Date: Wed, 3 Feb 2010 12:37:08 -0500
From: "Bob Camp" <[email protected]>
Subject: Re: [time-nuts] EFRATOM LPRO-101
To: "'Discussion of precise time and frequency measurement'"
<[email protected]>
Message-ID: <[email protected]>
Content-Type: text/plain; charset="iso-8859-1"
Hi
I would second that as well.
I got an LPRO in that had a "rattle" when it arrived. After some thought, I
popped the can and figured out what the issue was. Since it might have been
gong back, I kept Bob in the loop during the entire process. He was very
supportive through it all. Definitely a first class guy.
Bob
-----Original Message-----
From: [email protected] [mailto:[email protected]] On
Behalf Of Doug Parker
Sent: Wednesday, February 03, 2010 9:23 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] EFRATOM LPRO-101
Hi,
?
I'd like to confirm that Bob is a very good seller and been great with the
things I've had off him.
?
He's also a member of this group
?
Doug
G4DZU
--- On Wed, 3/2/10, EB4APL <[email protected]> wrote:
From: EB4APL <[email protected]>
Subject: Re: [time-nuts] EFRATOM LPRO-101
To: "Discussion of precise time and frequency measurement"
<[email protected]>
Date: Wednesday, 3 February, 2010, 13:20
Hi,
Bob Mokia is a very good seller and you'll not have problems with him.?
In? order to avoid paying the return shipping and the delays I would ask
him to let you open the unit and try to find an obvious fault such a
loose connection.? When I bought my Rb (from another source) I opened it
even it was working ok, and I found one of the connector pins touching
its PCB pad without any trace of solder on it; nevertheless it worked
perfectly but it was just a question of time before the unit will begin
to act intermittently so I fixed it.? Thinking about I realized that
maybe this unit showed problems and it was retired from active service
and maybe it has very low hours on it.
Do your unit sweeps its frequency above and bellow 10 MHz when warming
up the first minutes? it is the normal operation before a lock is
obtained and if it is not done it will never lock the oscillator to the
Rb frequency.
Good luck,
Ignacio, EB4APL
-----------------------------------------------------------------------
Bob Camp wrote:
Hi
A little more data from a group of recent units:
The XTAL pin is looking at the VCXO control voltage, It should move around
(sweep) at turn on. It will settle in the vicinity of 6 volts. I have seen
units anywhere from 5.5 to 7.2 volts on this pin.
The BITE is a logic level pin, it's either high or low.
The lamp voltage on "early" units runs around 6 volts. On "later" units
it's up around 9 volts. This may be a better bulb, or it just may be a
different photo detector. Every time I've asked about lamp voltage, the
answer has been "in range is ok". Trying to predict life from lamp voltage
apparently is more complex than it might seem.? I would not throw out a 6
volt unit simply because it's not a 9 volt ....
Bob
On Feb 3, 2010, at 7:08 AM, Roy Phillips wrote:
???
Hi Jim
This would appear to be a faulty unit, I also purchased the same unit
from Bob Mokia two weeks ago and its performing very well.? It gives a
frequency of 10.000,000,005 Mhz.
One very obvious fault is the BITE voltage - - this should initially be
4.5 volts, which falls to less than 30 mV. within a few minutes. The XTAL
Control Volts, after a short while falls to 6.145 V., so yours, at 14.67 V,
would seem to be wrong ! The Lamp volts settle to 6.736 V, and the Ext? C
Field is 2.426 V. , which checks with your result.
I'm? sure you will get a replacement.
Roy
--------------------------------------------------
From: "Jim" <[email protected]>
Sent: Tuesday, February 02, 2010 9:08 PM
To: <[email protected]>
Subject: [time-nuts] EFRATOM LPRO-101
? ???
Hello all.
I received an EFRATOM LPRO-101 last week from eBay Fluke 1, and wired it
up last night. The results are not good.
Using a HP 5370B counter with calibration less than one year old, the
frequency reads = 10,000,146,012.9
Power Applied = 24.9Vdc @ 0.30 amps after warm up
Lamp Voltage = 5.9 Vdc
Ext C Field = 2.437
XTL V Mon = 14.67
Bite = 4.57
Bite never goes low.
I just sent an email to the outfit I purchased it from, but would like
to get this working instead of waiting a month or so.
Is it dead?? Any thing I can do to revive it?
Any suggestions?
Thanks to all.
Jim N0OBG
--