Buy another one, they are so cheap :-)

Out of the last lot of five that I bought from the same source, four work very well and one is marginal - sometimes it locks but it's noisy, sometime does not lock. Lamp and xtal voltage are ok, but it jumps up and down by 1E10. Very positive overall, IMHO.

73 - Marco IK1ODO


At 22.08 02/02/2010, you wrote:
Hello all.

I received an EFRATOM LPRO-101 last week from eBay Fluke 1, and wired it up last night. The results are not good.

Using a HP 5370B counter with calibration less than one year old, the frequency reads = 10,000,146,012.9
Power Applied = 24.9Vdc @ 0.30 amps after warm up
Lamp Voltage = 5.9 Vdc
Ext C Field = 2.437
XTL V Mon = 14.67
Bite = 4.57

Bite never goes low.

I just sent an email to the outfit I purchased it from, but would like to get this working instead of waiting a month or so.

Is it dead?  Any thing I can do to revive it?

Any suggestions?

Thanks to all.

Jim N0OBG


_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to