rectify that.
I have Z3801 as my workhorse reference. I built a little adapter for it
to provide time and 1pps data to NTP running on my Linux system and a
buffered 10 Mhz feed to my lab for a
Congratulations! It's a wonderful instrument, isn't it?
I can answer a few of your questions.
Set/clear ref - after power-on or a reset, the ref value is (should be)
zero.
When set-ref is pressed, the current time interval average value is
stored. This remains set until a power-off/reset or you
alignment.
What I'm missing is a spectrum analyzer, so I haven't done the
frequency multiplier alignment on either of my 5370s. Any tricks for
doing that without one?
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
Chuck Harris wrote:
wje wrote:
Freq vs time
Having just gone through the process of finding, acquiring, and fixing
an HP 5061A, here are a few pointers for qualifying one for purchase.
The biggest problem you're likely to face is a beam tube that's at
end-of-life. If there is an electronics failure (like mine, read on),
these clocks are
you toss around the 'stupid' when discussing oscillator design.
Some folks still have a close relationship to those old 00105-6xxx
oscillators.
Dave
- Original Message -
From: wje [EMAIL PROTECTED]
To: time-nuts@febo.com
Sent: Friday, June 27, 2008 4:32 AM
Subject: [time-nuts
failure was because the cable fried and shorted power to ground)
Bill Ezell
--
They said 'Windows or better'
so I used Linux.
Bruce Griffiths wrote:
wje wrote:
Yes, but my comment is rather specific... my oscillator failed because
the heater pass transistor shorted.
This sent the oven
The heater winding for my 5061A crystal oven doesn't seem to be bifilar
wound, but it's a little difficult to tell. Since the heater works, I
don't want to risk poking around too much. However, one lead comes in
at one end, the other lead is at the other end.
BTW, my oven failed
Also, check the Cs oven meter reading. If it's zero, the heater
controller/heater/thermocouple is suspect. Normally, when you turn the
unit on, the Cs oven will read up near the top of scale, gradually
declining as the oven warms up. Mine goes down to about 10 steady-state.
If it's doing this,
65 matches
Mail list logo